From: DON CRAMER
Sent: Monday, November 01, 2004 2:06 AM
Subject: [hp_agilent_equipment] E4411A spectrum analyzer low gain fixed

I thought I should document repair to an E4411A (alias ESA-L1500A) 1.5Ghz spectrum analyzer in the event anyone here might find the information useful.
?
The symptom was low gain (~28dB low) but with flat response and all other functions ok.? The HP service manual only goes down to the module level, which is doubly unfortunate for this instrument because the one RF assembly module contains everything from the front panel input connector all the way back to the 2nd IF input.? The manual helped confirm the fault was in the RF assembly module, but at that point they say don't service it, get a replacement instead.? That would be ok except the replacement price is $6120 exchange.? For an analyzer that probably went for $8K new, I bet they don't get too many takers.? As the SA was initially free, I didn't have too much to lose by violating the recommendation and taking a peek inside the RF assembly.
?
The clever RF assembly comprises one PCB, with castings that sandwich the board in between providing circuit isolation.? The service manual does give some information as to what's in there via a coarse block diagram and a page or two of explanation.? I was able to find a shorted transistor (Q6) in what I believe is the 2nd mixer (could be the tail end of the 1st IF).? This was located using the diode function of a DMM; I did not have to power up the assembly.? Normal was about a 0.45V diode drop from "dot" (I thought base, but turned out to be gate) to the other leads.
?
The transistor is a "pill package" 4 lead part with "211" as the sole ID marking.? Some hunches lead to the identity of the part being most likely an HP (Agilent) ATF-21186 GaAs FET, which is used in numerous locations in the front end of the RF assy.? Regarding the 0.45V drop, what I was measuring I believe is a schottky barrier gate structure.? I don't work in RF; FETs in my line are all insulated gate (enhancement mode) devices, so this has been an interesting experience.
?
The bad news was the ATF-21186 was discontinued in 2001.? Newer parts don't use the pill package.? However, I did find a small lot of '21186 parts (9805 lot code) via a helpful broker at about $14 each, with a required lot buy of 16 pieces.? My analyzer is SN prefix "US37.." so the parts are less than a year apart in manufacture date.? Upon arrival I was pleased to find the new parts looked identical to the old.? BTW, a "-STR" suffix on the ATF-21186 designates the part packaged in a strip of 10.?
?
I did not attempt to "match" devices or anything--I just replaced the bad part with one of the new ones.? Input welcome if someone knows better.
?
One tool I found helpful in putting the RF assy castings back together with the zillion M3 screws used is a good torque screwdriver to get everything down evenly.? The tightening torque is 9 in-lbs on everything (casting screws, casting to instrument screws, and instrument case screws).? As there is no give to the housings, it is crucial all screws get tightened down the same amount.? Over tightening one screw will make the adjacent one loose for instance.? It takes two of three passes around them all to get to final torque.? I used a Proto 6106 (now apparently owned by Stanley) driver which supposedly will not over torque screws even with multiple attempts to retighten them.
?
The replacement transistor seems to have solved the problem.
?
One other hint.? The one?piece?rubber membrane key pad has thin walls where the raised key caps drop down to the sheet.? They are easily torn if you do something silly like drive the key cap sideways through overzealous cleaning for instance.? I found a reasonable fix was to build a "hinge" between the torn key cap and base sheet using a flowable silcone product made for car windshield gasket sealing: Permatex Windshield and Glass Sealer.??I found a?1.5oz tube of the stuff available at the local hardware store.
?
Don C
Beaverton, OR
?
?
?
?


E4411A spectrum analyzer low gain fixed

DON CRAMER
 

开云体育

I thought I should document repair to an E4411A (alias ESA-L1500A) 1.5Ghz spectrum analyzer in the event anyone here might find the information useful.
?
The symptom was low gain (~28dB low) but with flat response and all other functions ok.? The HP service manual only goes down to the module level, which is doubly unfortunate for this instrument because the one RF assembly module contains everything from the front panel input connector all the way back to the 2nd IF input.? The manual helped confirm the fault was in the RF assembly module, but at that point they say don't service it, get a replacement instead.? That would be ok except the replacement price is $6120 exchange.? For an analyzer that probably went for $8K new, I bet they don't get too many takers.? As the SA was initially free, I didn't have too much to lose by violating the recommendation and taking a peek inside the RF assembly.
?
The clever RF assembly comprises one PCB, with castings that sandwich the board in between providing circuit isolation.? The service manual does give some information as to what's in there via a coarse block diagram and a page or two of explanation.? I was able to find a shorted transistor (Q6) in what I believe is the 2nd mixer (could be the tail end of the 1st IF).? This was located using the diode function of a DMM; I did not have to power up the assembly.? Normal was about a 0.45V diode drop from "dot" (I thought base, but turned out to be gate) to the other leads.
?
The transistor is a "pill package" 4 lead part with "211" as the sole ID marking.? Some hunches lead to the identity of the part being most likely an HP (Agilent) ATF-21186 GaAs FET, which is used in numerous locations in the front end of the RF assy.? Regarding the 0.45V drop, what I was measuring I believe is a schottky barrier gate structure.? I don't work in RF; FETs in my line are all insulated gate (enhancement mode) devices, so this has been an interesting experience.
?
The bad news was the ATF-21186 was discontinued in 2001.? Newer parts don't use the pill package.? However, I did find a small lot of '21186 parts (9805 lot code) via a helpful broker at about $14 each, with a required lot buy of 16 pieces.? My analyzer is SN prefix "US37.." so the parts are less than a year apart in manufacture date.? Upon arrival I was pleased to find the new parts looked identical to the old.? BTW, a "-STR" suffix on the ATF-21186 designates the part packaged in a strip of 10.?
?
I did not attempt to "match" devices or anything--I just replaced the bad part with one of the new ones.? Input welcome if someone knows better.
?
One tool I found helpful in putting the RF assy castings back together with the zillion M3 screws used is a good torque screwdriver to get everything down evenly.? The tightening torque is 9 in-lbs on everything (casting screws, casting to instrument screws, and instrument case screws).? As there is no give to the housings, it is crucial all screws get tightened down the same amount.? Over tightening one screw will make the adjacent one loose for instance.? It takes two of three passes around them all to get to final torque.? I used a Proto 6106 (now apparently owned by Stanley) driver which supposedly will not over torque screws even with multiple attempts to retighten them.
?
The replacement transistor seems to have solved the problem.
?
One other hint.? The one?piece?rubber membrane key pad has thin walls where the raised key caps drop down to the sheet.? They are easily torn if you do something silly like drive the key cap sideways through overzealous cleaning for instance.? I found a reasonable fix was to build a "hinge" between the torn key cap and base sheet using a flowable silcone product made for car windshield gasket sealing: Permatex Windshield and Glass Sealer.??I found a?1.5oz tube of the stuff available at the local hardware store.
?
Don C
Beaverton, OR
?
?
?
?


Re: Can't ID Equipment

gerbo04
 

I have looked at pictures of the 5345A and the dimensions all match
up. Thanks for the ID JM. I got so many google hits on 05345-XXXXX
that I was led to believe that was a generic prefix for HP boards.

It appears that it does not have the HPIB option, there is one slot
open and there are two aluminum blank cover plates on the back where
perhaps the connector would likely be. A couple of the boards have
many custom gold cover HP ICs, all socketed. Anyone wants parts let
me know. (Never heard it called a 'hot rock box', I like that.)

It also has:

05345-60031 88802F PLL Buffer Multiplier
05345-60045 88802F
05345-60007 PS
05345-60006 PS

(I need a knob or two for a 1720A-family scope and an 8620C generator
if anyone wants to swap stuff.)

Thanks again,
Gerry W.


--- In hp_agilent_equipment@..., Scott Newell
<newell@c...> wrote:
At 04:30 AM 10/30/2004 +0000, gerbo04 wrote:

kind? The fan is in the front which is odd. The boards are labeled
as
follows:

05345-60044 88802F
05345-60013 744414
05345-60011 88802F
05345-60050 88809F
05345-60009 73803L

Can anyone identify this thing?
As JM explained, it's certainly a 5345A.

If you do decide to part it out, let me know...I'd be interested in
the A12
(HPIB) board (which you may not have--it's not on the list you
posted), a
couple of EECL chips to fix my 5370, and maybe the hot rock box.

--
newell N5TNL


Re: 3406A sampling meter/peeling faces

 

开云体育

You could use a paint on mask that you can get at hobby shop to protect the mirror.

-----Original Message-----
From: daveolla [mailto:grobbins@...]
Sent: Saturday, October 30, 2004 5:21 PM
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] 3406A sampling meter/peeling faces


Greetings, recieved a 3406A sampling meter free, now I need to
brushup on? what it is good for,? and what benefits from sampling as
opposed to regular RF meters.
The question is concerning those Hp meter faces from a certain era
that peel and crack at the mirror ends (glued so well to brass
mirrored back and contraction forces). There is a clear film that
flakes up and off over the lettering and scale, the flaking stops at
the letters, as if that helps hold it down. The stuff is water
soluable, so if it is wiped it smears. My 428B meter also had this
same problem except letters below also were flaking, had to photo
copy a new one, a pain since I tried dozens of copiers around town
and found most could not copy at 100%,(laid ruler along side
meter face when copied so as to compare ), also cutting out a nice
mirror slot again is shunned.
The flaking has been nicely brushed off and I would like to overspay
(masking the mirror) with something (even water will remelt the film)
fast drying to prevent trouble below, wrinkling etc. Anyone else have
ideas to save face. Iron on films are out of the question since heat
on my 428B meter face caused fine cracks everywhere like fine art
paintings.
Hopefully that is all that is wrong with this meter since the lifting
film stopped the needle from moving......... unless this occurred
after the probe was found to be bad.
Dave Robbins
And In the Halloween spirit, might I say, BOO






Re: 3406A sampling meter/peeling faces

 

Greetings, recieved a 3406A sampling meter free, now I need to
brushup on what it is good for, and what benefits from sampling as
opposed to regular RF meters.
The advantage of the sampler is that it has good linearity over a wide
dynamic range, unlike a diode detector, which goes from approximately
linear to approximately square law.

I don't have any advice on the meter face. Maybe it would be best to
connect a DPM in place of the analog meter.

David DiGiacomo


3406A sampling meter/peeling faces

 

Greetings, recieved a 3406A sampling meter free, now I need to
brushup on what it is good for, and what benefits from sampling as
opposed to regular RF meters.
The question is concerning those Hp meter faces from a certain era
that peel and crack at the mirror ends (glued so well to brass
mirrored back and contraction forces). There is a clear film that
flakes up and off over the lettering and scale, the flaking stops at
the letters, as if that helps hold it down. The stuff is water
soluable, so if it is wiped it smears. My 428B meter also had this
same problem except letters below also were flaking, had to photo
copy a new one, a pain since I tried dozens of copiers around town
and found most could not copy at 100%,(laid ruler along side
meter face when copied so as to compare ), also cutting out a nice
mirror slot again is shunned.
The flaking has been nicely brushed off and I would like to overspay
(masking the mirror) with something (even water will remelt the film)
fast drying to prevent trouble below, wrinkling etc. Anyone else have
ideas to save face. Iron on films are out of the question since heat
on my 428B meter face caused fine cracks everywhere like fine art
paintings.
Hopefully that is all that is wrong with this meter since the lifting
film stopped the needle from moving......... unless this occurred
after the probe was found to be bad.
Dave Robbins
And In the Halloween spirit, might I say, BOO


Re: Can't ID Equipment

John Miles
 

Hi, Scott --

What do you need for the 5370? I have a 5370B parts mule around here.

-- john KE5FX

-----Original Message-----
From: Scott Newell [mailto:newell@...]
Sent: Saturday, October 30, 2004 8:50 AM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] Can't ID Equipment



At 04:30 AM 10/30/2004 +0000, gerbo04 wrote:

kind? The fan is in the front which is odd. The boards are labeled as
follows:

05345-60044 88802F
05345-60013 744414
05345-60011 88802F
05345-60050 88809F
05345-60009 73803L

Can anyone identify this thing?
As JM explained, it's certainly a 5345A.

If you do decide to part it out, let me know...I'd be interested
in the A12
(HPIB) board (which you may not have--it's not on the list you posted), a
couple of EECL chips to fix my 5370, and maybe the hot rock box.

--
newell N5TNL





Yahoo! Groups Links








Re: Can't ID Equipment

Scott Newell
 

At 04:30 AM 10/30/2004 +0000, gerbo04 wrote:

kind? The fan is in the front which is odd. The boards are labeled as
follows:

05345-60044 88802F
05345-60013 744414
05345-60011 88802F
05345-60050 88809F
05345-60009 73803L

Can anyone identify this thing?
As JM explained, it's certainly a 5345A.

If you do decide to part it out, let me know...I'd be interested in the A12
(HPIB) board (which you may not have--it's not on the list you posted), a
couple of EECL chips to fix my 5370, and maybe the hot rock box.

--
newell N5TNL


Re: Can't ID Equipment

John Miles
 

That's a 5345(A?) counter. Reciprocal (computing) counter, one of the best
ever made. Good to 500 MHz, 18 GHz+ with plugins. Crystal oscillator will
be a 10545 or 10811; either are very good-quality standards.

Unit as a whole is not good for anything without the front panel.

-- john KE5FX

-----Original Message-----
From: gerbo04 [mailto:gjwinga@...]
Sent: Friday, October 29, 2004 9:30 PM
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Can't ID Equipment




I came across a piece of HP gear that has the entire faceplate and
front frame missing, got it cheap. Can't identify what it is. It is a
mainframe type package with an empty space on the right side for a
plug-in. Total about 16" wide 5" high and 20" deep. Inside there are
seven plug in boards (eight slots), a big crystal in an oven, and an
impressive power supply and fan. Maybe a signal generator of some
kind? The fan is in the front which is odd. The boards are labeled as
follows:

05345-60044 88802F
05345-60013 744414
05345-60011 88802F
05345-60050 88809F
05345-60009 73803L

Can anyone identify this thing?

Gerry W.







Yahoo! Groups Links








Can't ID Equipment

gerbo04
 

I came across a piece of HP gear that has the entire faceplate and
front frame missing, got it cheap. Can't identify what it is. It is a
mainframe type package with an empty space on the right side for a
plug-in. Total about 16" wide 5" high and 20" deep. Inside there are
seven plug in boards (eight slots), a big crystal in an oven, and an
impressive power supply and fan. Maybe a signal generator of some
kind? The fan is in the front which is odd. The boards are labeled as
follows:

05345-60044 88802F
05345-60013 744414
05345-60011 88802F
05345-60050 88809F
05345-60009 73803L

Can anyone identify this thing?

Gerry W.


HP 8754A Operation Guide - Have depot maintenance manual...

Gary
 

Hello everyone;

I have the 8754A Network Analyzer manual, including the maintenance
and alignement and so on... about 4 inches thick, however, it is
missing the operators section. Likely someone pulled it out of the
binder because it was used so frequently.

If you have an operators guide that covers procedures for various test
setups, I'd appreaciate a copy... I do have a scanner, however the
whole manual is 4 inches thick; I'd be happy to provide copies of a
portion on a trade.

Gary
NZ5V


Re: Finding HP/Agilent Manuals and parts

Dave Brown
 

Tks Walter- URL survived the posting OK-pity re the 54200A though!
Cheers
DaveB, NZ


Finding HP/Agilent Manuals and parts

wshawlee2
 

The highly sought after 54200A manual is not at the following site,
but many other HP/Agilent items ARE. It's worth some time to get
familiar with this part of the Agilent site, as it does contain many
very useful PDF manuals for new and old equipment. I notice many of
these being sold on ebay, but you can just download them for free here:



it is a huge URL, so I hope it survives yahooization. if you have
trouble, email, me, I will send it to you directly.

many more manuals for olde rgear are available on the LOGSA site (link
below), and there are many dealers with manuals (printed). have a
look at these pages for stuff on our site and elsewhere:



you can also find HP parts cross-reference data on our site here:


I have a pretty extensive Tektronix service resource page on line here:


I have been trying to make time to do a similar page for HP, but too
much work right now. Plus, there is not as much user group activity
for these products, so there's not as much info to post there.

all the best,
walter


Re: 8662A and VCOs

John Miles
 

Ooof. I don't think you want to use those in the same room as an 8662A.
Noise city.

-- john KE5FX

Mike:

If you continue to have trouble with those VCOs, have you considered
an update with the "VCO on a chip" approach? A while back I had a
bit of a "wrestling match" with a Systron Donner freq. synthesizer
that refused to lock reliably, and I refused to lose that match.

A company in Philadelphia called Amplifonix makes some nice VCOs(if
a bit pricey) .
Another approach would be to use one of the Analog Devices PLL chips
that has the on-board VCO, and just use its VCO "core." The ADF4360-
7 looks like it might do the trick for you (if you ignore its
programmable PLL sections, you may need to throw in an ECL flip flop
to divide its output by two). If you go the Analog Devices route
may I also recommend the KOA KQ line of surface mount inductors (I
have found that they approach my s parameter simulator's
(Ansoft) "ideal" inductor up to around 2 GHz in a number of
circuits).


54200A gpib

Bob Japundza
 

Thomas, I managed to find my NI AT-GPIB/TNT card and got it working
(I think.) Actually the driver setup from NI is pretty good at
telling you the correct jumper setup. When clicking on the
autoscale button inside your application, I get the following error
dialogs:
Unable to write to device ibsta=$8100 iberr = 2
lbsta ===> ERR CMPL
iberr ===> ENOL

Any ideas anyone? In the NI gpib explorer the scope shows up in the
list, and will beep when I try to send a command thru the NI
interface, but the scope will say WARNING: waiting to recieve data,
then the NI program will time out. I have a suspicion that I am
able to send commands to the scope but cannot read the responses
from the scope.

Regards,
Bob


HP 8591a

 

Hello all
Have been lucky to get a HP 8591a spectrum analyzer.
Would really like to have some information about this fine instrument-
the agilent home page seems only to help with 8591e.Not sure on the
differences.
Have You information on this-user guide-service guide??
Yours
Hardy


Re: 8662A and VCOs

 

开云体育

Thanks for all of the suggestions. I mainly bought the 8662A for it’s low phase noise characteristics, so I could use it in a phase noise test set-up. So, I would like to leave it with it’s original design as those VCOs were designed to be very low noise items. You guys should see how they are assembled. Each unit is wrapped in a mu-metal ?shield and the construction is fantastic, especially considering the vintage. I have not given up, just taking a break from it. Still hoping to find someone who knows something about the alignment of them. I believe the varicaps are OK in both units. I think what happened initially was that in one of the units one of the 2N5397 JFETs died and then substituted with a 2N4416, but, not aligned properly after the substitution and then the JFETs were also no longer matched. I would like to find a dozen or so 2N5397s but have had no luck so far. I would settle for a beat to heck 8662A with two good intact oscillators (VCOs). Each VCO covers from about 320 to 640 MHz and has 5 or 6 pin diode switched inductors driven by the control logic to select the proper inductor for a given range. There are three miniature pots on each VCO. One I believe controls the voltage to the FETs and each of the other pots controls the current through each FET to set the proper operating point. I sure would like to finish this project up as I have a lot of other ones waiting in the queue, but, it may be a while. Thanks for all the ideas. Please keep them coming. Regards - Mike ??

?

?

Mike B. Feher, N4FS

89 Arnold Blvd.

Howell, NJ, 07731

732-901-9193

?

?

-----Original Message-----
From: phila_renewal [mailto:phila.renewal@...]
Sent: Friday, October 29, 2004 7:32 AM
To:
hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 8662A and VCOs

?


Mike:

If you continue to have trouble with those VCOs, have you considered
an update with the "VCO on a chip" approach?? A while back I had a
bit of a "wrestling match" with a Systron Donner freq. synthesizer
that refused to lock reliably, and I refused to lose that match.

A company in Philadelphia called Amplifonix makes some nice VCOs(if
a bit pricey)
Another approach would be to use one of the Analog Devices PLL chips
that has the on-board VCO, and just use its VCO "core."? The ADF4360-
7 looks like it might do the trick for you (if you ignore its
programmable PLL sections, you may need to throw in an ECL flip flop
to divide its output by two).? If you go the Analog Devices route
may I also recommend the KOA KQ line of surface mount inductors (I
have found that they approach my s parameter simulator's
(Ansoft) "ideal" inductor up to around 2 GHz in a number of
circuits).

Last but not least, to take a look at some very creative alternative
VCO circuits and layouts, the older Wavetek sweep generators (I'm
thinking specifically of, I belive it was the 2002A) use wideband
VCOs to get the sweeps (OK, I guess what else would you use to
generate sweeps -- but anyway if you can get a look at a manual for
one of those old generators, they have some interesting ideas).

My "update" circuits end up looking crude because, so far at least,
once I get them working, I leave them as is (I prototype mostly
using the "Jim Williams" solder to the ground plane and let
everthing else hang just above it approach, or using small boards
with manufacturer's suggested layout for particular parts but then
there are always at least a couple more parts, again, just soldered
above the ground plane), but I've always thought that updating could
be a matter of making a small custom circuit board that would be
like your own "hybrid" in essence to replace the functional block
you are updating.

Best of luck,

-Keith

--- In hp_agilent_equipment@..., "factory_peasant"
wrote:
>
> the VCOs are hand built and tested on the line, so they had to be
> sent back into the factory for repair or replacement. through the
> late 90s many of the components vital to these VCOs like the glass
> pin diodes were becoming very difficult to find and were of poor
> quality. your best bet will be to try and locate another of the
VCOs
> from a different unit. they are all identical regardless of 8662A
or
> 8663A. each unit required an identical pair of the VCOs so there's
a
> few of them out there to be sure.
>
>
> --- In hp_agilent_equipment@..., "Mike Feher"

> wrote:
> > I am trying to get my 8662A to work. I bought it knowing that it
had
> > problems.? I now have most of it working but believe that one of
> the VHF
> > (320-640 MHz) VCOs is not behaving the way it should. There are
two
> of these
> > VCOs in the unit, essentially identical. They use a pair of
2N5397
> N channel
> > JFETs in an oscillator and it is varactor tuned. There is no
info
> on them at
> > all in the manuals as it is a factory repair part only. I wonder
if
> any one
> > on here has any experience with them that they could share.
Thanks -
>? Mike??
> >
> >?
> >
> >?
> >
> > Mike B. Feher, N4FS
> >
> > 89 Arnold Blvd.
> >
> > Howell, NJ, 07731
> >
> > 732-901-9193






NEED A PAGE FROM 3456A MANAUL

NR1DX
 

Any body out there with a 3456A manual?

I need a copy of pages 8-13 & 8-14 I recently came by a copy of this manual and this one page is missing.I'll be happy to send a SASE for a paper copy or an emailed scan would also do the trick

Thanks
Dave
NR1DX





Dave
NR1DX
nr1dx@...

"A man who picks a cat up by the tail learns a lesson he can learn no other way" .... Mark Twain


Re: 8662A and VCOs

phila_renewal
 

Mike:

If you continue to have trouble with those VCOs, have you considered
an update with the "VCO on a chip" approach? A while back I had a
bit of a "wrestling match" with a Systron Donner freq. synthesizer
that refused to lock reliably, and I refused to lose that match.

A company in Philadelphia called Amplifonix makes some nice VCOs(if
a bit pricey) .
Another approach would be to use one of the Analog Devices PLL chips
that has the on-board VCO, and just use its VCO "core." The ADF4360-
7 looks like it might do the trick for you (if you ignore its
programmable PLL sections, you may need to throw in an ECL flip flop
to divide its output by two). If you go the Analog Devices route
may I also recommend the KOA KQ line of surface mount inductors (I
have found that they approach my s parameter simulator's
(Ansoft) "ideal" inductor up to around 2 GHz in a number of
circuits).

Last but not least, to take a look at some very creative alternative
VCO circuits and layouts, the older Wavetek sweep generators (I'm
thinking specifically of, I belive it was the 2002A) use wideband
VCOs to get the sweeps (OK, I guess what else would you use to
generate sweeps -- but anyway if you can get a look at a manual for
one of those old generators, they have some interesting ideas).

My "update" circuits end up looking crude because, so far at least,
once I get them working, I leave them as is (I prototype mostly
using the "Jim Williams" solder to the ground plane and let
everthing else hang just above it approach, or using small boards
with manufacturer's suggested layout for particular parts but then
there are always at least a couple more parts, again, just soldered
above the ground plane), but I've always thought that updating could
be a matter of making a small custom circuit board that would be
like your own "hybrid" in essence to replace the functional block
you are updating.

Best of luck,

-Keith

--- In hp_agilent_equipment@..., "factory_peasant"
<factory_peasant@y...> wrote:

the VCOs are hand built and tested on the line, so they had to be
sent back into the factory for repair or replacement. through the
late 90s many of the components vital to these VCOs like the glass
pin diodes were becoming very difficult to find and were of poor
quality. your best bet will be to try and locate another of the
VCOs
from a different unit. they are all identical regardless of 8662A
or
8663A. each unit required an identical pair of the VCOs so there's
a
few of them out there to be sure.


--- In hp_agilent_equipment@..., "Mike Feher"
<n4fs@e...>
wrote:
I am trying to get my 8662A to work. I bought it knowing that it
had
problems. I now have most of it working but believe that one of
the VHF
(320-640 MHz) VCOs is not behaving the way it should. There are
two
of these
VCOs in the unit, essentially identical. They use a pair of
2N5397
N channel
JFETs in an oscillator and it is varactor tuned. There is no
info
on them at
all in the manuals as it is a factory repair part only. I wonder
if
any one
on here has any experience with them that they could share.