Re: HP 8590 e series card cage
Could not find 08590-60532, but the 1251-8912 is TE Connectiy 5-534204-5 Hope this helps [email protected]> wrote:
By
Kenneth Harrison <bibpakak507@...>
·
#132987
·
|
Re: Agilent N5230A Power Supply (0950-4599) repair: Looking for Schematic & general repair advice
Can't provide too much, but typical failure on these is an ectrolytic near the AC input. It will cause the symptoms described. [email protected]> wrote:
By
Kenneth Harrison <bibpakak507@...>
·
#132986
·
|
Re: HP 8590 e series card cage
Ken, Thanks ,as follows. 08590-60532 I believe that the connector on the card cage is the male and is a Amphenol 30x2 The part number from the CLIP is connector HP part 1251-8912 -this will be the
By
Mike Lockley
·
#132985
·
|
Re: HP 8591A 60x2 connector HP
Dear Mike, No doubt I either have the 60x2 connector or can source one quickly. Please contact me today at WA1ESO@... John WA1ESO
By
JP Nery
·
#132984
·
|
Re: HP 8594E low signal level indicated
I was also going to point out the 8594E has no YTF or SYTF.? Unfortunately HP tried to re-use the same block diagram for too many models. I wanted to add that cross-referencing the assembly numbers
By
Mark Litwack
·
#132983
·
|
Re: HP 8594E low signal level indicated
The block diagram in the service manual is confusing because it includes many items that are not in my unit. So, I am pausing in my troubleshooting to create my own block diagram to reflect how my
By
Rik Rasmussen WA4BAN
·
#132982
·
|
Re: HP transistor equivalent for 1854-0247?
1854-0247 TI SA-6325 2N3866A wrote:
By
Kenneth Harrison <bibpakak507@...>
·
#132981
·
|
Re: HP 8594E low signal level indicated
------ Hi Rik, There is no YTF on a 8594E, block diagram shows that with a dotted box. Mixer is also single band. Greg's suggestion of looking at after the first mixer is good. One slight correction
By
Ozan
·
#132980
·
|
Re: HP 8590 e series card cage
Get me the board part number, and I will see what I can find [email protected]> wrote:
By
Kenneth Harrison <bibpakak507@...>
·
#132979
·
|
HP 8590 e series card cage
Does anyone out there know what 30x2 connector fits into the card cage connector receptacle on the HP 8590 series ( and probably others) SA ? I have a option 130 narrow BW pcb which i am working on
By
Mike Lockley
·
#132978
·
|
Re: HP 8594E low signal level indicated
Bruce, Can you see the 0hz LO level¡ªwhat is it¡ªshould be around 0-to -10db ? Mixers are not as easily shot as in the 8591 SA so that is good-but it is a possibility The 70 db front end attenuators
By
Mike Lockley
·
#132977
·
|
Re: HP 8594E low signal level indicated
Rik, It sounds like your problem is a little deeper inside the instrument But if it ever turn out to be the step attenuator (HP part # 33321-60036) this experience might offer some past dealings with
By
Greg Muir
·
#132976
·
|
Re: HP 8594E low signal level indicated
OK - First confirm the gain of the YTF. There should be an expected signal level specified and the expected output (input to the next stage) - you should be able to calculate the stage gain; then
By
Bruce
·
#132975
·
|
Re: HP 8594E low signal level indicated
When I said I was feeding the First Converter, that was apparently icorrect. I was looking at the wrong block diagram. I am now looking at the block diagram for the 8594E. The block diagram does
By
Rik Rasmussen WA4BAN
·
#132974
·
|
Re: HP 8594E low signal level indicated
Reasonable - 1st converter would be suspect, but can you inject signal into the 2nd converter (at appropriate level) to confirm low 1st converter gain? The CLIP should give expected signal level at
By
Bruce
·
#132973
·
|
Re: looking for HP85101C display keys plastic frame
Nice 3D model too!
By
John Griessen
·
#132972
·
|
Re: HP 8594E low signal level indicated
I am feeding the CAL OUT (300 mHz -20 dBM) directly to the input of the First Converter, using the hardcable that normally connects to the output of the attenuator. The HP-8594E is showing the signal
By
Rik Rasmussen WA4BAN
·
#132971
·
|
Re: HP 8594E low signal level indicated
Sounds reasonable - I would "work" the attenuator a bit to be sure it is reliable - these attenuators can be erratic. Cheers! Bruce Quoting Rik Rasmussen <rikrasmussen@...>:
By
Bruce
·
#132970
·
|
Re: HP 8594E low signal level indicated
I have the CAL OUT connected to the front panel INPUT, with the 300 mHz -20 dBm signal. I have my tinySA Ultra connected to the output connector on the attenuator. The signal is reading -24 dBm on the
By
Rik Rasmussen WA4BAN
·
#132969
·
|
Re: HP 8594E low signal level indicated
If you elect to inject a signat AFTER rather that THROUGH the attenuator be very careful - you risk damaging the first mixer. It is much safer to disconnect the attenuator input (sig gne side), inject
By
Bruce
·