Keyboard Shortcuts
ctrl + shift + ? :
Show all keyboard shortcuts
ctrl + g :
Navigate to a group
ctrl + shift + f :
Find
ctrl + / :
Quick actions
esc to dismiss
Likes
- HP-Agilent-Keysight-Equipment
- Messages
Search
Re: Suffering from RIFA anxiety!
On 2022-09-15, at 07:41, Tom Lee <tomlee@...> wrote:
To me it seems the most relevant aspect of moisture is that it goes into the paper that is part of the metallized paper construction of the safety capacitors. Whether that (the effect of the moisture itself or of the resulting increases in capacitor loss) in turn increases pressure and leads to further cracking (or whether moisture is contributing to the cracking of the epoxy directly) I don¡¯t know. My mental model was treating the shrinkage cracking and the moisture ingress as a simple sequence of causations for the incidents we are seeing. In the end, though, does it actually matter why the caps crack? Isn't it sufficient simply to know that they do crack?It may not matter, but enquiring minds still want to know. People who are risk takers and want to keep the bad caps in place for devices that always have been in dry air (and, maybe, are further risk- or damage-reduced by only being used with 110 V), knowing more about the process may even be of further interest. Since RIFA/KEMET (certainly not WIMA) inexplicably appears to be still making caps with the same construction, and we don¡¯t know what is in current Schaffner filters, the impact may not be limited to the 1980s/1990s vintage. Gr¨¹?e, Carsten |
Re: Suffering from RIFA anxiety!
My understanding is the opposite of yours. Despite being hygroscopic, epoxy can and does shrink. I once used epoxy as support structures in some coaxial capacitive sensors for inline process monitoring of injection mo(u)lding, and the dimensional change was readily measurable. I made measurements only because of failures in the field.
toggle quoted message
Show quoted text
I don't claim to know how shrinkage and moisture uptake could be compatible, but I suspect that moisture-induced changes (increases) in cross-linking may have something to do with it, in much the same way that heat in heatshrink works. As I am four years shy of a chemistry degree, though, you may rightfully take issue with that hypothesis. But the shrinkage is an empirical fact. In the end, though, does it actually matter why the caps crack? Isn't it sufficient simply to know that they do crack? --Tom -- Prof. Thomas H. Lee Allen Ctr., Rm. 205 350 Jane Stanford Way Stanford University Stanford, CA 94305-4070 On 9/11/2022 16:05, Dave McGuire wrote:
On 9/11/22 18:41, Tom Lee wrote:Linear vs. SMPS would not be expected to matter, in any case. The problem is one of moisture ingress. The encapsulant shrinks and cracks over time, letting moisture in. At some point the capacitor breaks down and you get a multimedia spectacular. None of those processes is particularly sensitive to what circuitry lies downstream.? It's my understanding that it doesn't happen quite like this. My understanding is that the epoxy is hygroscopic, and its mass increases as it absorbs moisture.? The mass of epoxy swells, causing first crazing, then full cracking of the outer shell. Leakage increases and the resultant current flow causes heating, until it goes bang.? I believe the end stage there is thermal runaway. |
Re: Suffering from RIFA anxiety!
Will the RIFA waxed paper caps sold today have the same failure mode in 20 or 30 years? They appear to have similar/same construction.
I recently installed a new RIFA PME271 Y cap in my 8753D (against my better judgment, but it was the only cap with the right lead spacing) so I guess I'll let you know in 20 years. |
E4982A Broken hard disk.
QuyetN
hi everybody!
Hi guys
Currently, I have an E4982A device that has a hard drive failure and cannot work. and around me no one has this device so I can borrow the hard drive to check the rest of the components.
so I ask everyone, who has E4982A, can you help me create a backup file of the hard drive (using a Terabyte Image to create a ghost of the entire hard drive) and send me the download link?
I would like to thank everyone. |
Re: 8660C restoration attempt
Is the fan running at a high speed? They are a bit noisy. Bad bearings or dust/gunk may be slowing it down and not allowing enough air to pass. I have 3 8660 B versions. I do agree the extension module is tricky and I don't have the extender kit. I know one of my extenders has an issue and I believe its a varicap or the voltage feeding it. But lifting the monster on the bench is a concern as I already have an operational one on the bench along with other heavy hp gear. Good luck Paul WB8TSL |
Re: fiche Re: [HP-Agilent-Keysight-equipment] HP microfiche collection / scanning
¿ªÔÆÌåÓýHello--Regarding fiche as an information-storage medium, interested parties will profit reading from this book: Double Fold: Libraries and the Assault on Paper??? by?"...The ostensible purpose of a library
is to preserve the printed word. But for fifty years our country¡¯s
libraries¨Cincluding the Library of Congress¨Chave been doing just the
opposite, destroying hundreds of thousands of historic newspapers and
replacing them with microfilm copies that are difficult to read, lack
all the color and quality of the original paper and illustrations, and
deteriorate with age.
With meticulous detective work and Baker¡¯s well-known explanatory power,?Double Fold?reveals a secret history of microfilm lobbyists, former CIA agents, and warehouses where priceless archives are destroyed with a machine called a guillotine. Baker argues passionately for preservation, even cashing in his own retirement account to save one important archive¨Call twenty tons of it. Written the brilliant narrative style that Nicholson Baker fans have come to expect,?Double Fold?is a persuasive and often devastating book that may turn out to be?The Jungle?of the American library system....." Your local public library may have a copy (or borrow one via intrlibrary loan). 73-- Brad? AA1IP |
Re: HP microfiche collection / scanning
Peter -
toggle quoted message
Show quoted text
When (if) you get a chance - could you see if there is anything on the 8510B Cheers! Bruce Quoting Peter Brown <peter@...>: Hi Bruce,It would be interesting to see the quality of the scans.Sample scans in the files section |
Re: HP microfiche collection / scanning
Now that I re-read your first?email, I see the 1969-1979 date range. The HP-113AR was introduced in 1960 and replaced by the 113BR and 114BR within a year or two. So they would not be in a fiche set with those dates.? On Wed, Sep 14, 2022 at 2:43 PM Peter Brown <peter@...> wrote: Hi Jeremy, --
Jeremy Nichols 6. |
Re: HP microfiche collection / scanning
Hi Bruce,
>It would be interesting to see the quality of the scans. Sample scans in the files section? ?A temporary directory for photographs and help relating to emails and posting > ... the 8510B (both the 85102B, and the 85101B) If anything. Nothing on these - it looks like the instrument was introduced in the 1985 catalogue and so is outside the date range of the fiches that I have - 1969 - 1979 (-ish) Cheers Peter |
Re: HP microfiche collection / scanning
Peter -
toggle quoted message
Show quoted text
It would be interesting to see the quality of the scans. Also, when (if ?) you have any time, I'd like to know what you have on the 8510B (both the 85102B, and the 85101B) If anything. Cheers! Bruce Quoting Peter Brown <peter@...>: Hi All, |
Re: 8660C restoration attempt
Ruben -
toggle quoted message
Show quoted text
Any frequency "wobble" is likely due to the frequency extension module (11661B). Ihis is easy to check if you have the 8660 RF test plug in. Just put a spectrum analyzer on the various outputs and see which are unstable. Without the 86601 (0-500MHz) plug, it is a bit difficult to separate mainframe problems from 11661B problems (but the test module will help with that) I do not remember the 8660s running all that hot, but it has been probably 30 years or so. Anyway, the fun starts not and you will not have to worry about what you are going to do with your "sree" time :-) Cheers! Bruce Quoting RubenRubio <rubenrb2019@...>: Hi there again! |
Re: HP microfiche collection / scanning
HI Jeremy
>Would love to see a sample I have added a subdirectory in the files section under? ?A temporary directory for photographs and help relating to emails and posting It contains a poor quality sample from a scratched original and a good sample from a clean original All of the earlier fiches are photographs of actual manuals - some of the later ones appear computer generated and may be of better quality Peter |
Re: VNA Calibration - Port Impedance - a late life ponder
Dan,
toggle quoted message
Show quoted text
Here is my take on this: When you do a full two port calibration and then apply that calibration, the measurements you get are of the device under test (DUT).? The calculations remove errors due to the raw characteristics of the VNA ports. When you do the direct connection to port B, the DUT is not port B, but rather the non-existent (infinitely short) transmission line at the reference plane.? That DUT has an S11 and S22 of zero, and an S12 and S21 of one. --John Gord On Wed, Sep 14, 2022 at 11:23 AM, dan.meeks222@... wrote: Hello Group - |
Re: HP microfiche collection / scanning
Peter, thanks for looking for the 113AR! For Chris¡¯s 5263, I had a look, can¡¯t find it in my HP catalogs, but the faceplate looks identical to the 5262. It¡¯s possible the 5262 manual might work for him.? On Wed, Sep 14, 2022 at 2:43 PM Peter Brown <peter@...> wrote: Hi Jeremy, --
Jeremy Nichols 6. |
Re: HP microfiche collection / scanning
Hi Chris,
>I would be looking for an 5263A Time-interval unit, Operation and Service. It is one of the plugin units for a 5245L Electronic Counter. I had a quick look around that instrument type number but no 5263A unfortunately 5262A - 05262-9007 Apr 15 1970 - op svc - pfx 516 5262A - 05262-9008 Aug 1970 - op svc - pfx 928 5262A - 05262-9008 May 1973 - op svc - pfx 928 5262A - 05262-9025 May 1974 - op svc - pfx 1124A 5264A - 05262-9006 Apr 1970 - op svc - pfx 616 5264A - 05262-9007 Apr 1970 - op svc - pfx 1124A 5264A - 05262-9004 Rev 3 Apr 1973 - op svc - pfx 832, 1124A 5264A - 05262-9010 Jan 1974 - op svc - pfx 1124A Peter |
8660C restoration attempt
Hi there again!
Today a new problem child arrives at home, ¡®79 8660C sig gen with 86634A phase modulator, 11661B freq xtender and 86603A RF section. Not a big fan of this HP era, but a few bucks for 33 kg of seventies stuff¡ nobody becomes bitter by a candy.? Here you have my first impressions: Cosmetic issues, some rust in the aluminium frame, dust inside, two keys broken (the stud still in, so easy repairable), attenuator knob in RF section needs glue too, so, not a big deal on that side. When I opened the pannels, saw a fantastic Rifa capacitor exploded, inmediately extirpate. Then I¡¯ve started the unit using variac, it slowly turns on without smoke or explosions. Keys, switches, RPG, anunciators, led display and meters roughly working at first sight in both mainframe and plugins. Let¡¯s try to see something at scope. Power seems ok, a bit over the numbers I read on meter, but not bad.?Frequency is 5.6 MHz over the selection on the display?and keep this behavior. You can see this in my SA screen. Obv it doubles the error at 2 GHz as you can see in my counter. Also the spectra is wobblin¡¯ a lot with 100 kHz of RBW, with less resolution the skirt isn¡¯t a skirt, only high noise. Terrible. Obviously have a loop problem. But first things first. Power supply rails are quite spot on in dc measurement,?but ripple is terrible.?Figures are three and four times?over the specs. Also I¡¯ve cheked unregulated rails for ripple and two of them have 250 and 300 mV over the specs. In the scope (AC coupled) the waveforms look weird, like a ramp or triangle. I think some caps are dry and maybe some diodes are blown in bridges. Besides this, I saw some corrosion and blackened area in the A20 board, near C4 and C5 capacitors. Tomorrow I¡¯ll take appart the caps and boards to evaluate them. During measurements I noticed that the corner of the instrument that holds the power supply guts are quite hot. Actually the whole machine was, but power related parts the most. Main transformer itself reaches 50?C (122?F) and the rectifier bridges almost 55?C (131?F). I think that isn¡¯t normal. I¡¯ll try an experiment: leave the unit cool down and connect it again but without turn the main switch on, thus if a capacitor is drawing current, I¡¯ll be noticed with some temperature increment. Anyway, I¡¯ll disassemble all parts to check them properly. Some work to be done. Again, opened ears to every chap that wants to contribute with an experience or tip. I¡¯ll be posting updates of my adventure. Mainframe op & service manual was easy to find, but nothing about the other three parts. Regards, Ruben ![]()
A919B818-C2B6-49B3-98C2-3E86B159A3D8.jpeg
![]()
A384A32B-2FA1-4177-AF1A-C8AC4C30E754.jpeg
![]()
09FCD108-CEB3-482E-B86C-4A01D2010390.jpeg
![]()
72097670-837B-47A5-A7C2-4DA3A6058B23.jpeg
![]()
D11510C9-3AB7-4849-8E5F-E6A9D18B1000.jpeg
![]()
E0E4DD1D-4DBF-46D8-9A31-55F9C3FA01B6.jpeg
![]()
32916862-BDC0-474A-B788-F0ABF014B08D.jpeg
![]()
62DBE511-930F-445E-BF88-B97DB1765186.jpeg
![]()
5AAC079A-1211-4EB4-8E30-C53A6366A05B.jpeg
![]()
864F53BC-8D2B-4FBE-A8E0-5A61D5267735.jpeg
![]()
D7B10521-646B-4657-86BE-B3AA8E6B3E69.jpeg
|
Re: HP microfiche collection / scanning
Would love to see a sample. The price sounds reasonable, especially with Sterling at a new low compared to the US dollar.? On Wed, Sep 14, 2022 at 2:14 PM Peter Brown <peter@...> wrote: Hi All, --
Jeremy Nichols 6. |
Re: HP microfiche collection / scanning
Hi All,
I spoke to the fiche scanning house today - after review of the scanning process for a set of sample fiches that I sent them (HP 5480A) they were able to more accurately price any further job. The fiches that I sent were typical of the earlier samples in the collection (some clean others quite scratched) and so required some post processing to separate the pages which put the per fiche price up a bit.? Their quoted cost is 1.07 UK pounds for each 14 x 7 fiche.? This works out at around 1.1 UK pence per page with a small discount likely for quantity Would it be of use for me to upload some of the resulting .pdf files for people to get an idea of quality?? 100 pages runs to around 3MB Peter |
VNA Calibration - Port Impedance - a late life ponder
Hello Group -
My whole life I have been calibrating VNAs and then verifying the cal by connecting Through, and looking for very good RL and IL across the band. I'll set up my confusion / question this way: let's say I have two VNAs that are perfectly calibrated. Now I take port 1 of VNA A and connect it to port 1 of VNA B. What I am doing is measuring VNA B's actual impedance on port 1. Which is not going to be perfect - or necessarily close to it. The "true" impedance of any VNA port is not as good as you measure after a cal with a through. Why is that? After a perfect cal, with a perfect through, S11 and S22 should be showing the actual "true" impedance of the other port (port 1 is looking at Port 2, and vice-versa. Can someone tell me what is wrong with my thinking on this? I must be wrong... but I don't know why. Thanks in advance - Dan |
to navigate to use esc to dismiss