Re: HP 8662A Power up - overvoltage shutdown
I recommend another weekend and replace all tear drop tantalums, it will happen again and again over time if you do not remove all. Regards Gerald VK3GM
By
Gerald
·
#126759
·
|
Re: HP 8662A Power up - overvoltage shutdown
WB8TSL: .... The 8663 supply is pretty much the same. The 20 V reg is pretty simple and can be traced out if needed. But check that the 20 V is not shorted to ground. The smoke makes me think
By
Gerhard_DK4XP
·
#126758
·
|
Re: E4418B power supply schematic?
Hi Tom thank you for the link to the power supply I have uploaded it to the E4418-9 file along with service and user information Regards Paul B
By
Paul Bicknell
·
#126757
·
|
Re: HP 8116A woes with E42 error
I can assure you: A steady-state voltage of approx. 10.5V (mind the polarity: Q502's emitter 10.5V more positive than its base, as may be concluded from 3) and 4)) between Q502's emitter and its base
By
Raymond Domp Frank
·
#126756
·
|
Re: HP 8116A woes with E42 error
---- If 1V-1.5V measurements are with no power applied then Q502 is bad as Raymond also suggested. |Vbe| shouldn't exceed ~ 0.7V even in-circuit if power is not applied.
By
Ozan
·
#126755
·
|
Re: HP 8116A woes with E42 error
I double checked my DC measurements for the cathode of CR502 and the emitter of Q502 and they were the same as above. Put a scope on pin 6 of U502, and there is minimal AC. 52 mV max. These are
By
Swiss
·
#126754
·
|
Re: HP 3465A offset when input shorted
Quick update on this one, mostly replying for completion. I located the suggested 2N4117A from Langrex here in the UK for a measly ?3.50 which was acceptable coin for this DMM. Arrived today. Did a
By
Chris Smith <me@...>
·
#126753
·
|
Re: HP 8116A woes with E42 error
... as Ozan correctly says. Raymond
By
Raymond Domp Frank
·
#126752
·
|
Re: HP 8116A woes with E42 error
I assumed you were looking with a minimal amplitude AC signal. If incorrect, you're unnecessarily complicating things and making measurements with a DC meter essentially useless. Raymond
By
Raymond Domp Frank
·
#126751
·
|
Re: 8657B distortion
Thanks Jim.? I don''t know where to look for the input to the attenuator so I better start reading the manual.? Your diagnosis is what I hope is the problem, since repairing that may not be as
By
Bob Albert <bob91343@...>
·
#126750
·
|
Re: HP 86200B detectors.
The 86200 is a entirely different architecture as the regular HP sensors , it is a diode based sensor that contains correction constants inside a EEPROM , if I remember they were based on the 85037
By
Lothar baier <Lothar@...>
·
#126749
·
|
HP 86200B detectors.
Hi. Apologies if I've asked this here before, but the grey cell doesn't recall doing so... I have two 86200B 50 Ohm 10 MHz to 3 GHz detector probe assemblies (with cables) that came with my slightly
By
Dave_G0WBX
·
#126748
·
|
Re: HP 8116A woes with E42 error
--- DC voltages do not add up to correct values, are these voltages in a state where output also has a large AC signal on it? For example cathode of CR502 is at -9.446V, which is the same node as base
By
Ozan
·
#126747
·
|
Re: HP 8116A woes with E42 error
The combination of 3) and 4) would indicate Q502 has open base-emitter. Raymond
By
Raymond Domp Frank
·
#126746
·
|
Re: servicing 8671B Synthesized CW Generator 2.0-18 Ghz unit (first power up)
First red flag is the missing seal on the top screw which indicates someone been in there before . The phaselock message normally clears once the unit warms up but it could also be a indication of a
By
Lothar baier <Lothar@...>
·
#126745
·
|
Re: E4418B power supply schematic?
Hi Tom, Thanks for the datasheet. I saved it for next time. This time I got lucky - another one of those resistors was open. It was the same value as the one I replaced several years ago, so I still
By
Razvan
·
#126744
·
|
Re: 8657B distortion
I sounds like there may be a problem with the reverse power protection.? There are PIN diodes in series with zener diodes to ground in the protection circuit and it sounds like perhaps one of the
By
Jim Summers
·
#126743
·
|
Re: servicing 8671B Synthesized CW Generator 2.0-18 Ghz unit (first power up)
Hello Lothar and HAKE community I just got cables shipped to me, and planned to use a semi rigid microwave test cable for early check out between 8671b generator and 8566b spectrum analyzer. I had a
By
Samudra N3RDX S21X
·
#126742
·
|
Re: E4418B power supply schematic?
Hi Razvan, I don't have a schematic, but mine has an Artesyn NFN40-7608. The datasheet is here. https://datasheetspdf.com/pdf-file/454677/ETC/NFN40-7608/1 Is the standby light on, or do you get
By
Tom B
·
#126741
·
|
Re: 8562A service manual
According to the manual on the Keysight site, 08562-90124 should be a complete set of the 'latest' CLIPs. The CLIP for A4 2923A prefix should be 08562-90128. You'd have to ask Dave at Artek whether
By
Orin Eman
·
#126740
·
|