Re: E4407B strange peak shapes at low RBW
Hi Sandra, What is a NSCOM reset? Do you mean factory reset / preset on the machine itself or some command?
By
Kalle Kempe
·
#126792
·
|
Re: HP 8510B Service manual CLIP
There is much similarity between the A and B versions. The schematics for the A are available. Can you work from that? Cheers! Bruce Quoting "tom_iphi via groups.io" <iphi@...>:
By
Bruce
·
#126791
·
|
Re: E4407B strange peak shapes at low RBW
The problem isn't exactly new , back when the ESG generators came out we had to buy a bunch of equipment including a 89440A and E4419B because the service software only supported those and actually
By
Lothar baier <Lothar@...>
·
#126790
·
|
Re: HP8640B no rf out
Another plastic-disk finger contact symptom and cause. This is the disk on the back of the step attenuator. SYMPTOM: Intermittent: LEVEL mode meter reads zero, RF still output. CAUSE: A1A1S1 Meter
By
Dave Wise
·
#126789
·
|
Re: E4407B strange peak shapes at low RBW
Keysight will no longer sell the licenses for the N7800A unless you are a large company and want to do your own calibrations but the catch is that you HAVE to use the equipment that's written into the
By
Richard Parrish
·
#126788
·
|
Re: 4395A and other project units FS
Hi Lothar, Is the 4395A?still available? If Yes. Can you send it to Switzerland? Best regards Rico
By
Rico Sonderegger
·
#126787
·
|
Re: Keithley 236 firmware version A10
Hi, I finally found a link on EEVblog and was able to find and install the new firmware successfully. Thanks! De : [email protected] <[email protected]> De
By
Yves Tardif
·
#126786
·
|
hp 8407A Service Cable
All, Would anyone have a service cable for the 8407A Network Analyzer for the 8412A, 8413A ,and 8414A/B plugins? By looking at the number of pins in the Amphenol connector, it appears to be the same
By
Christopher Tucker
·
#126785
·
|
Re: Keithley 236 firmware version A10
Hi Yves, I own a 237 (high voltage version). These devices share the same firmware. Only a resistor on the mainboard decides whether the 236 or the 237 version. I have the A10 version and can send it
By
Rico Sonderegger
·
#126784
·
|
Re: HP 8662A Power up - overvoltage shutdown
Good quality new tantalums with appropriate voltage ratings. Digikey has a good range. I did send a PM. Regards Gerald VK3GM
By
Gerald
·
#126783
·
|
Re: HP 8510B Service manual CLIP
Hi folks, I have a broken 8510B display section that I'm trying to fix and I have been chasing a schematic for a long time. Has anyone spotted the mentioned B-supplement? I know those display sections
By
tom_iphi
·
#126782
·
|
hp 3721A Correlator
I am looking to acquire an hp 3721A Correlator, in working condition, if possible. I am also looking for any manuals or application notes. Thanks in advance. Christopher
By
Christopher Tucker
·
#126781
·
|
Re: E4407B strange peak shapes at low RBW
HP used discrete filters for the IF filters , those filters are constructed using multiple poles that are coupled and each pole is adjusted individually to form a specific filter shape and bandwidth
By
Lothar baier <Lothar@...>
·
#126780
·
|
Re: E4407B strange peak shapes at low RBW
FWIW I had a similar issue after a processor board repair. I ended up doing to NSCOM clear of the unit then a Align All and it went away for me. Sandra
By
Sandra Carroll <smgvbest@...>
·
#126779
·
|
Re: E4407B strange peak shapes at low RBW
Hi Are you using an external test signal?-it seems so. First i would make a self alignment with the inbuildt 50 Mhz cal signal. Hardy Fra: [email protected]
By
hardyhansendk
·
#126778
·
|
Keithley 236 firmware version A10
Hi, All the instruments in my lab are mainly HP/Agilent/Keysight, and two from tektronix, but now I have two Keithley sources, a 236 and a 238 functional received very recently. I was able to find the
By
Yves Tardif
·
#126777
·
|
Re: HP8920A display onto computer monitor
The SimmConn Labs LCD replacement display does a nice job, although may be more than some want to spend. As I recall, it can bring out a VGA port as an option. Also, I have a CRT display removed from
By
Tom via Metronet
·
#126776
·
|
Re: HP8920A display onto computer monitor
Looks like at least one enclosure available. I would prefer metal as I have a enough RF floating around. **** GBS-8200 CGA/EGA/YUV to VGA Video Converter Board Case ****
By
RFI-EMI-GUY
·
#126775
·
|
Re: HP8920A display onto computer monitor
There is more info on the 8920 group /g/HP8924-Family-Spectrum-Analyzers/topics I've used both Gonbes 8200 (8220 is same but 2 VGA outputs) + LM1881, the Sont DSC104 and a
By
Robert G8RPI
·
#126774
·
|
Re: HP8920A display onto computer monitor
There is no why spend so much money on that https://tabajara-labs.blogspot.com/2021/01/como-ligar-um-monitor-vga-no-hp-e8285a.html?m=1 Enviado do meu Tele-Movel Em ter, 31 de mai de 2022 16:14,
By
Alexandre Souza
·
#126773
·
|