Re: Scope Probe parts
You might check the catalog at probemaster.com
These are not Agilent/HP probes, but their own OEM designs, but there's some chance that they will have accessories which will fit your probes.? I found, a few years ago, that their screw-in probe tips are a good fit for the probes on the venerable HP 8405A vector voltmeter (and are much less expensive than new-old-stock HP). |
Re: 8566B 1dB scale trace not linear at bottom of screen
This is most probably an 8566 adjustment problem. Look over the adjustment procedure and you will find a section where the attenuators are adjusted. I'll try to find time to look up the exact section, but I remember doing this and correcting similar errors.
toggle quoted message
Show quoted text
Quoting DW <wilson2115@...>: Here is a thought, if someone with a 8566 could reproduce my results by injecting a -20dBm 10 Mhz signal and setting the reference level to -20dB and set the scale to 1dB. Then with the signal generator start dropping the signal from -20dBm to -30dBm in 1dB increments. Does the trace start to fall off towards the bottom of the screen compared to 2dB and 5dB? |
Re: 8566B 1dB scale trace not linear at bottom of screen
Here is a thought, if someone with a 8566 could reproduce my results by injecting a -20dBm 10 Mhz signal and setting the reference level to -20dB and set the scale to 1dB. Then with the signal generator start dropping the signal from -20dBm to -30dBm in 1dB increments. Does the trace start to fall off towards the bottom of the screen compared to 2dB and 5dB?
|
Re: Capturing screenshots on 8934C
¿ªÔÆÌåÓýHi, ? I found that for screen capture about 8924C? ? BTS software ? ? De?: [email protected] <[email protected]> De la part de w9dkc ? Sorry, yes. 8924C |
Hp8558b Spectral analyzer (1/2)
hello,
i have repard the displays This analyzer never allowed me to display anything except once , with a generator and a 5v signal which is a bit strong May you tell me if the?if the BNC connector has been replaced by an N. the plug information is not clearly visible? ---> see picture May you tell me if the frequency scanning operation seems normal to you (I have repaired the displays)? --->see the vid¨¦o Thank's for your returns cdt |
File Notifications
#file-notice
[email protected] Notification
The following files have been uploaded to the Files area of the [email protected] group. By: Ken Eckert <eckertkp@...> The following files have been uploaded to the Files area of the [email protected] group. By: Ken Eckert <eckertkp@...> |
Re: Asking help to experienced people who repaired a 3585A spectrum analyzer
Hi Alberto,
no worries, one purpose of this forum is to contribute and share experience when someone is in trouble with a HPAK instrument, so just stay on. I assume that you have checked that, on A34, the existence of the 100kHz clock signals around U9, 10, 14 and the signal at TP3, correct? Something happening at U15C, U3B? But not signs of life at TP2? BTW, it is very helpful to go through the functional descriptions in Vol.I Section VI to understand what is happening inside the 3585A. If completed, check the following: - fault isolation procedure in Vol.II from pg. 8-1 (hint about which loop is not locked on pg. 8-7) - the LEDs indicate which loop might be affected, see Vol.II table 8-4 (pg. 8-10) That should give you an indication about if the problem really is located on A34. Chris |
FS: Original Tek 576 Op/Serv Manual and Type 172 Programmable Test Fixture $US200.00 for both
The 576 manual is the August 1984 revision and in excellent condition.
The "Type 172 Programmable Test Fixture" for the 576 is in likewise excellent shape. I want to sell them as a pair, they'll be shipped from Calgary, AB, Canada and UPS Ground into the ConUS I guess might run $US50-75.00, depending. The 172 comes with a programming card and I swear up and down I have a slew of original pins and an extra couple of programming cards sitting in one those places where I'll be -sure- to be able to locate them. They're included once I do. A cleaned up, OCR-d copy of the manual from the TekWiki is here: PayPal pretty much works but for unknown "reasons" that I ascribe to brain dead AI it has begun randomly withholding payments into and transfers out of the account, so I'll ask that any payment be sent as, "Money to those you trust" or whatever they call it these days. Hit me off list please at sales ¨C at ¨C pearl-hifi ¨C dot ¨C com Liam |
Re: Asking help to experienced people who repaired a 3585A spectrum analyzer
Hi Chris,
I just checked before the A21 unit, without pulling out it, according to the Service Group B of Section 11: when I started directly from the "Service" chapter, the "LOCAL OSC. UNLOCKED" error led me to that Group. By following the "Main Local Oscillator Troubleshooting Tree" (11-151/11-152), I found perfectly within the tolerances the following signals:
Now, by restarting from Section 5 - as you advised - the first step I should begin from is ¡ì 5-7 "90 MHz Reference Board Adjustment (A21)": in fact I just checked the voltages of the Low Voltage Power Supply and I should prefer to avoid checking its current protection. NB In Section 11 Group B the frequency counter and the RF synthesizer had to be externally sync'd to the 10 MHz reference output of the 3585A, but in Section 5 this is true only for the synthesizer (see ¡ì 5-5). Therefore I will sync the frequency counter to my GPSDO 10 MHz reference. Last (by now): you wrote that in you troubleshooting you bought a second unit as donor, to help you focus the possible fault in case of multiple failures. I have not this possibility, unless buying certain replacement boards: it is a little expensive approach and meanwhile I got almost all the ICs of the A34 board that I considered possibly involved, by following the schematic diagrams up to the TP2, where I cannot find the expected "pulses"... I didn't replace all of them, but now I'm dubious if a spare A34 board could be a good choice. PS if this posting begins boring someone, please tell me: I will go on in private form. Tnx de Alberto, IZ2EWV |
Re: A homemade diode power sensor for HP meters
This has been an interesting thread. Like everyone lots of dead power sensors. Whats funny is now even the dead sensors seem to have magical value. The actual meters are nice. Something about a analog readout for peaking power while tuning. But that said I do agree that the modern ic power sensors good to about 6 Ghz seem like a reasonable fit. But I believe that most of the circuits used in the meter may not be needed. If I get around to this I would tend to set the gain of a opamp in the meter through the db switch. The sensors have a predictable voltage out. A small?part of that range needs to be mapped into each power switch range. Easier said than done. The alternative and I think perhaps far harder is to compress the IC sensor into what a diode would have put out. Then use the chopper and such as is. Best of luck to anyone who goes further. Regards Paul |
HP4140 offset?
I have a question for any HP4140B users out there... Mine seems to power up with an offset of around -230fA when on the 1pA range. After about an hour's warmup this drops to around -120fA but no lower. And the 'Zero' button has no effect (presumably as it says it can only compensate for +/- 100fA?).
I have tried adjusting A1's offset as per the 'adjustments' section of the manual, but it doesn't change things. It seems to be a 'real' current - leakage possibly (the triaxial socket is clean, not looked at the A1 input board yet). Has anyone had a similar problem and if so did you manage to cure it? |
Re: Copying 4145A floppy - How to do it / Conclusion
After researching the matter of copying 4145A diskettes I've concluded that my
earlier statements regarding the diskette format were totally wrong. I have now made copies, copies of copies, etc. of the original HP4145A diskette that came with my 4145A - and it was simple and straight-forward. I was simply WRONG about the diskette being hard sector. I was told that by the previous owner - who is a technically astute person - and I just believed him. I thought I had checked it myself, but that must have been an "illusion". Today I used IMD to read my original HP diskette on my "copy stuff PC". I used a floppy drive capable of single to quad density to create the "copy diskettes". The parameters I used were 40 tracks, single side and double step. My original diskette read in flawlessly and created an IMD image file. I then used that file to create another diskette. I then read that diskette to create a second IMD file - and used that file to produce another diskette (copy of a copy). I also used a 4145A IMD image I retrieved from the Internet and used it to create a third diskette. I tested out all of the diskettes on my HP 4145A - and ALL worked flawlessly. In short, the diskette is NOT weirdly formatted! It is single density (FM - 250K), 9 sectors/track, 256 byte sectors for a total of 360 sectors on the diskette. Sorry that I didn't research this properly before I made invalid comments on this list and wasted Dave McGuire's time (and energy) and others who were following our discussion... Best, Lyle BTW: I studied the 4145A schematics and the chip associated with the floppy controller - which is a Fujitsu MB8866. It only is capable of handling a single index and its single density FM format and frequency is the same as that of an IBM 3740. It is also capable of MFM double density (not used in the HP 4145A). ========= On Fri, 4 Jun 2021 18:39:27 -0400 "Dave McGuire" <mcguire@...> wrote: On 6/4/21 5:24 PM, Lyle Bickley wrote:Wow, ok.Uhh, waitaminute. Is it possibly the case that there are twoWhen I got my 4145A (and a floppy diskette that would boot) I talked with -- 73 NM6Y Bickley Consulting West "Black holes are where God is dividing by zero" |
hp 35670A option 1D2 swept sine
I think that this has been asked before, however I am not aware of a definitive answer.
My question is were or how can I obtain the 1D2 option for the 35670A? This appears to be a software upgrade (via floppy disk) but I have seen references to ROM upgrades. I am confused. My thanks for any assistance with this. Mark |
Re: Copying 4145A floppy - a suggestion
Maybe should be making a new topic for "Reading/Writing Unusual Disks"
For those interested in Kryoflux/Greaseweasel, also have a look at I killed my Kryoflux (dont ask...) and found FluxEngine when looking for a replacement. I have the parts on my bench waiting so I cannot vouch for it, but it's a very economical alternative to KryoFlux. |
Added lots of hard to find FETs and MOSFETs to the Stuff page.
walter shawlee
So many people have been chasing hard to find FETs recently that I spent some time digging though all our semi inventory, and posted up the ones I thought might be useful for fixing test gear and RF items. You can see them here on the stuff day page:
?
https://www.sphere.bc.ca/test/stuffday.html#fets
?
I have put the Tek parts there too (and I have more to post this week, also HP parts). I will try and make sure there is a rich assortment there to look through. I also added three interesting FET substitution guides at the top of the section, to help you find a replacement for that toasted part. There are 2N4000/500 series parts, J-series parts, and M-series parts, plus Tek 151-1xxx parts so far. I also have siliconix ultra low leakage current diodes and CL diodes I will add in shortly.
?
all the best,
walter (walter2 -at- sphere.bc.ca)
sphere research corp.
https://www.sphere.bc.ca/test/index.html |
to navigate to use esc to dismiss