开云体育

Date

Re: spectrum HP 8558b good but missing part

 

This is a really common problem. Generally, the contacts fall off somewhere inside
the unit, but they may go missing. You can fabricate some from suitable shim stock.
Here's one person who did that:



It's a bit of a ways down. I'm not sure what the original material was (stainless? a
nickel alloy?) but given the age of the equipment and the use it will likely get, I don't
think it would matter a lot. I think some have made up contacts from a piece of stainless
steel feeler gauges (like you would use in auto repair).

good luck

On Tue, Sep 03, 2019 at 02:56:08PM -0700, johnlegros wrote:
hello , have a nice working spectrum 8558b with electron beam memory? , all is correct have checked , but the issue the time division switch does not work , in auto mode, , opened it and found a missing part in the switch , from the service manual sent with this spectrum analyser from new only 40 years ago part number is 5061-5425 pronger switch , it is the metal leaf part which is missing , checked and did not find it anywhere , so question is does anyone have or make these ? thanks





!DSPAM:5d6ee55b212076544458839!
--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@... | Unix & Windows


Re: Spare feet (5060-0727)?

 

The bumpers sound like a good idea. Stick on ones are available at most hardware stores in addition to general merchandise stores like Walmart and Target. Even some drugstores.
Make both the same and no one will ever notice:-)
Good luck with your project.

On 9/3/2019 4:38 AM, Robert G8RPI via Groups.Io wrote:
As a quick alternative you could slide the feet off the one with feet and put some 3M "bump-on" or similar self adhesive feet / bumpers on both so they match. Target sell the 3M bumpers for less than $5 a pack

(I'm English and I knew that Target sold them :-)
Robert G8RPI
--
Richard Knoppow
dickburk@...
WB6KBL


Re: E4418B power meter clearing Ram battery failed

 

Thank you all. That is what I meant.


Re: E4418B power meter clearing Ram battery failed

 

开云体育

High an alternative to FFF could be FFP????? fit for purpose

?

The @ is instead of saying Ha You

?

TTFN??


From: [email protected] [mailto:[email protected]] On Behalf Of Jim Ford
Sent: 03 September 2019 17:23
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] E4418B power meter clearing Ram battery failed

?

Glenn, @ means the comment is directed at the particular person.? As in @Glenn Little if I were addressing you.

?

FFF is "Form, Fit, and Function"

?

Jim Ford?

?

Sent from my Verizon, Samsung Galaxy smartphone

?

-------- Original message --------

From: Glenn Little <glennmaillist@...>

Date: 9/3/19 8:56 AM (GMT-08:00)

Subject: Re: [HP-Agilent-Keysight-equipment] E4418B power meter clearing Ram battery failed

?

What do the "@" signs mean?
And what is FFF?

I have never seen these used before.

Glenn

On 9/3/2019 11:16 AM, victor smith wrote:

@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:??

?
-----------------------------------------------------------------------
Glenn Little?????? ?????????ARRL Technical Specialist?? QCWA? LM 28417
Amateur Callsign:? WB4UIV??????????? wb4uiv@...??? AMSAT LM 2178
QTH:? Goose Creek, SC USA (EM92xx)? USSVI LM?? NRA LM?? SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 

No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15884 - Release Date: 08/14/18
Internal Virus Database is out of date.


Re: 3336B optional 10Mhz Osc. problems

walter shawlee
 

The OCXO Oven works fine, heats up, and stabilizes, but is off frequency at roughly 10,000,250Hz.
the can is soldered shut, doesn't really seem to lend itself to a repair.

regards,
walter

--
Walter Shawlee 2
Sphere Research Corp. 3394 Sunnyside Rd.
West Kelowna, BC, V1Z 2V4 CANADA
Phone: +1 (250-769-1834 -:-
+We're all in one boat, no matter how it looks to you. (WS2)
+All you need is love. (John Lennon)
+But, that doesn't mean other things don't come in handy. (WS2)
+Nature is trying very hard to make us succeed, but nature does not depend on us.
We are not the only experiment. (R. Buckminster Fuller)


Re: E4418B power meter clearing Ram battery failed

 

开云体育

Hi victor

?

Thank you for the link unfortunately there are 3 people selling this item and non of them will ship to the UK

?

Thank you again? I have a stock of old batteries that I thought / suspected as being no good but from recent information it could be a known fault with the 4418

So It will come back to the repair bench for investigation

?

Regards Paul B

?


From: [email protected] [mailto:[email protected]] On Behalf Of victor smith
Sent: 03 September 2019 16:17
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] E4418B power meter clearing Ram battery failed

?

@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:?

No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15884 - Release Date: 08/14/18
Internal Virus Database is out of date.


spectrum HP 8558b good but missing part

 

hello , have a nice working spectrum 8558b with electron beam memory? , all is correct have checked , but the issue the time division switch does not work , in auto mode, , opened it and found a missing part in the switch , from the service manual sent with this spectrum analyser from new only 40 years ago part number is 5061-5425 pronger switch , it is the metal leaf part which is missing , checked and did not find it anywhere , so question is does anyone have or make these ? thanks


Re: 3336B optional 10Mhz Osc. problems

 

An OCXO that's that far off likely has oven failure, so it never warms up to the turnover point. The worst case is a burned out heater, but I've usually found other, simple, fixable causes. If you can easily open it up, you can check the basics even without schematics. I recall for instance, having a couple of OCXOs (I think these are Ovenaire brand too) from 8568/8566 SAs that were dead cold. I found somewhere in the guts, a cermet trimmer pot in the oven temperature control circuit, that had its wiper open - both units had the same problem. On the first, I replaced the pot with an identical type, with the rotor set to the same position, to at least land close to the original setting. This restored proper oven operation, and it was good to go, although no longer set at the exact right temperature.

I found that the original pot's wiper contact could be restored by turning the rotor back and forth a number of times. I think the harsh heat environment and out-gassed fumes from the insulation and components in there degraded the contact, but cleaning it up was enough to work again. I think on the second unit, I marked the rotor position, and just worked it over a bunch of times, then reset it to the same spot.

Since then, I have piled up several more of these dead units for repair, and plan to fix them en masse, assuming they all have this problem. BTW, depending on where the OCXO was used, the insulation may be degraded mechanically - the guts will thump around inside when shaken - so it should be restored too. Every one from an 8566 has been in this condition, while the same kind from 8568s are solid. The 8566 abuses the oven due to the poor choice of mounting location, thermally speaking. I have often lamented this in 8566 (especially cooling and fan) repair threads.

Ed


Re: 3336B optional 10Mhz Osc. problems

 

Hi,

Does the heater of the oven work? I had a similar case with a 3225b. The internal trimmer to set the oven disintegrated over the years. I carefully disassembled the oven and fixed.
I measured to current consumption to first diagnose that the oven did not work. It was also obvious that the case was vold to thr touch.
The frequency is way off when the oven is cold.


Szabolcs


walter shawlee <walter2@...> ezt írta (id?pont: 2019. szept. 3., Ke 21:54):

I have a 3336B that seems to be made of parts from both an A and B unit, but had the optional high performance ovenized oscillator installed (hp pcb 03325-66509) which is an Ovenaire OSC 73-52.

sadly for me, while I got everything else up and running correctly, the Ovenaire oscillator is WAY off frequency (+250hz), and can't be adjusted via the EFC or mechanical adjustment. if I use the external clock input, everything works correctly, with no offset.

Does anybody have either a good spare board or the Ovenaire oscillator available? Happy to buy or swap parts.

it has one other odd issue I have not seen before in any hp synthesized generator, it has a flutter or approximately 0.5-1Hz up and down (residual FM) at 10Mhz out. Is this some kind of loop instability that can be adjusted, or just how it works? my 5335A easily resolves 0.01Hz at this frequency, so this is very visible to me.

all help appreciated (best off list),
-walter (walter2 -at- )
sphere research corp.


Re: 3336B optional 10Mhz Osc. problems

Terry Maurice
 

开云体育

Walter, I am interested too, so please keep all replies on-list.

Terry

On 9/3/2019 3:54 PM, walter shawlee wrote:

I have a 3336B that seems to be made of parts from both an A and B unit, but had the optional high performance ovenized oscillator installed (hp pcb 03325-66509) which is an Ovenaire OSC 73-52.

sadly for me, while I got everything else up and running correctly, the Ovenaire oscillator is WAY off frequency (+250hz), and can't be adjusted via the EFC or mechanical adjustment. if I use the external clock input, everything works correctly, with no offset.

Does anybody have either a good spare board or the Ovenaire oscillator available? Happy to buy or swap parts.

it has one other odd issue I have not seen before in any hp synthesized generator, it has a flutter or approximately 0.5-1Hz up and down (residual FM) at 10Mhz out. Is this some kind of loop instability that can be adjusted, or just how it works? my 5335A easily resolves 0.01Hz at this frequency, so this is very visible to me.

all help appreciated (best off list),
-walter (walter2 -at- sphere.bc.ca)
sphere research corp.


3336B optional 10Mhz Osc. problems

walter shawlee
 

I have a 3336B that seems to be made of parts from both an A and B unit, but had the optional high performance ovenized oscillator installed (hp pcb 03325-66509) which is an Ovenaire OSC 73-52.

sadly for me, while I got everything else up and running correctly, the Ovenaire oscillator is WAY off frequency (+250hz), and can't be adjusted via the EFC or mechanical adjustment. if I use the external clock input, everything works correctly, with no offset.

Does anybody have either a good spare board or the Ovenaire oscillator available? Happy to buy or swap parts.

it has one other odd issue I have not seen before in any hp synthesized generator, it has a flutter or approximately 0.5-1Hz up and down (residual FM) at 10Mhz out. Is this some kind of loop instability that can be adjusted, or just how it works? my 5335A easily resolves 0.01Hz at this frequency, so this is very visible to me.

all help appreciated (best off list),
-walter (walter2 -at- sphere.bc.ca)
sphere research corp.


Re: E4418B power meter clearing Ram battery failed

 

Steve Hendrix mentioned:
(snip) "This may or may not be relevant to your situation, but I recall a bizarre problem" (snip)

Years ago a tech replaced a suspected bad?backup battery in a PC.? But what he didn't know was that the one removed was a rechargeable type.? He installed a non-rechargeable in its place.? It took about a day before the explosion happened.? Fortunately the traces on the board where the hole appeared were able to be jumpered across and things got working again - after another rechargeable battery was installed.

Greg


Re: E4418B power meter clearing Ram battery failed

 

开云体育

Thanks

I am now enlightened.

Glenn

On 9/3/2019 12:22 PM, Jim Ford wrote:
Glenn, @ means the comment is directed at the particular person.? As in @Glenn Little if I were addressing you.

FFF is "Form, Fit, and Function"

Jim Ford?



Sent from my Verizon, Samsung Galaxy smartphone

-------- Original message --------
From: Glenn Little <glennmaillist@...>
Date: 9/3/19 8:56 AM (GMT-08:00)
Subject: Re: [HP-Agilent-Keysight-equipment] E4418B power meter clearing Ram battery failed

What do the "@" signs mean?
And what is FFF?

I have never seen these used before.

Glenn


On 9/3/2019 11:16 AM, victor smith wrote:
@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:??

-- 
-----------------------------------------------------------------------
Glenn Little                ARRL Technical Specialist   QCWA  LM 28417
Amateur Callsign:  WB4UIV            wb4uiv@...    AMSAT LM 2178
QTH:  Goose Creek, SC USA (EM92xx)  USSVI LM   NRA LM   SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 

-- 
-----------------------------------------------------------------------
Glenn Little                ARRL Technical Specialist   QCWA  LM 28417
Amateur Callsign:  WB4UIV            wb4uiv@...    AMSAT LM 2178
QTH:  Goose Creek, SC USA (EM92xx)  USSVI LM   NRA LM   SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 


Re: E4418B power meter clearing Ram battery failed

 

On 9/3/19 11:22 AM, Jim Ford wrote:
FFF is "Form, Fit, and Function"

Oh good. I was worried it might be F-ing language. :-)


Re: E4418B power meter clearing Ram battery failed

 

开云体育

Glenn, @ means the comment is directed at the particular person.? As in @Glenn Little if I were addressing you.

FFF is "Form, Fit, and Function"

Jim Ford?



Sent from my Verizon, Samsung Galaxy smartphone

-------- Original message --------
From: Glenn Little <glennmaillist@...>
Date: 9/3/19 8:56 AM (GMT-08:00)
Subject: Re: [HP-Agilent-Keysight-equipment] E4418B power meter clearing Ram battery failed

What do the "@" signs mean?
And what is FFF?

I have never seen these used before.

Glenn


On 9/3/2019 11:16 AM, victor smith wrote:
@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:??

-- 
-----------------------------------------------------------------------
Glenn Little                ARRL Technical Specialist   QCWA  LM 28417
Amateur Callsign:  WB4UIV            wb4uiv@...    AMSAT LM 2178
QTH:  Goose Creek, SC USA (EM92xx)  USSVI LM   NRA LM   SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 


Re: HP8714B source module extender needed

 

开云体育

Thanks, Tam.

Jim



Sent from my Verizon, Samsung Galaxy smartphone

-------- Original message --------
From: Tam Hanna <tamhan@...>
Date: 9/2/19 2:06 PM (GMT-08:00)
Subject: Re: [HP-Agilent-Keysight-equipment] HP8714B source module extender needed

Component Level Information Package.


Document with the schematics of a Hewlett Packard TMI.

With best regards
Tam HANNA 

Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at  
On 2019. 09. 02. 23:02, Jim Ford wrote:
Uh, dumb question:? what is the clip?? Or CLIP?

Thanks.?

Jim Ford?



Sent from my Verizon, Samsung Galaxy smartphone

-------- Original message --------
From: "Dave_G0WBX via Groups.Io" <g8kbvdave@...>
Date: 9/2/19 1:47 AM (GMT-08:00)
Subject: Re: [HP-Agilent-Keysight-equipment] HP8714B source module extender needed

From: hp70000freak
Date: Sun, 01 Sep 2019 06:57:29 PDT

Hello,

I had the same issue long time ago. It was a faulty RF amplifier stage. As far as I can remember schematics are in the net and the amplifier was something like ATF 53142 or so.?

I soldered thin coax onto the board in order to pickup the signal stage by stage and find the fault.

?Cheers,

Hi.

Thanks for that, and yes I suspect it's just an internal amplification issue.

Yes, I also found some schematics in the past, but also note that ArtekManuals have the full service manual + clip.

??? ????? and just put 8714B in the search box.

Need to find some space to do stuff like this, as those modules with the covers removed (for access) are rather vulnerable to mechanical damage.

Regards to All.

Dave B
(G0WBX)

-- 
Created on and sent from a Unix like PC running and using free and open source software:


Re: E4418B power meter clearing Ram battery failed

 

开云体育

If you have a message going to many, it is customary to use @ to identify a specific recipient.


Such as:

- - - - -

Subject: INVITATION TO PARTY
Sent to: Schlomo, Omo and Dromo


Hey Chaps,

let us meet at Burger King at Oktogon.

@SCHLOMO: the security guard really liked the beer you brought. Can you tell him which brand it was?


With compliments

Bomo

- - - -

With best regards
Tam HANNA 

Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at  
On 2019. 09. 03. 17:56, Glenn Little wrote:

What do the "@" signs mean?
And what is FFF?

I have never seen these used before.

Glenn


On 9/3/2019 11:16 AM, victor smith wrote:
@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:??

-- 
-----------------------------------------------------------------------
Glenn Little                ARRL Technical Specialist   QCWA  LM 28417
Amateur Callsign:  WB4UIV            wb4uiv@...    AMSAT LM 2178
QTH:  Goose Creek, SC USA (EM92xx)  USSVI LM   NRA LM   SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 


Re: E4418B power meter clearing Ram battery failed

 

开云体育

What do the "@" signs mean?
And what is FFF?

I have never seen these used before.

Glenn


On 9/3/2019 11:16 AM, victor smith wrote:
@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:??

-- 
-----------------------------------------------------------------------
Glenn Little                ARRL Technical Specialist   QCWA  LM 28417
Amateur Callsign:  WB4UIV            wb4uiv@...    AMSAT LM 2178
QTH:  Goose Creek, SC USA (EM92xx)  USSVI LM   NRA LM   SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license" 


Re: E4418B power meter clearing Ram battery failed

 

@Steve H., great story...
@Paul B : I got what appears to be FFF identical to the original battery here:?


Re: E4418B power meter clearing Ram battery failed

 

开云体育

Excellent root cause analysis!


Peter

On Sep 3, 2019, at 8:28 AM, Steve Hendrix <SteveHx@...> wrote:

At 2019-09-02 04:50 PM, Richard Knoppow wrote:
?? Assuming the battery test shows the voltage under load I wonder if the old battery got noisy. Batteries do sometimes. I was taught that putting a capacitor across a battery may be good practice because it shorts out the noise. Perhaps does not apply to this device.

This may or may not be relevant to your situation, but I recall a bizarre problem quite a while back (this would have been in the late 1980's) where a purchased CPU board we used in a design as a subcontractor to a prime contractor for NASA, used a primary lithium coin cell as a memory backup battery. After the unit ran continuously and flawlessly for a couple of months, at somewhat elevated temperatures, it started having random resets, becoming worse over time, until it wouldn't boot at all. Since it took months to duplicate, you can well imagine it was tedious to troubleshoot.

It turned out that the designers of that CPU board also used the backup battery as a reference voltage for the power-fail comparator. Running at elevated temps (around 100F / 30C or thereabouts as I recall), the diode used to switch in the backup battery leaked about 1 ?A. This was enough to slowly "charge" the non-rechargeable battery, causing its output voltage to slowly rise, over a period of months. Eventually it got high enough to make the comparator see the main supply voltage pretty close to the reference, and all it took was a little noise to trigger a power-fail reset. As the battery voltage continued to rise, it took less and less noise, so resets became more frequent until they became continuous.

IIRC our solution was to add a 1 megohm resistor across the battery, guaranteeing that it was never charged by diode leakage. This of course meant that under most conditions it had a slightly increased discharge rate, but the battery was way over-specified, with a theoretical lifetime in centuries, so we didn't care about the additional discharge.

Steve Hendrix