Date

Re: HP 3852a extended memory modules info

 

On 5/3/24 13:11, Maciej Kawalkowski, Liberty Electronics Design and Service LLC wrote:
The Hp 9000 DIO board has the same pin out where it matters and looks like the same board type.
Yes, it does look to be the same thing.

I too have a 3852A by the way. It has never given me trouble, so I've never needed to open it up.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: HP 3852a extended memory modules info

 

On 5/3/24 12:59, Maciej Kawalkowski, Liberty Electronics Design and Service LLC wrote:
HP-9000 DIO RAM
Your post helped a lot . I think I found the cloned memory info by Dr. Scott M. Baker :
<>
That's not the project I worked on; the one I was involved in is just pure RAM, 7.5MB to fill the address space. No ROM. There was no discrete logic; everything was implemented in a GAL, and it was all surface-mount.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: 8594E CAL: SPAN SENS FAIL

 

On Fri, May 3, 2024 at 05:38 AM, Scott Townley wrote:
Ozan,
Thank you for the tips.? My observations:
1. I was able to verify +15VF, +5A and +10UF voltages directly on A7.? I need to get an extender to really dig into it :-(
2. In the system diagnostics, +10V Reference (presumably +10UF?) is quite unstable
3. In verifying +10UF, I measured at one end of R105, which connects +10UF to V(REF)+ of the DAC in the third converter ref_cal circuit.? Not knowing which end was which, it was easy enough to find the +10UF (and checking with an o-scope it's quite stable, unlike the system diagnostic), but as it happens the other side of R105 has a very small voltage on it (less than 100 mV).? This doesn't seem right to me.??
As I await my extender, any further insights would be appreciated!
===
Hi Scott,
Other end of R105 goes to a virtual ground inside the IC , seeing ~0V is expected. Current through R105 becomes the reference current for the DAC.

+10V test voltage should go through muxes mentioned in the previous message and show up as 1V at top of R18 (ANA_TEST). If you see 1V then this board is fine, the issue is either connectivity between this board and the CPU board or the MUX on the CPU board is damaged.?

Instead of extenders I usually solder kynar wires to critical test points and take out for measurement. At the measurement end I put different number of knots to distinguish which wire is which. A battery powered soldering iron works well for these small soldering jobs.?
Ozan


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Hi,

I was also looking into these boards to add the 3rd channel and I found
out that all boards on eBay that have more than 3GHz you need to set the
SCALE to 4 otherwise it will not count correctly.

Below are the links for 12GHz and even 18GHz version:

18GHz Model:


12GHz Model:


Regards,
Razvan

On 03/05/2024 19:23, Ed (scskits) via groups.io wrote:
Tony:

The 53131A had three options for the channel 3 input, 3GHz, 5GHz, and
12GHz.

I have never seen anything on EBAY over 3GHz installed in a unit.

I have seen Chinese Channel 3 boards for 3GHz (a lot of them) and 6GHz
(just 1) currently listed.
I do not see the 12GHz board listed at this time, although it was listed
in the past.

I did a board a few years back using the HMC432 (divide by 2) and two
HMC434 (divide by 8) that could get to 6GHz or higher.

A less expensive version would use the HMC432 driving an MC12079
programmed to divide by 64, have the layout for it but never made it.

Never did any follow up on these because the Chinese clone boards were
so inexpensive it was not worth pursuing.

ed


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Tony:

The 53131A had three options for the channel 3 input, 3GHz, 5GHz, and 12GHz.

I have never seen anything on EBAY over 3GHz installed in a unit.

I have seen Chinese Channel 3 boards for 3GHz (a lot of them) and 6GHz (just 1) currently listed.
I do not see the 12GHz board listed at this time, although it was listed in the past.

I did a board a few years back using the HMC432 (divide by 2) and two HMC434 (divide by 8) that could get to 6GHz or higher.

A less expensive version would use the HMC432 driving an MC12079 programmed to divide by 64, have the layout for it but never made it.

Never did any follow up on these because the Chinese clone boards were so inexpensive it was not worth pursuing.

ed


Re: HP 3852a extended memory modules info

 

Dave
The Hp 9000 DIO board has the same pin out where it matters and looks like the same board type.


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Jared:

I was looking at the 53131A specs and the minimum input spec for 100MHz to 2.7GHz is -27dBm and from there to 3Ghz is -21dBm.
I guess actual levels would be less than that , but not significantly or Agilent would have taken credit for it.

For the channel A and B inputs the levels range from 20mV to 40mV (-21dBm to -15dBm if terminated in 50 ohms).

Your board must have a lot of gain at the low end for the minimum input to be below -50dBm for frequencies below 800Mhz as seen in the PNG image you posted.

Like you, I also noticed that the input protection diodes affected the signal levels. More effect than I expected.?

I am currently using two Mini Circuits ERA amplifiers for the gain stage with a 3dB attenuator in line with the input.

My minimum input levels from 750MHz to 3GHz are somewhat in line with the Agilent spec, -34dBm to -21dBm.
At 500Mhz, the input level drops to -13dBm, I think my input cap and those in line may be too small.

ed


Re: HP 3852a extended memory modules info

 

On Fri, May 3, 2024 at 10:01 AM, Dave McGuire wrote:
HP-9000 DIO RAM
Dave

Your post helped a lot . I think I found the cloned memory info by Dr. Scott M. Baker :


Maciej


Re: HP 3852a extended memory modules info

 

Ed

The connector is part of the machine and the memory card uses PCB edge as connector.
There is a raiser inserted in onto the 3852A processor board like the one below.
The memory card interfaces this raiser.
I was thinking of making a card using? SRAM .? The 4Mbit chips? R1LP0408DSP-5SI (512kx8) would be a good choice.? 2 chips for 1MB card.
Some things I am not sure of are:
  • how the? addressing works on these systems
  • how the detection of RAM is determined? by the CPU
  • how DTACK signal is generated
  • what address range should be used (HP had dip switched)
  • how address decoding would be done with dip switches for picking range


Maciej


Re: HP 130A restoration

 

If the CRT has an intact heater & shiny silver getter, it is probably fine. Lack of EHT is more likely to be a problem, especially if there are leaky paper based capacitors in the EHT circuit, or a bad step up transformer.
Also there are several 5AQP type CRTs on ebay, two which look to have been pulled from old oscilloscopes and one very expensive NOS.

David


Re: HP 8360-Series Sweeper CLIP soon available on KO4BB

 

开云体育

Hi Tom

In the folder 07 Recent uploads it is listed as but no to download

HP 83620 86340 86350 Synthesized Sweeper Service Manual-HP8360-SeriesComponent-LevelServiceManualVol1&2

Kind regards

Kurt

Fra: [email protected] <[email protected]> P? vegne af tom_iphi via groups.io
Sendt: 3. maj 2024 11:56
Til: [email protected]
Emne: Re: [HP-Agilent-Keysight-equipment] HP 8360-Series Sweeper CLIP soon available on KO4BB

?

Am I too stupid to find it or hasn't it been verified yet?
I'm excited to see the CLIPs.

Tom DG8SAQ


Re: HP 130A restoration

 

One more generic question: up to now I have never seen any evidence that the CRT is working, if I get to the point of having the whole circutry working again and I discover that the CRT is not recoverable will I be able to find a replacement? I already checked the heater and that one seems to have the right resistance, the look of the gettered surface seems to have the right look too. Thank you for your comments.


Re: HP8563E Blank display

 

SER is 3305A00996
Don't know if that helps,
Michel


Le?ven. 3 mai 2024 à?17:37, F6DZK Michel via <michel.m.bourdon=[email protected]> a écrit?:
Hi David,
I saw that DIMM like socket on Agilent pictures. However my HP8563E is older (don't know the year) and A6 PCB has no such board.
Regards, Michel

Le?ven. 3 mai 2024 à?17:29, David via <david=[email protected]> a écrit?:
The first thing you should try is to reseat the power supply module in the main power supply. This module is similar to a memory module in an old PC, and it plugs into a DIMM socket.


Re: HP8563E Blank display

 

Hi David,
I saw that DIMM like socket on Agilent pictures. However my HP8563E is older (don't know the year) and A6 PCB has no such board.
Regards, Michel


Le?ven. 3 mai 2024 à?17:29, David via <david=[email protected]> a écrit?:
The first thing you should try is to reseat the power supply module in the main power supply. This module is similar to a memory module in an old PC, and it plugs into a DIMM socket.


Re: HP8563E Blank display

 

The first thing you should try is to reseat the power supply module in the main power supply. This module is similar to a memory module in an old PC, and it plugs into a DIMM socket.


Re: HP 130A restoration

 

I’ll have to look at the schematic again, as I didn’t notice the dependence (also) of 88V on the 300V supply. HP simplified their ‘scope power supplies subsequently, making one supply the standard for regulating all the others.?

Jeremy?


On Fri, May 3, 2024 at 3:55 AM Bepi via <pepicima=[email protected]> wrote:
Sorry a correction, by disconnecting the J9P interpanel connector I have disconnected the V33 heater so I could not have had the right voltage out anyway from the -150 supply, regardless I was getting around -165 on C140B. The dV across V34, 80+V, is about right.
Riconnecting J9P only and not J10P the 88V supply should not have worked regularly also since there the 300 V supply contribution is missing? , I am getting 31V instead of 88. Around 200V on the V28 and V29 plates. I tried to regulate the 150V supply and I could do this verya ccurately.


HP8563E Blank display

 

I have a blank display on my HP8563E. The CRT is fine but I found out that -15 VDC power supply is down (LED on A2 is off). I measured nearly 0 V on TP303 (vs TP301) on A6.
Anyone came across the same problem?

Best regards, Michel F6DZK


Re: HP 3852a extended memory modules info

 

On 5/3/24 07:20, Ed (scskits) wrote:
If the connector is still available, it would not be difficult to make a 1MB or 8MB static RAM board.
I did one not too long ago for the Data I/O UNISITE that also uses the MC68000.
The original UNISITE board was DRAM with up to eight 9 bit modules.
With the static RAM, the parity was disabled.
Nice!

If that is in fact an HP-9000 DIO RAM board, they've already been cloned. A fellow over in .eu did the design, I had a small run of PCBs done and built a few, and eventually got it working. I had intended to do a production run and sell them, but there wasn't a lot of interest on the vinthpcom list, surprisingly, so there it has sat for the past few years.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: HP 8360-Series Sweeper CLIP soon available on KO4BB

 

开云体育

Hi Tom,

?

The file is in the RECENT DOWNLOAD list. We have to wait for verification before we can do the download.

?

?

Yves

?

De?: [email protected] <[email protected]> De la part de tom_iphi via groups.io
贰苍惫辞测é?: 3 mai 2024 05:56
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 8360-Series Sweeper CLIP soon available on KO4BB

?

Am I too stupid to find it or hasn't it been verified yet?
I'm excited to see the CLIPs.

Tom DG8SAQ


Re: 8594E CAL: SPAN SENS FAIL

 

Ozan,
Thank you for the tips.? My observations:
1. I was able to verify +15VF, +5A and +10UF voltages directly on A7.? I need to get an extender to really dig into it :-(
2. In the system diagnostics, +10V Reference (presumably +10UF?) is quite unstable
3. In verifying +10UF, I measured at one end of R105, which connects +10UF to V(REF)+ of the DAC in the third converter ref_cal circuit.? Not knowing which end was which, it was easy enough to find the +10UF (and checking with an o-scope it's quite stable, unlike the system diagnostic), but as it happens the other side of R105 has a very small voltage on it (less than 100 mV).? This doesn't seem right to me.??
As I await my extender, any further insights would be appreciated!