开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育
Date

Re: HP 3325A Function Generator A3 Mixer Troubleshooting U15 And R106

 

Hi Al,
I read the timebase as 2ms/div. Those flat top waveforms are 8-10ms long and well defined. They look like IR drops more than ripple. All the signals in mixer section are 30MHz-50MHz range; however, divider section of A3 (same board) has slower signals.?

Even with this waveform on +5V I don't think it is the root cause of missing signal at TP7. ?
Ozan


On Tue, Oct 10, 2023 at 01:53 PM, @GP53 wrote:
Ozan,

Finally got to the power supply to check ripple. The +15 is 29mV, the -15 is 15mV, and the +5 is 23mV. But the waveform at the +5 on the scope(ac coupled) is troublesome. The waveforms at both 15V looked pretty normal.


Re: semi-OT: Cleaning smoke smell out of electronics

 

Depends on what the equipment is. ?I’ve been in the same predicament in the past. ?

I’ve had good luck with a good surfactant soap, things like Simple Green and such. ?Take the equipment apart as best as possible and rinse the case and chassis well, and get in there with the soap and a short bristle brush/toothbrush and go at it. ?Rinse well. ?Warm air blower dry. ?PC boards the same. ?I’ve found 70’s and after vintage components are pretty well sealed for a short bath. ?Older stuff with paper and/or cotton insulation, not so well. ?Point to point wiring is a bitch to get in and at all the nooks and crannies. ?Old electolytics probably aren’t sealed well due to aging, but more than likely they’d need replacing anyway.

Once it’s fully and throughly dried out, be prepared to clean and lube switch contacts, shafts, pots, plugs and connectors, etc. ?

I avoid smoker equipment because of the time and effort involved. ?It’s gotta be a real good deal to go through the cleaning dance to make it worth it.

Have fun. ?Wear rubber gloves as all the soap and wet does a number on the natural skin oils in your hands. ?Ask me how I know this…


semi-OT: Cleaning smoke smell out of electronics

 

I just received a piece of gear that positively REEKS of cigarette smoke. I probably should have asked if it was from a non-smoking environment before buying, but it's too late now. What's the best way to get rid of the smell.

--
--Eric
_________________________________________
Eric Garner


Re: help repairing 8133a

 

开云体育

Hi,

?

Looking at this trace on the graph, what catches my attention is the fact that the control voltage applied to the YTO makes jumps at the start and end of the YTO range, and is linear in places, but mostly the fact that part is curved.

?

Maybe the YTO driver is more complex than I imagine. Perhaps there is switching in the transconductance amp powering the YIG coils, and one or more of these circuit sections are distorting...hard to say.

?

I saw a friend of mine yesterday, and he had three 8133As available to do some quick testing. On 2 of these generators tested, there was a relay problem, one was on the CH2, normal output, and the other was on the reversed output. Their problems were exactly the same as on my 8133A which I ended up repairing. No doubt this problem is recurring on these generators.

?

I measured the FM modulation at Spectrum at 500 MHz on these two generators, and I had the same result as on mine...this would seem normal behavior.

?

I had the opportunity to use a CS803C and SD-22, and it was possible to watch the signal in PRBS mode. However, in the trigger menu it was necessary to use the EXTERNAL DIRECT trig, to switch from VARIABLE PERSISTENCE mode.

?

For the CH2 relay, if the repair fails, this part is available on eBay.

?

De?: [email protected] <[email protected]> De la part de Xtremexp
贰苍惫辞测é?: 12 octobre 2023 16:04
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] help repairing 8133a

?

Thanks Yves,

Collecting data for frequency accuracy analysis is lengthy task. I wrote a powershell script to set frequency, read 8133A internal frequency counter reading (and the external frequency counter connected to the ch1 output.
I stepped frequency 1Mhz from 34Mhz to 3.5Ghz. The plot below shows the frequency error (set-reading from 8133A internal counter). Indeed there is some frequency issue in some bands, which does not have explanation.


--------------
Regarding ch2 relay issue, I plan to open my unit tomorrow and try to fixt it. Let's see how it goes. Thanks for your step by step instruciton.

Thanks!


Re: help repairing 8133a

 

Thanks Yves,

Collecting data for frequency accuracy analysis is lengthy task. I wrote a powershell script to set frequency, read 8133A internal frequency counter reading (and the external frequency counter connected to the ch1 output.
I stepped frequency 1Mhz from 34Mhz to 3.5Ghz. The plot below shows the frequency error (set-reading from 8133A internal counter). Indeed there is some frequency issue in some bands, which does not have explanation.


--------------
Regarding ch2 relay issue, I plan to open my unit tomorrow and try to fixt it. Let's see how it goes. Thanks for your step by step instruciton.

Thanks!


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

I would not use a switching supply on the prescaler board.
I think at about 50mA, a DPACK or SOT-223 regulator should do the job and stay relatively cool..
Even the SOT-89 should be OK as the most of the heat will go to the ground plane through the tab.
And on that note, the tab should be embedded in the ground plane, no relief connection.
That is assuming that the tab is the ground pin on the regulator.


Re: Book draft update

 

Gianni,

The HP Book looks amazing (I already enjoyed the Tektronix one which partly led me astray into acquiring a 7904) - some amazing pin-sharp photographs and equipment journeys.

But clearly in England our education system focuses on the things that really matter(?). Should the latin quote not be "ab uno disce omnes" ?(Virgil, Aeneid Book 2) - From this one [proof of their treachery] you may understand them all (i.e. not the infinitive discere)....

Alan


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

There are a number of such chips, one particularly nice one is an isolated DC supply 5 volts to 5 volts. TEA1-0505 (Traco), this is rated for 200 ma with an isolation of 1500 volts.? I was somewhat surprised to find these chips (and there are bigger ones that are adjustable and at more current, no TO-220 form factor, though).

Harvey

On 10/11/2023 3:15 PM, warrens@... wrote:
Re the L78M05CDT you could also try a Traco Power TSR 2-2450.

This is a switchmode drop in replacement for LM7805 with a 2mA idle current, rated for 2A. It does not need heatsinking. I've used them when repairing the power supply in multimeters which have closed cases and no active cooling (Fluke 8000 series bench meters).

Warren


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Re the L78M05CDT you could also try a Traco Power TSR 2-2450.

This is a switchmode drop in replacement for LM7805 with a 2mA idle current, rated for 2A. It does not need heatsinking. I've used them when repairing the power supply in multimeters which have closed cases and no active cooling (Fluke 8000 series bench meters).

Warren


Re: Book draft update

 

Hello,

thank you so much for the great book - just put it on Instagram!


Tam


Re: help repairing 8133a

 

开云体育

Hi,

?

For the relay problem:

- I decided to try the disassembly again and it worked correctly this time.

- step#1: remove the cover, there are two screws, one of which is hidden under the label.

- step #2: you have to dismantle the relay coil to go further and it is difficult.

- step#3: I had no choice but to cut the screw head, drill and remove the debris from the screw. The replacement screw was a 2-56.

- step #4 and #5, shows the RF base including the mechanics of the relay without the activation coil.

- step #6: dismantling the external mechanics of the relay.

- step #7: separation of the RF and mechanical part of the relay. At this point the goal was to clean the gold-plated blades and contacts and reassemble that part. Reassembly was quite difficult because it requires precision. After this step was done, I tested with the VNA and when I manually pushed on the blade above the spring that we see in steps #8 and #9, everything was going well with the VNA.

- After step #7, I redid the assembly without the cover and tested everything to realize that it was not yet working correctly.

- I ended up understanding that the problem was the distance between the blade pushing on the spring base which didn't have enough pressure. So I did the disassembly in step #6, I adjusted the thin blade attached to a metal part, so that the blade barely touches the base of the spring as shown in step #9... and this solved the problem.

?

So, here is my advice: do steps 1,2 and 3 and look at the blade shown in 8 and 9. If it does not touch the base of the spring, do step 6, adjust the blade to get a result like step 9 and reassemble everything without the cover and do a test on your workbench. The relay operates at +15V (97mA), pay attention to polarity. The test can be done with an oscilloscope, if you do not have a VNA or Spectrum. Choose a low frequency for the test first, for example 50 MHz, you will quickly see if the signal passes when the relay is activated. If unfortunately, the relay still has the same erratic behavior, then you will have to go further and do step 7...and that is not easy at all.

?

----------------------

?

I did the same test as you to compare, and it's difficult to explain this behavior.

?

--------------------

?

I also did this automated performance test, which is the error of the output frequency vs the programming of that frequency.

The acceptance limit is ±0.5%, but typically the error should be close to ±0.1%. This test checked all programmable frequencies from 31.3MHz to 3.5GHz in steps of 100kHz, therefore 34688 frequencies, so this test lasted several hours. The error result also includes that of thermal drift during the test.

?

?

-------------------------

?

Tomorrow I'm going to go see a friend to do a comparative test regarding frequency stability on another 8133A.

I will probably have the opportunity to do a test on a Tektronix 11801B with SD-26 and I will check the trigger mode with a signal from the 8133A. I will give you my comments if this test can ever be done.

?

?

?

De?: [email protected] <[email protected]> De la part de Xtremexp
贰苍惫辞测é?: 9 octobre 2023 20:46
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] help repairing 8133a

?

Yeah,

I did ran the command and I get following, so It seems calibration is OK (it is linear)

:DIAG:CAL:YIG?
+2.02000E+09,+9.55000E+02,+2.56030E+09,+1.54500E+03,+3.02723E+09,+2.05600E+03,+3.49605E+09,+2.56700E+03,+3.95992E+09,+3.07200E+03

I re measured and plotted the set frequency and the output frequency on ch1 (using frequency counter). I see below plot. Yellow trace is the YIG frequency from 2 to 4Ghz (expected) and the blue trace is the actual frequency output, x axis is the set frequency. I measured manually from 33 Mhz to 500Mhz. Its clear that the output frequency has some issues and not linear with set frequency.


I still do not understand how 12 bit DAC translates to the YIG output, May be some bit of DAC is an issue?.

By the way, My timing board was coming from Israel, and due to what's going on over there, it is not certain if I will get it or not.?

If you end up fixing RF relay, then I will follow your pursuit and fix mine also. :)

Thanks alot.


Re: Help needed: simple THT soldering in Yorktown, Virginia, USA

 

Hello Folks,

sorry for being slow in responding to youse all.


I managed to return to my office yesterday, and FedEx the board. It should arrive by Friday, so I think this is the fastest way.


Thanks again for your willingness to help - I was hoping for someone to be "round the corner" who could just pop over. But alas, the Americas are a large and wide space...


Be that as it may, yes, I will teach Kiddo soldering one day. But we need to burn some ESP32 boards quickly...


Tam

--
With best regards
Tam HANNA

Enjoy electronics? Join 21k4 other followers by visiting the Crazy Electronics Lab at


HP 8665A latest firmware

 

Does anyone have a dump of the most up-to-date firmware for the HP 8665A that they can share?
I believe that it is version 4.2.1. I'd like to update my sweeper that is at 3.4.0

thanks,
? ?Michael


Re: HP 3325A Function Generator A3 Mixer Troubleshooting U15 And R106

 

Ozan,

Finally got to the power supply to check ripple. The +15 is 29mV, the -15 is 15mV, and the +5 is 23mV. But the waveform at the +5 on the scope(ac coupled) is troublesome. The waveforms at both 15V looked pretty normal.


Re: Book draft update

 

Gianni I downloaded the low resolution documents for HP, Tek, and Surplus and will say really great reads from a technical and historical perspective. Very nice pictures and details. Why things are the way they are.
Just a great read.
Thank you so much for your hard work.
Regards
Paul
WB8TSL


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

If it doesn't work, then change the incantation, go have lunch, and start with a new chicken.

Harvey

On 10/10/2023 3:51 AM, Tom Lee wrote:
No worries -- this was all new stuff to all of us at one time or another.

As for black magic, it's got that rep, sure, but after a while you realize that it's all pretty straightforward. All you need is a wizard's hat and a chicken. Adjust the latter as necessary.

-- Cheers,
Tom


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

No worries -- this was all new stuff to all of us at one time or another.

As for black magic, it's got that rep, sure, but after a while you realize that it's all pretty straightforward. All you need is a wizard's hat and a chicken. Adjust the latter as necessary.

-- Cheers,
Tom

--
Prof. Thomas H. Lee
Faculty Co-Director, SystemX Alliance
Director, Stanford-Samsung Research Initiative

Allen Ctr., Rm. 205
420 Via Palou Mall
Stanford University
Stanford, CA 94305-4070

On 10/10/2023 12:46 AM, Jared Cabot via groups.io wrote:
It makes sense. I'm still learning this black magic high frequency stuff, so I may have a few more basic-level questions to come. Bear with me.... :)


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

On Tue, Oct 10, 2023 at 11:10 AM, Ed Breya wrote:
You don't have to care about the diode reverse voltage ratings - when anti-parallel, each can only see forward voltages, unless the current rating is exceeded. What matters is that (indicates the RF power it can clamp at the input), and the zero-bias capacitance (for speed).
Of course.... That makes perfect sense.


The prescaler inputs are symmetric. You can treat it as a differential amplifier, and put the input signal to either one, while the other is bypassed - it will toggle one opposite edges if swapped. The ECL DC bias (Vbb) is usually developed by some sort of internal regulator circuitry, and each input connects to it through some medium resistance maybe in the hundreds of ohms range. You can put the DC bias into either one, just like the signal. What you want to do is upset the bias level a little bit to reduce the gain. HP used a selected R value for pulling the bias, and you can do the same if you can figure out how much is right. If you're designing for "production," it may be better to use a fixed resistor in series with a pot, to effectively select a value that's right for the particular part.
Yeah, this application doesn't care if the signal is inverted so we are free to do what we want.
This one will be just a small run to cover my initial costs so a selected resistor would be fine, and then the full design files will be released, so maybe I'll make provision in v2.0 for a selected resistor and a trimpot depending on the builders preference.


You can get some good ideas by studying the old ECL parts and documents. My favorite info source for this is the Motorola from the 1980s - I think the data book was called "MECL-III" family. Later generations like 10H and up started improving the performance, always tending faster, and with temperature-compensated Vbb,, lower operating voltage, and more function variety. A good proxy for a prescaler input section is the 116 or 216, for instance. These basic models extend from MC10216, to 10H216, to very high speed into the microwave 100E series and such. The 116 and 216 are triple line receivers, among the most highly used types, for interfacing small signals to ECL. Picture a single part with the three stages cascaded for high gain, so you can toggle the following flip-flops with mV input signals - if you have it biased right. The line receiver type parts almost always include access to the Vbb for setting up these scenarios with external resistors. The prescalers have the same setup with internal resistors, and don't provide access to Vbb. But, you can measure it - it's roughly the DC voltage present on the input pins.

Ed
I'll check that out, thanks!



On Tue, Oct 10, 2023 at 12:13 PM, Tom Lee wrote:
Hi Jared,

I assume that a decimal point went AWOL and that you meant 0.26pF. I
have never seen a 26pF diode with a 4V breakdown -- not because it's
impossible, but because there is no market for one.
Ah yup, a typo there, 0.26pF is what I meant. :D


A good guide for how much is too much is to calculate the capacitance
reactance at the highest frequency, and compare it to what shunts it in
that circuit (50 ohms, e.g.). At 1GHz, say, a 1pF cap presents about a
160 ohm reactance, which is larger, but not very much so, than 50 ohms.
At 10GHz, it's 16 ohms, which begins to look more like a short compared
to 50 ohms. You get the idea.

-- Cheers
Tom
It makes sense. I'm still learning this black magic high frequency stuff, so I may have a few more basic-level questions to come. Bear with me.... :)



Thanks!
Jared