¿ªÔÆÌåÓý

Date

Re: Emails to [email protected] blocked

 

Based on recent painful experience, I've discovered a couple of things:

1. Some ISPs block based on listing in some questionable databases. In particular, the "UCEPROTECT" lists range from including just the sender IP, to the whole subnet of that IP, to the whole ASN (assigned network number) of that IP, which can block millions of addresses. Smart ISPs assign only some weight in their spam filters to databases like that. But it seems not all are smart, or they get accidentally misconfigured. Our ISP recently had a big fight with MSFT over this.

2. Big ISPs like Google and Microsoft have gotten more picky about the sending system having correct DMARC, DKIM, and SPF records in their DNS. If you have your own domain and set up your own DNS, you may not have those records set up. That didn't used to matter much, but we just solved several problems by getting our records fixed (we had some outdated info after our ISP made changes to their system).

John
----

On 5/28/22 14:18, Paul Amaranth wrote:
Actually, that means the IP of your email server is blocklisted. If
you do a reverse lookup on that, it resolves to mail-pj1-f53.google.com.
You can look up IPs on line to see which, if any, blocklist it appears
in. A couple sites are mxtoolbox.com and www.debounder.com. This server
does appear in a two or three blocklists.
What you need to do is contact whomever manages your email system and have
them apply to be removed from the blocklist. This has to be from
an administrator of the domain, not a user, and any spam issues have to
be corrected.
Sometimes there may be some configuration issues at the mailer that
rejects the mail. On my system I was getting transient blocklist failures
rejecting perfectly valid senders and I had to disable that feature.
Paul
On Sat, May 28, 2022 at 07:52:43AM -0600, John Griessen wrote:

On 5/27/22 18:03, Matt Huszagh wrote:
[209.85.216.53]:42889 is in an RBL
This means your IP is blacklisted. is there a URL for remedy in the message? You can ask to be unblacklisted.
It can be because your ISP allows too many bad customers sometimes.







!DSPAM:62922936133621664620935!


Re: Emails to [email protected] blocked

 

On Sat, May 28, 2022 at 02:18:47PM -0400, Paul Amaranth wrote:
Actually, that means the IP of your email server is blocklisted. If
you do a reverse lookup on that, it resolves to mail-pj1-f53.google.com.

You can look up IPs on line to see which, if any, blocklist it appears
in. A couple sites are mxtoolbox.com and www.debounder.com. This server
does appear in a two or three blocklists.
Oops, typo. That was www.debouncer.com

--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@... | Unix/Linux - We don't do windows


Re: Emails to [email protected] blocked

 

Actually, that means the IP of your email server is blocklisted. If
you do a reverse lookup on that, it resolves to mail-pj1-f53.google.com.

You can look up IPs on line to see which, if any, blocklist it appears
in. A couple sites are mxtoolbox.com and www.debounder.com. This server
does appear in a two or three blocklists.

What you need to do is contact whomever manages your email system and have
them apply to be removed from the blocklist. This has to be from
an administrator of the domain, not a user, and any spam issues have to
be corrected.

Sometimes there may be some configuration issues at the mailer that
rejects the mail. On my system I was getting transient blocklist failures
rejecting perfectly valid senders and I had to disable that feature.

Paul

On Sat, May 28, 2022 at 07:52:43AM -0600, John Griessen wrote:

On 5/27/22 18:03, Matt Huszagh wrote:
[209.85.216.53]:42889 is in an RBL
This means your IP is blacklisted. is there a URL for remedy in the message? You can ask to be unblacklisted.
It can be because your ISP allows too many bad customers sometimes.







!DSPAM:62922936133621664620935!
--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@... | Unix/Linux - We don't do windows


Re: checking line module setting for 110V / 220V

 

Many thanks for your great feedback.
It gave me enough courage to pull the little card out.
I found I had a pair of pincers with a hook on the tip.
No idea what it was originally for, but worked great.
:-)

Best regards,
Tom


Re: HP8640B no rf out

Lothar baier
 

¿ªÔÆÌåÓý

The 8557/58/59 used the same kind of disks you could possibly find a parts mule on ebay though there is no guarantee that its not going to suffer from the same disease as this was a common problem.
Take note though that the springs are attached to the disks by means of small plastic bumps that are molted to secure them , thermoplastics become briddle over time and the bumps shear off.
The way i fixed them back at tucker was to get small self tapping screws and use a dremel tool to drill holes in the disks


On May 28, 2022, at 11:28, kim.herron@... via groups.io <kim.herron@...> wrote:

?
You can try contacting Sphere Research in Vancouver BC, Canada.? They have a lot of HP parts.? Those wipers are part of the discs that are in the switch assy and no part number exists for the wipers as such.? However, there may be other HP gear that uses something similar that would yield the items that you need to fix your generator.? You can also keep an eye out for an entire
switch assy from an 8640B on E-Bay.? That stuff is out there.? The reason that those wipers break off is because of the gears cracking.? Those cracked gears force the tension spring in the rear of that switch assy? to collapse and when that happens the wipers on those discs are forced to flex beyond their design limits and that's what breaks them off the discs.? Make sure your gears are not coming apart.



On 28 May 2022 at 8:06, Donald Prins wrote:

Hi Ellud,

Can you tell me where one can obtain those wipers? I have a HP8640B that is missing 3 in total on the FM/RANGE switches. Your help will be much appreciated.

Donald?
PE1AHJ


Kim Herron? W8ZV
kim.herron@...????
1-616-677-3706


Re: HP8640B no rf out

 

¿ªÔÆÌåÓý

You can try contacting Sphere Research in Vancouver BC, Canada.? They have a lot of HP parts.? Those wipers are part of the discs that are in the switch assy and no part number exists for the wipers as such.? However, there may be other HP gear that uses something similar that would yield the items that you need to fix your generator.? You can also keep an eye out for an entire
switch assy from an 8640B on E-Bay.? That stuff is out there.? The reason that those wipers break off is because of the gears cracking.? Those cracked gears force the tension spring in the rear of that switch assy? to collapse and when that happens the wipers on those discs are forced to flex beyond their design limits and that's what breaks them off the discs.? Make sure your gears are not coming apart.



On 28 May 2022 at 8:06, Donald Prins wrote:

Hi Ellud,

Can you tell me where one can obtain those wipers? I have a HP8640B that is missing 3 in total on the FM/RANGE switches. Your help will be much appreciated.

Donald?
PE1AHJ


Kim Herron? W8ZV
kim.herron@...????
1-616-677-3706


Re: HP8640B no rf out

 

¿ªÔÆÌåÓý

In the archive you will find discussion of this, including at least one account of improvising new contacts out of something else.

?

Dave Wise

?

From: [email protected] <[email protected]> On Behalf Of Donald Prins via groups.io
Sent: Saturday, May 28, 2022 8:06 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B no rf out

?

Hi Ellud,

Can you tell me where one can obtain those wipers? I have a HP8640B that is missing 3 in total on the FM/RANGE switches. Your help will be much appreciated.

Donald?
PE1AHJ


Re: HP8640B no rf out

 

Hi Ellud,

Can you tell me where one can obtain those wipers? I have a HP8640B that is missing 3 in total on the FM/RANGE switches. Your help will be much appreciated.

Donald?
PE1AHJ


Re: Emails to [email protected] blocked

 

On 5/27/22 18:03, Matt Huszagh wrote:
[209.85.216.53]:42889 is in an RBL
This means your IP is blacklisted. is there a URL for remedy in the message? You can ask to be unblacklisted.
It can be because your ISP allows too many bad customers sometimes.


Re: Info Need on a RF FET for a Agilent E4438C signal generator

Lothar baier
 

¿ªÔÆÌåÓý

Its a common issue that you have a lot of excess gain at lower frequencies and then roll off at higher frequencies normally you can combat this by using resistive feedback between drain and gate , it will change your input impedance and drop gain at lower frequencies!
I think the best approach would be to find a MMIC that will work , i looked through offerings from different companies and nobody seems to make discretes in SOT89 anymore?


On May 28, 2022, at 03:13, Peter Hansen via groups.io <oz1lpr@...> wrote:

?
Hello Lothar I do not need 27dBm output. As it is not the final Fet. but I think the gain of the SHF-0189 is too little. It has around 20dB at low freq but only 10 to 12dB at 4Ghz. so I might need more gain
best regards Peter

Fra: [email protected] <[email protected]> p? vegne af Lothar baier <Lothar@...>
Sendt: 26. maj 2022 18:56
Til: [email protected] <[email protected]>
Emne: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator
?

The problem is finding a HEMT in SOT89 with 27dBm ,? used to those parts were made as driver or pre-driver for Power amps but this is all done with MMIC now ,? I remember seeing a GaN HEMT made by MACOM that was SOT89 but I believe it was a 28V part !

?

As far as MMIC you can try the TQP7M9103 , its specified 400-4000MHz but you can use it at lower frequencies , this part does require external matching and only produces 13-14Db gain at 4GHz but can drive 27dBm

?

?

From: [email protected] <[email protected]> On Behalf Of Peter Hansen via groups.io
Sent: Thursday, May 26, 2022 11:23 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

Hello Lothar It is a Fet. I does have a biasing system where is measures to drop over the Drain supply and then give bias to the Gate. The circuit does include matching network. I added a SHF-0189 in the circuit and it works but too low output for the High output option mounted. I can find traces on the Net over the M11X but only that it is a HFET.

The Fet is in the ALC loop so the amplitude is not really a problem as long as there is enough. I am afraid that is not the case. I wonder if a MMIC would work in the spot. I need one working from 250khz to 4Ghz. the 4 to 6Ghz path is another board so the Part only need to go to 4Ghz.

sorry no schematics over the E44XX series

best regards Peter OZ1LPR


Fra: [email protected] <[email protected]> p? vegne af Lothar baier <Lothar@...>
Sendt: 26. maj 2022 06:19
Til: [email protected] <[email protected]>
Emne: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

Do you have the schematic of the stage ?? if so can you share it ?

Let me render a few comments / thoughts on this :

  1. The first thing that needs to be established is whether the original part is a FET or a MMIC amplifier ,? you can tell by looking at the schematics , if there is a bias applied to both input and output or just the output of the block , if bias is only applied to the output than its more than likely a MMIC amplifier if bias exists on both gate and drain than it¡¯s a FET/HEMT !
  2. If the part is a FET than there should be matching as well as feedback networks present , since this is a stage operating over a wide frequency range resistive feedback between gate and drain will most certainly be required in addition to lumped input and output matching circuits
  3. Based on the gain and output power estimate you provided as well as the frequency range of the generator I question the assessment of this part being a FET and rather assume this being a MMIC as it would be hard to reach 20Db gain and 27dBm P1Db up to 4GHz
  4. If the part is a FET using matching networks it will be hard to find a drop in replacement , if you were to find a part in a SOT-89 package you will most certainly have to re-design the matching networks as S-parameters will be different

?

?

From: [email protected] <[email protected]> On Behalf Of Peter Hansen via groups.io
Sent: Wednesday, May 25, 2022 4:16 AM
To: [email protected]
Subject: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

I am rapiring a E4438C generator. There is a FET shorted in the output of chain of the 4Ghz path. The FET is called M11X1005. I can see traces of it existing but no data. The SHF-0189Z can be used but not directly. The M11X1005 is using 6.5V on the drain but the SHF-0189 uses 8V and has higher gain and higher output.

Does anyone have either a M11X1005 Fet, data or a suitable repleacement.

I would assume something having 20dB of gain and +27dBm output.

best regards Peter


Re: Info Need on a RF FET for a Agilent E4438C signal generator

 

¿ªÔÆÌåÓý

Hello Lothar I do not need 27dBm output. As it is not the final Fet. but I think the gain of the SHF-0189 is too little. It has around 20dB at low freq but only 10 to 12dB at 4Ghz. so I might need more gain
best regards Peter

Fra: [email protected] <[email protected]> p? vegne af Lothar baier <Lothar@...>
Sendt: 26. maj 2022 18:56
Til: [email protected] <[email protected]>
Emne: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator
?

The problem is finding a HEMT in SOT89 with 27dBm ,? used to those parts were made as driver or pre-driver for Power amps but this is all done with MMIC now ,? I remember seeing a GaN HEMT made by MACOM that was SOT89 but I believe it was a 28V part !

?

As far as MMIC you can try the TQP7M9103 , its specified 400-4000MHz but you can use it at lower frequencies , this part does require external matching and only produces 13-14Db gain at 4GHz but can drive 27dBm

?

?

From: [email protected] <[email protected]> On Behalf Of Peter Hansen via groups.io
Sent: Thursday, May 26, 2022 11:23 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

Hello Lothar It is a Fet. I does have a biasing system where is measures to drop over the Drain supply and then give bias to the Gate. The circuit does include matching network. I added a SHF-0189 in the circuit and it works but too low output for the High output option mounted. I can find traces on the Net over the M11X but only that it is a HFET.

The Fet is in the ALC loop so the amplitude is not really a problem as long as there is enough. I am afraid that is not the case. I wonder if a MMIC would work in the spot. I need one working from 250khz to 4Ghz. the 4 to 6Ghz path is another board so the Part only need to go to 4Ghz.

sorry no schematics over the E44XX series

best regards Peter OZ1LPR


Fra: [email protected] <[email protected]> p? vegne af Lothar baier <Lothar@...>
Sendt: 26. maj 2022 06:19
Til: [email protected] <[email protected]>
Emne: Re: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

Do you have the schematic of the stage ?? if so can you share it ?

Let me render a few comments / thoughts on this :

  1. The first thing that needs to be established is whether the original part is a FET or a MMIC amplifier ,? you can tell by looking at the schematics , if there is a bias applied to both input and output or just the output of the block , if bias is only applied to the output than its more than likely a MMIC amplifier if bias exists on both gate and drain than it¡¯s a FET/HEMT !
  2. If the part is a FET than there should be matching as well as feedback networks present , since this is a stage operating over a wide frequency range resistive feedback between gate and drain will most certainly be required in addition to lumped input and output matching circuits
  3. Based on the gain and output power estimate you provided as well as the frequency range of the generator I question the assessment of this part being a FET and rather assume this being a MMIC as it would be hard to reach 20Db gain and 27dBm P1Db up to 4GHz
  4. If the part is a FET using matching networks it will be hard to find a drop in replacement , if you were to find a part in a SOT-89 package you will most certainly have to re-design the matching networks as S-parameters will be different

?

?

From: [email protected] <[email protected]> On Behalf Of Peter Hansen via groups.io
Sent: Wednesday, May 25, 2022 4:16 AM
To: [email protected]
Subject: [HP-Agilent-Keysight-equipment] Info Need on a RF FET for a Agilent E4438C signal generator

?

I am rapiring a E4438C generator. There is a FET shorted in the output of chain of the 4Ghz path. The FET is called M11X1005. I can see traces of it existing but no data. The SHF-0189Z can be used but not directly. The M11X1005 is using 6.5V on the drain but the SHF-0189 uses 8V and has higher gain and higher output.

Does anyone have either a M11X1005 Fet, data or a suitable repleacement.

I would assume something having 20dB of gain and +27dBm output.

best regards Peter


Re: Limits of the Port Extension on the Cal screen for calibrating at end of cable

 

As far as I am aware, the port extension feature only adds pure phase. You enter a delay parameter, or a physical length + velocity factor. I don't recall ever seeing any option to dial in any other parameters (e.g., loss), but that may be due to my rarely using port extensions, and then only to shift the reference plane by tiny amounts.

--Tom

--
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
350 Jane Stanford Way
Stanford University
Stanford, CA 94305-4070

On 5/27/2022 18:12, Dave McGuire wrote:
On 5/27/22 19:50, Karin Johnson wrote:
This is probably related to the storage of the cal correction factors in the memory of the 8753C.? Is it via floating point numbers, or integer numbers?
? I don't know about the 8753 in particular, but for the 8510, for example, the calibration correction factors are floating point numbers that are the coefficients for the third-order polynomial that describes the curve (C or L) in the frequency domain for the calibration standard in question.? I'd be very surprised if the 8753 did it any differently.

? Dave Kirkby should be able to chime in on this with more information.

??????????? -Dave


Re: Limits of the Port Extension on the Cal screen for calibrating at end of cable

 

On 5/27/22 19:50, Karin Johnson wrote:
This is probably related to the storage of the cal correction factors in the memory of the 8753C.? Is it via floating point numbers, or integer numbers?
I don't know about the 8753 in particular, but for the 8510, for example, the calibration correction factors are floating point numbers that are the coefficients for the third-order polynomial that describes the curve (C or L) in the frequency domain for the calibration standard in question. I'd be very surprised if the 8753 did it any differently.

Dave Kirkby should be able to chime in on this with more information.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Emails to [email protected] blocked

 

On 5/27/22 20:03, Matt Huszagh wrote:
I'm not really sure where to send this, but I figured some people on
this list might have an idea.
I sent an email to curator@... a few months ago to get access
to some software on their site. I never heard back. I reached out again
just now and received a response that my email had been blocked as junk
mail. I actually don't need the software anymore - found it on the
terrific hp-series300.net, but I'm concerned that my emails are being
flagged. Has anyone experienced this?
There was nothing suspicious about my original email or the follow-up. I
also don't use a graphical email client, so there are no funny images or
attachments in the email; just plain text. This is the blocked response
I got back.
** Message blocked **
Your message to curator@... has been blocked. See technical details below for more information.
The response from the remote server was:
550 "JunkMail rejected - mail-pj1-f53.google.com [209.85.216.53]:42889 is in an RBL: "
David Collins is currently the curator of hpmuseum.net. He's on the VintHPCom mailing list; if you post there he'll see it.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Emails to [email protected] blocked

 

I'm not really sure where to send this, but I figured some people on
this list might have an idea.

I sent an email to curator@... a few months ago to get access
to some software on their site. I never heard back. I reached out again
just now and received a response that my email had been blocked as junk
mail. I actually don't need the software anymore - found it on the
terrific hp-series300.net, but I'm concerned that my emails are being
flagged. Has anyone experienced this?

There was nothing suspicious about my original email or the follow-up. I
also don't use a graphical email client, so there are no funny images or
attachments in the email; just plain text. This is the blocked response
I got back.

** Message blocked **

Your message to curator@... has been blocked. See technical details below for more information.

The response from the remote server was:
550 "JunkMail rejected - mail-pj1-f53.google.com [209.85.216.53]:42889 is in an RBL: "

Matt


Re: Limits of the Port Extension on the Cal screen for calibrating at end of cable

 

I realized a couple of hours after sending the initial post, that I was incorrect with associating Calibration with the Port Extensions.
Here is, I hope a better explanation of what I am trying to do.
First off the measurement frequency is 14 MHz.? So not too beyond the capability of the 8753C.
I want to place the measurement plane at the end of a 100 foot length of RG213.? So I place the SOL standards, one at a time at the
end of the cable.? Supposedly the 8753C measurement plane is now at the end of the cable.
Now I can take S11 measurements relative to the measurement plane at the end of the cable.? My question should have been
" Is there a limit in terms of how long a cable can be before this calibration technique fails."
This is probably related to the storage of the cal correction factors in the memory of the 8753C.? Is it via floating point numbers, or integer numbers?
How large or small can the calibration correction values be???
I seem to be able to obtain valid data using this technique, but wanted to ask on the Keysight forum with some of the folks who actually designed
the instrument.? I think Dr. Joel it still active there.?

My rant about not being able to login to the Keysight forums still exists.? It seems their login/validation screen seems broken.
At one time I had an account over there but they have seemed to have changed the supporting vendor for the forum sites.

As always,
Regards,
Karin Johnson


Re: LCD replacement for vector displays

 

On Fri, May 27, 2022 at 2:56 PM Bill Higdon via <willard561=[email protected]> wrote:
This might open up up some possibilities for some of the vector display instruments

This appears to go in the other direction, e.g. it allows using games emulated with a custom version of the MAME arcade emulator to be displayed on an authentic old-style vector monitor._._,_._,_


Running HP-UX 9 remotely from a Linux host

 

Hello,

I have an HP 382 that I use to control an HP 3048A system. Currently, I
have a disk image containing an HP BASIC 6.3 installation that I serve
to the 382 with HPDrive. This works, but there are two main pain points:

1. Transferring files between the 382 and my modern Linux machine is
cumbersome. HPDir allows me to extract files from a disk image
containing an LIF filesystem. However, my disk image is formatted for
HFS. I could also serve an LIF-formatted drive and copy between them,
but still this is a bit painful (I won't go into the details, but happy
to explain further if someone wants).

2. Using my 382 system requires pulling out a VGA monitor and keyboard
that I only use for this purpose. I'd much prefer to run everything from
my Linux machine.

I expect that if I used HP-UX 9 instead of HP BASIC I could easily
resolve the first problem. HP-UX supports FTP that should make file
transfer over a network simple.

I'm curious about using HP VUE to also resolve the second annoyance. HP
VUE targets the X window system and allows remote connection over a
network. So, in principle this is simple, but with the 30 year time
gap between an implementation of the X window system for HP-UX 9 and the
one running on my current Linux machine, I expect there will be a few
challenges to sort out.

There's an old forum thread that goes into some of the difficulties that
arise. See



and linked threads.

Does anyone use this workflow or something similar? If so, would you be
able to provide any pointers about getting it running? I'm going to
investigate this myself either way, but if someone has already addressed
some of the challenges that pop up, I'd prefer not to duplicate the
troubleshooting time.

Thanks!
Matt


LCD replacement for vector displays

 

This might open up up some possibilities for some of the vector display instruments
https://hackaday.com/2022/05/26/color-vector-display-controller-brings-arcade-classics-back-to-life/


Re: checking line module setting for 110V / 220V

Lothar baier
 

¿ªÔÆÌåÓý

I used a little hook like the ones you can buy at a hardware store and glued it into a small plastic rod

?

From: [email protected] <[email protected]> On Behalf Of ??????? ????? via groups.io
Sent: Friday, May 27, 2022 1:29 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] checking line module setting for 110V / 220V

?

there is small hole on card
use a hook and pull, swinging the card from side to side

There is no locking except fuse lever