Never mind.. I found the part number. ?It's a B-1493, which doesn't show up anywhere but in the NSN system. ?From the characteristics given, the NTE121 or NTE179 should be a good sub. The B-1493 is a Germanium PNP, BVcbo = 60V Pd = 90W Iv = 7A Hfe = Unknown
Cheers, DaveM
|
The zener diodes seem fairly easy to replace. ?A quick search on Mouser's web site comes up with the following as suitable replacements: CR6 = MTZJ39SB R0G, a 36.3V 500mW Zener 2% ?$0.22 each CR8 = TZX6V2C-TR, a 6.2V 500mW 2% ?$0.19 each
The power transistor isn't a 2N3106. ?From what I can find, the 2N3106 is a large stud cased SCR. ?What's the HP part number for the transistor?
Cheers, Dave M
|
I'm in the process of downloading it (finally).
Looks like I should get the whole thing. Now I need to get a spare drive and burn it.
One problem: this scope only has a 44-pin adapter (3.5" to 2.5") and the drive only has 44-pins. Power comes in on a breakout connector on the adapter which a tiny connector (that doesn't mate well) provides 5V and ground.
A standard SSD has a SATA connector (?), so I need I need to get a drive that's compatible with the connectors in this scope.
|
I have a few 410Cs. I've restuffed the caps in a couple of them. For one of them, I designed a small PC board that holds modern caps and ran wires down to an adapter plate with solder posts. That worked but it was rather kludgy. I keep looking for a better solution for these. Maybe Hayseed Hamfest could be talked into making some replacements.
Good luck with it.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: driftwoodturning@... To: [email protected] Sent: Sunday, October 11, 2020 6:33:10 PM Subject: [HP-Agilent-Keysight-equipment] Hp-410c parts?
I have three Hp-410c meters. I bought the Ss boards from Ashley (kiss electronics) for one of them. I would like to replace ALL of the electrolytic capacitors, the two diodes on the a7 board, A7CR6 (1902-0026) and A7CR8 (1902-0049), maybe even replace the power transistor 2N3106 (I think that’s the number). Has anyone already done the work to find modern, current part numbers for these units? I am having little to no luck finding them. I would appreciate standing on your shoulders.
Thanks in advance, Jeff
|
I have three Hp-410c meters. I bought the Ss boards from Ashley (kiss electronics) for one of them. I would like to replace ALL of the electrolytic capacitors, the two diodes on the a7 board, A7CR6 (1902-0026) and A7CR8 (1902-0049), maybe even replace the power transistor 2N3106 (I think that’s the number). Has anyone already done the work to find modern, current part numbers for these units? I am having little to no luck finding them. I would appreciate standing on your shoulders.
Thanks in advance, Jeff
|
Re: Clarification : Brass gears for the 8640B
Many thanks to you Jeff, you have certainly saved me lots of trouble when explaining to my buyers how to install the gears :-P. (No of course they are not molded -- some people love calling them names as I've seen on multiple occasions but I rather enjoy it :-O) Jon - my gear sets are available for purchase off ePray. Just search for "Brass gears 8640B" and you'll hit my listing. Good luck with your 8640B restoration! Cheers - Ram
|
Hello,
just cackle, if I can be of use!
Tam
With best regards
Tam HANNA
Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at
toggle quoted message
Show quoted text
Don't worry, I save way more time with this group that I can ever
give back, so it's really my pleasure.?
I'll upload it an other way (not zipped, so everything can be
download once at a time). I'll post the link when it's ready!
|
Re: HP8656B attenuator input and output connectors
Hi Mark
The two attenuator modules in a HP 8656B are not the standard HP rectangular ones, they use a deep drawn metal top hat soldered or spot welded onto a metal base, very possibly they are irreparable.
I hate to admit this but I have almost sixty years experience in the electronics/RF world, there are not many connectors that I cannot identify and have adapters to and from.
Hope you are keeping well and avoiding the Covid virus, stay safe
George G6HIG Dover UK On Sunday, 11 October 2020, 01:42:35 BST, Mark Bielman <mbielman@...> wrote:
Hi George,
Good point. I work in the RF world so...
The units I have dealt with are (typically) both female SMA's on the body of the unit (rectangular shape) with the ribbon cable connectors for the solenoids.
Mark
|
Re: HP8656B attenuator input and output connectors
Hi George,
Good point. I work in the RF world so...
The units I have dealt with are (typically) both female SMA's on the body of the unit (rectangular shape) with the ribbon cable connectors for the solenoids.
Mark
|
Re: HP8656B attenuator input and output connectors
Hi Mark
I actually wonder at just how few know the difference between an SMA, SMB and SMC connector. There is at least one HP signal generator that uses an SMC on the attenuator input and an SMA on the output, but it is arguable if it is from the same era.
George G6HIG Dover UK On Sunday, 11 October 2020, 00:37:16 BST, Mark Bielman <mbielman@...> wrote:
All the HP attenuators from this era that I have seen use SMA connectors.
Mark
|
Re: HP8656B attenuator input and output connectors
The front connector required is definitely an SMA MALE the rear connector is definitely an SMA FEMALE. Direct adapters to BNC can be had on cheaply eBay, they are used for some Beofang Hand portables.
George G6HIG Dover UK On Sunday, 11 October 2020, 00:16:03 BST, Syd via groups.io <nhuq1@...> wrote:
In this unit the attenuator has input and output RF cables connected to it.. I would like to make a couple of test cables using RG316 cable with a BNC MALE on one end of both cables and the other end of the two cables with male and female connectors to measure the RF at the input and the outputs of the attenuator. I thought that these fittings, to and from the attenuator, were SMC, MALE? and FEMALE, but I was wrong!? What kind of MALE and FEMALE connectors do I need to connect to these cables and sockets to/from the attenuator? syd/wt1v
|
Re: HP8656B attenuator input and output connectors
All the HP attenuators from this era that I have seen use SMA connectors.
Mark
|
HP8656B attenuator input and output connectors
In this unit the attenuator has input and output RF cables connected to it.. I would like to make a couple of test cables using RG316 cable with a BNC MALE on one end of both cables and the other end of the two cables with male and female connectors to measure the RF at the input and the outputs of the attenuator. I thought that these fittings, to and from the attenuator, were SMC, MALE? and FEMALE, but I was wrong!? What kind of MALE and FEMALE connectors do I need to connect to these cables and sockets to/from the attenuator? syd/wt1v
|
Re: HP435 measuring head? kW's?
True, if you want to use the switch, but the 11683A-H01, which is required for later meters can?use?a voltage input which bypasses the switches.
The schematic here:
Shows that when using the external reference, the range switch isn't used at all.
The home-brew 11683A works as a 11683A-H01 with the REF SELECT switch set to EXT as in the procedure here on pages 42-46:
So - even the expensive switch isn't good enough in some circumstances!
Orin.
toggle quoted message
Show quoted text
On Sat, Oct 10, 2020 at 10:49 AM Dave Smith W6TE < w6te@...> wrote:
Building a 11683A isn’t as simple as using an old 8481A and a few precision resistors. The range switch in the HP 11683A is not your run-of-the-mill SP 10 pos rotary switch. HP used a very low resistance contacts in the switch. In fact,
there’s a procedure in the 11683A’s manual for checking the integrity of the switch. The switch has an in house part number and Keysight lists the switch at $3.000!
?
Gud luk
?
Dave
?
?
On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes <tholmes@...> wrote:
By the way, what is involved in calibrating a power meter, not the sensor?
An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.
11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):
Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.
I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.
Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.
|
Re: HP435 measuring head? kW's?
Keysight is very PROUD of that switch!
?
Dave
?
toggle quoted message
Show quoted text
From: [email protected] < [email protected]>
On Behalf Of Steve - Home
Sent: Saturday, October 10, 2020 11:18 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP435 measuring head? kW's?
?
At that price for the switch, my 11683A probably needs to be in the locked cabinet!
On Oct 10, 2020, at 12:46 PM, Dave Smith W6TE <w6te@...> wrote:
?
Building a 11683A isn’t as simple as using an old 8481A and a few precision resistors. The range switch in the HP 11683A is not your run-of-the-mill SP 10 pos rotary switch. HP used a very low resistance contacts in the switch. In fact,
there’s a procedure in the 11683A’s manual for checking the integrity of the switch. The switch has an in house part number and Keysight lists the switch at $3.000!
?
Gud luk
?
Dave
?
?
On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes <tholmes@...> wrote:
By the way, what is involved in calibrating a power meter, not the sensor?
An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.
11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):
Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.
I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.
Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.
|
Re: HP435 measuring head? kW's?
At that price for the switch, my 11683A probably needs to be in the locked cabinet!
toggle quoted message
Show quoted text
On Oct 10, 2020, at 12:46 PM, Dave Smith W6TE <w6te@...> wrote:
?
Building a 11683A isn’t as simple as using an old 8481A and a few precision resistors. The range switch in the HP 11683A is not your run-of-the-mill SP 10 pos rotary switch. HP used a very low resistance contacts in the switch. In fact,
there’s a procedure in the 11683A’s manual for checking the integrity of the switch. The switch has an in house part number and Keysight lists the switch at $3.000!
?
Gud luk
?
Dave
?
?
On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes <tholmes@...> wrote:
By the way, what is involved in calibrating a power meter, not the sensor?
An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.
11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):
Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.
I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.
Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.
|
Re: HP435 measuring head?
The good home had the system operational for a few years before it was sent to the bad place...
Steve WB0DBS
toggle quoted message
Show quoted text
On Oct 10, 2020, at 11:40 AM, Dave McGuire <mcguire@...> wrote:
? Ugh. That doesn't sound like a "good home" to me.
-Dave
On October 10, 2020 9:21:16 AM "Steve - Home" <steve-krull@...> wrote: I wish I still had it but it went to a good home when money was tight. I’ll check but I’m afraid it was probably crushed for scrap.
Steve WB0DBS
On Oct 9, 2020, at 10:06 PM, Dave McGuire <mcguire@...> wrote: ?On 10/9/20 10:52 PM, Steve - Home wrote:
Not a Cray but I had an Intel iPSC sitting in my living room for quite a while. Got a lot of attention from Intel and others. We would SO give that a home at LSSM, if you still have it, or know where it is.
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: Hp 8563E analyzer for sale refurbished price 1725Euro
Il 10/10/2020 1.50, Peter Hansen ha
scritto:
Hello Diego the unit is sold but I get them from time to
time as I do refurbish them I can keep you in mind if you like
best regards Peter
Dear
I have sent you a PM on your address
oz1lpr@...
BR
Diego Gustin
Dear
?yes please, when you have a 8563e please
send me a mail.
BR
?diego gustin
|
Re: HP435 measuring head? kW's?
Building a 11683A isn’t as simple as using an old 8481A and a few precision resistors. The range switch in the HP 11683A is not your run-of-the-mill SP 10 pos rotary switch. HP used a very low resistance contacts in the switch. In fact,
there’s a procedure in the 11683A’s manual for checking the integrity of the switch. The switch has an in house part number and Keysight lists the switch at $3.000!
?
Gud luk
?
Dave
?
toggle quoted message
Show quoted text
From: [email protected] < [email protected]>
On Behalf Of Orin Eman
Sent: Saturday, October 10, 2020 10:35 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP435 measuring head? kW's?
?
On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes <tholmes@...> wrote:
By the way, what is involved in calibrating a power meter, not the sensor?
An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.
11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):
Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.
I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.
Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.
|
Re: HP435 measuring head? kW's?
On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes < tholmes@...> wrote: By the way, what is involved in calibrating a power meter, not the sensor?
An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.
11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):
Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.
I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.
Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.
Orin.
|