Re: HP435 measuring head? kW's?


 

¿ªÔÆÌåÓý

Keysight is very PROUD of that switch!

?

Dave

?

From: [email protected] <[email protected]> On Behalf Of Steve - Home
Sent: Saturday, October 10, 2020 11:18 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP435 measuring head? kW's?

?

At that price for the switch, my 11683A probably needs to be in the locked cabinet!

Steve

WB0DBS

?

?



On Oct 10, 2020, at 12:46 PM, Dave Smith W6TE <w6te@...> wrote:

?

Building a 11683A isn¡¯t as simple as using an old 8481A and a few precision resistors. The range switch in the HP 11683A is not your run-of-the-mill SP 10 pos rotary switch. HP used a very low resistance contacts in the switch. In fact, there¡¯s a procedure in the 11683A¡¯s manual for checking the integrity of the switch. The switch has an in house part number and Keysight lists the switch at $3.000!

?

Gud luk

?

Dave

?

From: [email protected] <[email protected]> On Behalf Of Orin Eman
Sent: Saturday, October 10, 2020 10:35 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP435 measuring head? kW's?

?

On Sat, Oct 10, 2020 at 7:07 AM Tom Holmes <tholmes@...> wrote:

By the way, what is involved in calibrating a power meter, not the sensor?

?

An 11683A, 432A plus 478A-H75 sensor, a counter and a DMM.

?

11683A is rare, but you can make one from a dead 8481A (and there are plenty of those around):

?

?

Then you just need a stable voltage source in the range of a few hundred uV to 15.8 V.

?

I made one and using a Fluke 343A as a voltage source and 34461A to measure the voltage, my 346A gave readings well within specs.

?

Calibrating the power reference is difficult... it requires the H75 option for the 478A sensor (good luck finding that) and a somewhat complicated procedure involving measuring voltages from the 432A.

?

Orin.

?

Join [email protected] to automatically receive all group messages.