CHECK OUT THE WIKI The purpose of the wiki is mainly to allow you to find information on instruments, either from either
- The model number
- The function(s) listed below. Some instruments have multiple functions - for example, the 4195A is a VNA, spectrum analyzer and an impedance analyzer. Therefore the 4195A is listed in multiple categories
Please also check out HPWiki available here:
- Accessory kits - various types
- AC power analyzers - PA2201A and PA2203A
- AC power supplies 6811C, 6812C, 6813C
- Airlines
- Arbitrary waveform generators M8194A
- Amplifiers?493A, 495A?
- Attenuators (optical) 8156A, 8157A, 8158B, 81566A, 81576A,?
- Attenuators (RF) 8494A
- Attenuator set (500 ¦¸) 350C
- Attenuator set (600 ¦¸) 350D
- Attenuator switch driver
- Audio analyzers? 8903A, 8903B, 8903E,? ?
- Base station test sets
- Bit error rate testers (BERTs)
- Cables
- Capacitance meters U1701A, U1701B, 4272A, 4278A, 4279A
- Capacitor Bridge 4270A,
- Capacitor standards 16380A, 16380C,?
- Carrier noise test setsi
- Cesium frequency standards
- Clamp ammeters
- Close field probes
- Crystal Impedance E4915A, E4916A
- Data Acquisition Systems (DAQs)
- DC power analyzers
- DC power supplies 6030A , 6031A , 6032A, 6033A, 6035A, 6131C, 6621A, 6622A, 6623A, 6624A, 6627A, 6255A, 6645A, 6671A, 6672A, 6673A, 6674A, 6675A, 62003A, 62003C, 62003E, 62004A, 62004B, 62004E, 62005A, 62005B, 62005E, 62006A, 62006B, 62006E, 62010A, 62010C, 62010E, 62012A, 62012C, 62012E, 62015A, 62015C, 62015E, 62018A, 62018C, 62018E, 62024A, 62024C, 62024E, 62028A, 62028C, 62028E, 62048A, 62048C, 62048E
- Delay lines
- Detectors
- Device current waveform analyzers
- Digital communications analyzers
- Directional couplers
- Distortion analyzers 330B, 330C, 330D, 331A, 332A, 333A, 334A, 339A, 8903A, 8903B, 8903E,???
- Dynamic measurement DC source
- Electrometers
- Fading simulators
- Femto ammeters
- Filters
- Frequency counters 522B, 5342A 5343A 5352B
- Frequency standards?
- Function Generators ? 3310A,? 8165A,
- GPIB controllers, extenders, cables etc.
- GPS frequency standards
- Harmonic mixers
- High resistance meters 4339B
- High resistance meter fixtures 16008B
- HEV EV Grid Emulators and Test Systems
- In-circuit test systems
- Impedance analyzers 4195A, 4291A, 4291B, 4395A, 4396A, 4396B, 4294A, E4990A, E4991A
- Impedance Analyzer Accessories
- Impedance / Gain Phase analyzer 4194A
- Impedance Meter 4193A,
- Isolators
- LCR meters? U1701A, U1701B, U1731A,? U1731B, U1731C, U1732A, U1732B, U1732C, U1733C, 4191A , 4192A, 4194A, 4195A, E4196A,? 4216A, 4260A, 4261A, 4262A? 4263A, 4263B, 4271B, 4274A, 4275A, 4276A , 4277A, 4284A, 4285A, 4286A, 4287A, 4291A, 4291B, 4294A, 4332A, 4342A, 4395A, 4396A, 4396B, E4980A and E4980AL
- LCR meter calibration devices? 16380A 42030A? 42090A, 42091A and 42100A
- LCR meter accessories
- 2-Terminal BNCs.
- 4-Terminal Pair (BNC connectors)
- Cable extension 16048A, 16048D, 16048E, 16048G, 16048H
- DC current bias accessories 42841A, 42842A, 42842B, 42842C, 42843A
- DC voltage bias accessories 16065A, 16065C,
- Kelvin clips 16089A, 16089B, 16089C,16089E
- Lead Components 16047A,16047B, 16047D, 16047E
- Material 16451B, 16452A
- Probes 42941A
- SMD 16034E, 16034G, 16034H
- 2-port 16096A
- 7 mm (APC7)
- 2-Terminal BNCs.
- LCZ meters? 4276A, 4277A,
- Lightwave clock / data receivers
- Lightwave converter
- Lightwave component analyzer
- Lightwave measurement system mainframes
- Lightwave polarization analyzers 8509B
- Logic analyzers
- Nemo wireless network solutions.
- Noise and interference test set
- Noise figure analyzers
- Noise sources 346A, 346B. 346C ,
- Matching pads (50 ohm to 75 ohm or similar)
- Materials test equipment
- Microwave repeaters
- Microwave downconverters 70427A
- Microwave / THz sources
- Milliammeter 428B
- Milliohm meter
- Mobile communications DC source
- Modular instruments
- AXIe
- Data acquisition (DAQ)
- USB
- PXIe
- Modulation analyzers
- Multimeters 427A, 970A
- Optical attenuators
- Optical heads
- Optical sources
- Optical spectrum analyzers
- Oscilloscopes 120A, 120AR, 120B, 122A, 130A, 130B, 130BR, 130C, 140A, 140B, 141A, 150A, 150AR, 160B, 180A, 180AR, 180CD, 181A, 181AR, 181T, 181TR, 182C, 182T, 183A, 183B, 184A, 184B, 185A, 185B, 1200A, 1200B, 1220A, 1221A, 1703A, 1707A, 1707B, 1710A, 1710B, 1715A, 1722A, 1725A, 1726A, 1740A, 1741A, 1742A, 1743A, 1744A, 1746A, 1980A, 1980B, 5403A, 6000A, 6000L, 16533A, 16534A, 54100A, 5410B, 54100C, 5100D, 54111D, 54120A, 54120B, 54200A, 54501A, 54502A, 54503A, 54504A, 54520A, 54520C, 54540A, 54540C, 54542A, 54542C, 54600B, 54601A, 54601B, 54602B, 54603B,? 54645A, 54654N, 54710A, 54720A, 54750A, 54825N, E1428,?
- Oven controlled crystal oscillators (OCXOs)
- Pattern generators
- PCM terminal test set
- Phase noise measurement
- Pico ammeters
- Printers 2225
- Plotters 7470A, 7475A?
- Probes
- Protocol analyzers and exercisers.
- Power booster test sets
- Power meters 431A, 431B, 431C, 432A, 435A, 435B, 437B, 438A
- Power splitters
- Power supplies
- Pulse generators
- Q-meters 4342A?
- Q-meter calibration inductors 16470A
- Reflection transmission test set
- Return loss module (optical)
- Relays / switches / switch matrices (optical)
- Relays / switches / switch matrices (RF)
- Resistor standards 42030A?and 42100A
- S-parameter test sets
- Scalar network analyzers
- SCSI bus preprocessor interface E2324A
- Selective level meters 3746A
- Semiconductors
- Semiconductor parameter analyzers 4145A, 4155B, 4156B,
- Signal analyzers
- Signal generators / sweep generators / signal sources / oscillators 200CD, 201B, 209A, 204D,? 608A,? 8165A
- Software
- Source measure units
- Spectrum analyzers 4195A,???
- Switch control units
- SWR meter 415E?
- Time interval? counters
- Time mark generator 226A
- Timing and data state modules
- Torque wrenches
- Transmitter testers
- Trigger modules
- Ultrasound transducers
- Universal bridge? 4260A, 4265A, 4265B?
- Vacuum tube voltmeter 410C
- Vector Impedance Meter 4193A, 4800A, 4815A
- Vector Network Analyzers (VNAs) 4195A,? 8510A, 8510B, 8510C, 8753A, 8753B, 8753C, 8753D, 8753E, 8753ES, 8752ET, 8719A, 8719B, 8719C, 8719D, 8720A, 8720B, 8720C, 8720D, 8720ES, 8722A, 8722B, 8722C, 8722D, 8722ES,
- Vector Network Analyzers (VNA) calibration kits 85032B, 85032E, 85033C, 85033D, 85033E, 85050B, 85050C, 85050D, 85052B, 85052C, 85052D, 85054A, 85054B, 85054D, 85056A
- Vector Network Analyzer (VNA) verification kits
- Vector Signal Analyzer 89650S, 89600S
- Vector voltmeters 8405A, 8508A,
- VXI mainframes 70000B, 70000C
- Waveform and function generators
- Waveguide to waveguide and waveguide to coaxial transitions.
- Wireless 58 OTA chambers
- Wireless channel emulators
- Wireless network emulators
- Wireless communication test sets
?
OT: Looking for recommendations on 3D modeling software
This is off topic but the end result will be for HP products so please bear with me.
I am looking for recommendation on 3D modeling software (preferably freeware) and a company that I can then send the 3D model to? make a battery holder clamshells.? I propose to make half clamshell pieces that would then make a complete? battery holder by using 2 pieces that would fit together. Thank you, Victor |
Re: looking for a few knobs
¿ªÔÆÌåÓýIf you have any questions about 3D printers, feel free to contact me, also offlist!
We are quite happy with our RenkForce RF100v2 system after adding
a part cooler and an enclosure. Enclosures are a feature where the
printer house generally like to charge a lot of money, but you can
build one easily from a LACK table =>
I am quite happy with my design, albeit I regret not also making
the back openable to access the filament supply. Not having the
printer subject to winding, and being able to heat up its den, has
made my life easier and prints more constant. I think about adding
a little thermo heater to heat the build chamber up faster.
Another important advice: heated beds are a bad bet. Get a Pritt
gluestick...
Tam - - - - - With best regards Tam HANNA Enjoy electronics? Join 13500 followers by visiting the Crazy Electronics Lab at On 07.09.20 04:40, Dave McGuire wrote:
On 9/6/20 8:19 PM, Michael A. Terrell wrote: |
Re: looking for a few knobs
Sorry. I clicked on 'reply to sender', not to the group. Another reply to the thread came back to me, rather than the person I replied to. On Mon, Sep 7, 2020 at 8:21 AM Michael A. Terrell via <terrell.michael.a=[email protected]> wrote:
|
Re: looking for a few knobs
I received the files for the Bezels. Thanks. Most of this is work that would have needed to be done, and with the way things are right now, he may lose the house so he isn't far from being ruined. He's had a series of jobs. He seems to keep getting hired not long before a company or location closes. The big problem is that the previous owner was too damned cheap to do anything right. All she did was bitch about all the money that her dead husband had spent to have the home remodeled. He worked for the school board, and it's obvious that some of what was used came out of schools when repairs were made. Two breaker boxes have burnt, in the last 21 years. One would have burnt the place down, if I hadn't heard the noise as the main breaker burnt off of the buss bars. The other was on a pole, in the yard, but it burnt during the night, and killed everything but the shop building. My main complaint is the difficulty of doing the work with one good hand. Old age isn't for sissies! On Sun, Sep 6, 2020 at 10:40 PM Dave McGuire <mcguire@...> wrote: On 9/6/20 8:19 PM, Michael A. Terrell wrote: |
Re: HP 1817A TDR sampling head
¿ªÔÆÌåÓýHello Craig, ???? Glad to find a guy who knows about 1817A sampler. This state of the art equipment (for then time it was marketed) has fallen in the museum class but works perfectly when the rather fragile sampler diodes are in good condition. Your suggestion to use a beam lead microwave schottky diode is one of the solutions I envisaged to fix my 1817's. I also tried some very tiny microwave SMD diodes (0.3 x 0.6 mm), which are supposed to be very fast (SMS 7621-060 from Skyworks).? If I have the binocular microscope, I don't unfortunately possess the micro tweezers (clamps) necessary to manipulate the diodes and put them at the right place on the holder, and my attemps to getb them in place? and to fix them with soldering paste have failed up to now. May be beam lead diodes would be easier to handle ! ??? Another solution I'll try will be to machine an new holder, fitting in place of the original one, and using an old DO34 housed Germanium point diode, which also presents very small capacitance and offset voltage. It is slightly bigger than the genuine one, but that seems workable. I recall you that the assembly is composed of the diode, a series 2.2 K resistor, and a 2 pf capacitor towards 7mm line body. The new part would fit, but electrical performances (rise time, overshoot) might be degraded ! ??? Thank You anyway to help me renewing one of the most advanced equipment ever marketed by former HP firm. (joined two drawings of one of the sampling diode part - there are 2 opposed polarity diodes ). ???? Jean-claude
?On 07/09/2020 09:19, Craig Sawyers
wrote:
|
Re: HP 1817A TDR sampling head
¿ªÔÆÌåÓýI was thinking in terms of a beam lead microwave diode ¨C but held back from suggesting it given the difficulty in handling, mounting and soldering/conductive epoxy/lead bonding such physically tiny devices. But given the bandwidth of the 1817A sampler, that would seem to be the only easily procurable (and cheap!) part family. And they have the correct sort of capacitance (0.2pF) and inductance (<250pH) needed. Just don¡¯t sneeze when handling them ¨C and a good stereo microscope is essential. ? Craig ? From: [email protected] [mailto:[email protected]] On Behalf Of Mark Kahrs
Sent: 07 September 2020 03:56 To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP 1817A TDR sampling head ? Salut Jean-Claude: ? The 1817A sampler contains the incredible Grove sampler (see the US Patent, US 3,278,763 for the associated patentese, also see the Proc. IEEE paper by Grove for the theory of operation and detailed analysis).? IMHO, the Grove sampler was the last hurrah for non-monolithic microwave sampler construction.? Since this sampler was found in many HP instruments of the era, you can find repair instructions in the 1430C manual (and others).? Basically, the diode is a pill that is shimmed into contact with the stripline.? To replace this with a modern diode, you're going to have to fabricate a matching mechanical contrivance.? I suggest (without having done this), that you might be able to "print" a suitable diode holder and then insert a replacement diode (it would be tempting to use a beam-lead diode or find a suitable pill diode). ? For extra homework, I suggest replicating Grove's analysis in the Proceedings. ? Bonne Chance! ? ? On Wed, Sep 2, 2020 at 1:09 PM JCB_SFR <jean-claude.bernengo@...> wrote:
|
Re: HP 1817A TDR sampling head
Salut Jean-Claude: The 1817A sampler contains the incredible Grove sampler (see the US Patent, US 3,278,763 for the associated patentese, also see the Proc. IEEE paper by Grove for the theory of operation and detailed analysis).? IMHO, the Grove sampler was the last hurrah for non-monolithic microwave sampler construction.? Since this sampler was found in many HP instruments of the era, you can find repair instructions in the 1430C manual (and others).? Basically, the diode is a pill that is shimmed into contact with the stripline.? To replace this with a modern diode, you're going to have to fabricate a matching mechanical contrivance.? I suggest (without having done this), that you might be able to "print" a suitable diode holder and then insert a replacement diode (it would be tempting to use a beam-lead diode or find a suitable pill diode). For extra homework, I suggest replicating Grove's analysis in the Proceedings. Bonne Chance! On Wed, Sep 2, 2020 at 1:09 PM JCB_SFR <jean-claude.bernengo@...> wrote: Hello, |
Re: looking for a few knobs
On 9/6/20 8:19 PM, Michael A. Terrell wrote:
I don't have a printer yet, but I would appreciate the files for when I do get one.I will send them off-list. I own the house. The damage was caused by a tree that a neighbor won't trim, or remove. The drop was installed in 1964. I was already planning to replace all the old wiring, but not all at once. My old Dell optiplex 780 and Insignia monitor were still working with the line at 67 volts. Duke Energy cancelled the first service? call, because their idiot 'Smart Meter' reported that my service was working. I had to threaten to call the Florida PUCO before they sent out a crew. The lead tech rolled his eyes when I told him the neural was open. Then he saw the break, arching. Tired old triplex that had pulled loose from the street pole and was drooping below six feet. I should have called OSHA, because it was over a neighbor's yard where kids were playing. I thought Florida power, and then Progress Energy were bad, but Duke is even worse. I am going to install underground service, but I would like to have waited until they do a promised rebuild. The entire subdivision was wired over 50 years ago and it was not designed for electric stoves and air conditioning. It needs to be upgraded to at least four times the current capacity. It was designed for 60A service, and for snowbirds second homes. I have a 1200 square foot building for my workshop. It was already upgraded to a 100A breaker box but I am currently unable to work there.Wow. That is a substantial lawsuit waiting to happen. That neighbor needs to be *ruined*. -Dave -- Dave McGuire, AK4HZ New Kensington, PA |
Re: Agilent E1430A and E1485A driver software
On 6/09/20 10:07 pm, spam@... wrote:
Hi,I've just uploaded my SCPI driver collection to /g/vxi/ It has the serial uploader for E1430A as well as the GPIB version. For E1485A I found neither manual nor software.I've got the install/service guide pdf for the E1485C. Plus notes etc on upgrading the SPOS on that module if you don't have that already. There was a HP 35635T?Programming Toolkit?sold on tape with these cards in the old days to be used on e.g. HP V743 HP/UX and Windows.I managed to find a PDF of a Powerpoint presentation regarding this product but nothing else so far. Any hints where I could get this old software (HP 35635T?Programming Toolkit ) or parts of it today?I would give my left testicle... We've been having discussions recently on the VintHPcom group regarding having members check through their boxes of old floppies/HDD/s/tapes to look for software to try and get as much as possible archived in safe locations for future use. We'd be eternally grateful if anyone on this group that has such software could archive it as well. My immediate interest evolves around E1497A/B E1437A E1485B/D E6401A E6402A and WJ-9119 :^) |
Re: looking for a few knobs
Hi, Michael. In my area of Southern California, we had a similar problem with mid 1960's power wiring.? We kept having outages in our neighborhood every month or so, and San Diego Gas & Electric kept band-aiding the old wiring.? Until I posted a formal complaint to the Public Utilities Commission, squeaky wheel kind of guy that I am.? That lit a fire under their chair and got us new wiring within a few months.? That was probably 4 years ago, and we haven't had an outage since. So go ahead and file a complaint.? It worked for me and my neighbors.? Good luck! Jim Ford Laguna Hills, CA
------ Original Message ------
From: "Michael A. Terrell" <terrell.michael.a@...>
Sent: 9/6/2020 5:19:35 PM
Subject: Re: [HP-Agilent-Keysight-equipment] looking for a few knobs
|
Re: looking for a few knobs
I don't have a printer yet, but I would appreciate the files for when I do get one. I own the house. The damage was caused by a tree that a neighbor won't trim, or remove. The drop was installed in 1964. I was already planning to replace all the old wiring, but not all at once. My old Dell optiplex 780 and Insignia monitor were still working with the line at 67 volts. Duke Energy cancelled the first service? call, because their idiot 'Smart Meter' reported that my service was working. I had to threaten to call the Florida PUCO before they sent out a crew. The lead tech rolled his eyes when I told him the neural was open. Then he saw the break, arching. Tired old triplex that had pulled loose from the street pole and was drooping below six feet. I should have called OSHA, because it was over a neighbor's yard where kids were playing. I thought Florida power, and then Progress Energy were bad, but Duke is even worse. I am going to install underground service, but I would like to have waited until they do a promised rebuild. The entire subdivision was wired over 50 years ago and it was not designed for electric stoves and air conditioning. It needs to be upgraded to at least four times the current capacity. It was designed for 60A service, and for snowbirds second homes. I have a 1200 square foot building for my workshop. It was already upgraded to a 100A breaker box but I am currently unable to work there. On Sun, Sep 6, 2020 at 7:23 PM Dave McGuire <mcguire@...> wrote: On 9/6/20 7:16 PM, Michael A. Terrell wrote: |
Re: looking for a few knobs
On 9/6/20 7:16 PM, Michael A. Terrell wrote:
Is that the same size as the knobs on the HP 6236Bpower supplies? I am missing knobs and the inserts on several of these.I think these are smaller. Also, has anyone made replacements for the meter bezels for these?I've designed a replacement bezel for the 6205B, and 3D printed it, and it works well. I think the 6236B's meters might be the same size. Let me know if you want the files. I can't access my manuals right now, I am still working to complete repairs after I lost the neutral from the street and fried a lot of equipment. I currently have lights and a window A/C in one room, or I would look .up the part numbers. Also, I ended up with limited data on Hughesnet, after Spectrum refused to restor my service. The open neutral fried their coax drop when it tried to carry the Neutral current. I heard burning MOVs and tripping breakers before the house went dark. It looks like several thousand dollars worth of damage to the wiring and equipment. I was off line for about two months.Ugh!! I recommend calling lawyers and packing to move. -Dave -- Dave McGuire, AK4HZ New Kensington, PA |
Re: looking for a few knobs
Is that the same size as the knobs on the
HP 6236Bpower supplies? I am missing knobs and the inserts on several of these. Also, has anyone made replacements for the meter bezels for these? I can't access my manuals right now, I am still working to complete repairs after I lost the neutral from the street and fried a lot of equipment. I currently have lights and a window A/C in one room, or I would look .up the part numbers. Also, I ended up with limited data on Hughesnet, after Spectrum refused to restor my service. The open neutral fried their coax drop when it tried to carry the Neutral current. I heard burning MOVs and tripping breakers before the house went dark. It looks like several thousand dollars worth of damage to the wiring and equipment. I was off line for about two months. On Sun, Sep 6, 2020 at 11:17 AM Dave McGuire <mcguire@...> wrote:
|
Re: looking for a few knobs
I think you have been reading my mind - I was thinking of posting a
toggle quoted message
Show quoted text
similar statement. THANKS Quoting RFI-EMI-GUY <rhyolite@...>: I agree, Profanity is the crutch of the inarticulate motherfr |
Re: looking for a few knobs
Yup it gets old. I thought maybe this guy Tam had Tourettes and dismissed him as simply having a handicap,? but, is Tourettes actually a thing when using a keyboard? Now it seems that it is a political expression of being non-PC and anti liberal. Is being a nice person suddenly a "Liberal" problem in this world? Is there an ignore button on this thing?
toggle quoted message
Show quoted text
On Sun, Sep 6, 2020 at 11:48 AM, Siggi wrote:
|
Re: Agilent 54831M
Overall, the scope won't calibrate, but the two healthy channels will look calibrated after the procedure fails. The two bad channels will not work properly on the 1Meg input impedance after the failed calibration.
The failed 50ohm inputs are 99.9% due to bad input hybrids, it doesn't take much of DC to blow the input (even the 50ohm resistor is integrated). It appears there is a chinese guy on EBay, selling these hybrids (expect a chunk of PCB with the chip soldered? on it). I did not buy any and don't know anybody that did, so the risk is all yours if you decide to go for.? Or you may want to hunt for another 54831 as a part mule, this is what I did and now have spares for the rest of my life. |
Re: looking for a few knobs
I agree, Profanity is the crutch of the inarticulate motherfr
toggle quoted message
Show quoted text
On Sun, Sep 6, 2020 at 12:48 AM, Dave Seiter wrote:
|