Re: HP 8753D Connection Adapters?
On Mon, 7 Feb 2022 at 08:08, James Sawle (MD0MDI) <james@...> wrote: Thanks everyone, I must have spent ages just looking at the ends and wondering why the leads would not connect to the unit and why I had not been supplied the connection adapters! Opps! I did not see the start of this thread, and don't have time to look back. ? I have had customers report problems with Chinese ones above 3 GHz, and others say they are okay - I have never measured any myself.? If I had a 6 GHz test set, and intended using at 6 GHz, I would probably be tempted to buy the Keysight ones, despite the much higher price. However, the VNA cal kit should take out the errors introduced by the adapters, but if there's a considerable loss, as some have reported to me, it will degrade your performance. There's quite a bit of information on the 8753 here Dave |
Re: Adjustment procedure for 355C VHF Attenuator?
is the nylon glide, but the price is quite high, along with the screws. The resistors are all obsolete, but you might find some on Ebay or other sources. I would see if some phillips head nylon screws to replace the glides. According to the Keysight part description, the hole is 0.19" which would be a #10 screw. I buy hardware from Fastenal, or Grainger but there are other sources. may work, and they are much cheaper than $5.05 each from Keysight. is another option, if you don't care about the color. 50/$7.80
On Sun, Feb 6, 2022 at 7:22 AM Jared Cabot via <jaredcabot=[email protected]> wrote: Thanks for the info so far! |
Re: HP 1858-0015
开云体育Given the performance required, I suspect that a garden-variety transistor array (e.g., LM3046, etc.) might suffice electrically. I have no idea how much room you have to work with, but if you have sufficient space, you might be able to shoe-horn one of those more commonly available arrays, appropriately cross-wired.-- Cheers, Tom -- Prof. Thomas H. Lee Allen Ctr., Rm. 205 350 Jane Stanford Way Stanford University Stanford, CA 94305-4070 On 2/6/2022 02:06, Istvan Kovacs wrote:
Hello, |
Re: Cascaded amplifier for 8673D Watkins Johnson 6203-044 (HP 0955-0215 )
开云体育Thanks for the answer, do you have the information for the flatness? My amp must be defective, it has 26dB of gain at 10 MHz and 30 dB at 2 GHz. ? De?: [email protected] <[email protected]> De la part de pianovt via groups.io ? 38dB gain |
Cascaded amplifier for 8673D Watkins Johnson 6203-044 (HP 0955-0215 )
开云体育Hi, ? does anyone have any information about this Watkins Johnson 6203-044 or HP part number 0955-0215 amplifier? ? I am interested in its gain and frequency response in order to make a comparison with the one I have. ? |
Re: HP-8561E serial number decoding ?
开云体育That is not necessarily true. The numerical prefix refers to the Year and Week when that revision went into effect. It does not refer to when the first instrument of that revision was produced. It refers to when the Production Change Order was issued which would implement the Revision. This could be up to many weeks before any Instrument was built to this revision. Furthermore, if there were no more revisions to the product, the prefix could go on for years. Regards Bill Lauchlan ? From: [email protected] [mailto:[email protected]] On Behalf Of Sam Reaves
Sent: Sunday, February 06, 2022 6:39 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP-8561E serial number decoding ? ? The first on was made in 1994 and the second one was made in 1996. |
Re: 54503A
Amen Dave! I junked a 400MHz version of one of these scopes and gave away a 100MHz version (54501A?) to?one of?my interns. Service manual was?a joke, although I fixed a few of them. They have the worst UI of any scope I have ever used. I don't know what HP was thinking when they came out with these scopes. Obviously none of the engineers that designed that scope ever used an analog scope. It surprised me that HP went through with that project. The LeCroy scopes at the time were far beyond the performance of any of the HP or Tek scopes for that matter. Sam Reaves
ARS W3OHM
Owner / Moderator of: LeCroy Owners Group on Groups.io Sencore Owners Group on Groups.io Sprint Layout Group on Groups.io?? Pulsonix EDA Group on Groups.io LPKF Owners Group on Groups.io
Electronics and Mechanical Hardware Design Engineering Manager Staff Scientist Andritz Rolls Global Research Center (RETIRED) |
Re: HP 8753D Connection Adapters?
开云体育James, the APC-7 connector is sexless. You spin the collar to move the threaded portion in or out of the collar nut. You do the same on the on the mating APC-7 connector/adapter. The most common adapters are APC-7 to N-f, or APC-7 to SMA-f (APC-7 to N-m, or APC-7 to SMA-m are also available but less common).The APC-7 connector is rated to 18 GHz. Once you determine what your cable preference for the 8753D, then obtain the adapters and install them on the 8753D. I would recommend once installed you leave them on. ?APC-7 connectors and adapters are part of a dying breed and will become harder to find over time. Leaving them on will reduce the wear and tear on the APC-7 connectors on the 8753D. ?Those inputs are part of the internal reflection bridges and are expensive to replace. Here are some links to both types of adapters. Not recommending them other than Keysight ($$), but the adapters are reasonable if purchase used or from China (but research these carefully). Don Bitters |
Re: HP-8561E serial number decoding ?
HP instrument divisions had a number of serial number formats over the years:
One format was of the form:? YYWWCNNNNN Where: YY = an indication of the year when the product was introduced or a?major change was made to the product.? YY = the number of years?since 1960.? Thus YY = 23 indicates the year 1983. WW = the week of the year when the product was introduced or a major?change was made to the product; range of WW = 01-52. C = The country in which the product was manufactured.? The letter?"A" indicates the USA as the country of manufacture, "J" for Japan, etc... NNNNN = A unique number for each individual product within a product?model.? This was typically a sequential number beginning with?00101. Serial numbers <= 100 were usually pilot/proto instruments that? were for internal HP use/testing/development and not usually available to the outside world. The format of product serial numbers changed in the early 1990s (somewhere around 1993).?Note that on many instrument models, the serial number format was not changed until there was a major change in the model itself, or its manufacture location. NEWEST FORMAT The newest format is of the form:? CCYYWWNNNN Where: CC = The ISO code for the country in which the product was?manufactured.? The letters "US" indicates the USA as the?country of manufacture. Others include MY for Malaysia, SG for Singapore, JP for Japan, or GB for Great Britain.? YY = an indication of the year when the product was introduced or a?major change was made to the product.? YY = the number of years?since 1960.? Thus YY = 36 indicates the year 1996. WW = the week of the year when the product was introduced or a major?change was made to the product; range of WW = 01-52.? Use of?these two characters to indicate week of the year is optional.?Note that for some products, the two WW characters are not?used to indicate the week of the year, but are used as part of?a six character NNNNNN sequence number. NNNN = A unique number for each individual product within a product?model. This sequence number begins with 101.? For some products, the sequence number has six characters (NNNNNN) The computer/calculator side of HP have their own serial number format variations... |
Re: HP 8753D Connection Adapters?
开云体育And by the way, do not tighten the two collet nuts up against each other. That forces the mating surfaces apart and pretty much ruins the performance of the connector. Tighten the one that goes over the extended sleeve to pull the halves together. Also, there is a torque specification that is slightly more than finger tight. Seriously ‘gronking’ on it just damages things. ? From: [email protected] <[email protected]> On Behalf Of Kit
Sent: Sunday, February 06, 2022 5:44 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP 8753D Connection Adapters? ? Team, ? APC connectors are genderless. ? Regards, ? Kit |
Re: 54503A
Dave, Bob, ? Not only does the HP/Agilent/Keysight product line these days reflect the increased costs to design and produce a product given today’s higher wages and other huge operating expenses but I feel that there could be a little cost savings going on somewhere in there to minimize any residual cost impact to the customer from nonsense spending. ? The week before last I received a Keysight 2022 calendar that was shipped UPS ** overnight** from their promotions department.? I had already hung a calendar last month from another manufacturer that came to me via US mail. ? My watch and cell phone does a pretty good job of telling me the day and date without sinking the money barge. ? Greg |
Re: 54503A
On 2/6/22 13:03, Greg Muir via groups.io wrote:
It’s a shame that around this time HP was starting to “dumb down” their product documentation.Aside from paying less people to do technical documentation (to some extent) I do understand the profit motive to push assembly level replacement instead of repair but the end user also faces a considerable increase in trying to keep their equipment alive.Well, yeah, they "got suits". Their logic (if you can call it that) is "if it breaks, we'd be happy to sell them another one", and "we can't have people opening things up, they might do something stupid and sue us!" Screw high-minded civic thinking, respecting your customer, and trying to be the best. Surely THAT won't get you a good quarterly bonus. I love it when I see "NO USER SERVICEABLE PARTS INSIDE" on a piece of *test equipment*. "Wanna bet, bitch?" -Dave -- Dave McGuire, AK4HZ New Kensington, PA |
Re: 54503A
On 2/6/22 13:18, Bob Albert via groups.io wrote:
In the old days, HP did a commendable job.? As the years went by they became more profit oriented.Correction: "Short-term profit oriented". These suits can't see past the end of the quarter. Bill Hewlett and Dave Packard were looking a lot farther down the line than that. The decision to go public is often a big mistake when it comes to product value.It's common knowledge that being publicly-traded is generally bad for a company's customers and encourages self-destructive corporate behavior, when viewed in the long term. If you think about it, it makes sense, as the company's true customers are no longer those who purchase its products, but its stockholders. Everything boils down to "shareholder value". Disgusting. No longer do managers attempt to create products; they now attempt to make money.? Not that money wasn't always the ultimate goal, but some outfits carry the banner a bit too much.? A lot too much.Yes. And since suits typically have outsized egos, each successive generation of suits thinks they can "push things a little bit farther" than their predecessors, because of course they're better businessmen. So things deteriorate over time. I recently looked at Keysight's offerrings regarding the present successor to the original HP200.? They no longer seem to offer a reasonably priced audio oscillator.Analog oscilloscopes take real talent to design, digital oscilloscopes much less so. The former cannot easily be farmed out to India or China, while the latter can. Again, there's that short-term profit. Further, the current crop of tech guys, while I think they're generally doing great things, seem to lack any real knowledge of oscilloscopes. They think that an oscilloscope can be made by slapping a display on a microcontroller and connecting a wire to an A/D input. Sure, you can generally display a waveform that way, but it's going to be garbage. And the market is FLOODED with finished products labeled "oscilloscope" that are little more than that. These kids don't understand the difference, and they formulate assumptions based on incomplete knowledge, and then they vehemently defend those positions without being willing to learn about what they've missed. -Dave -- Dave McGuire, AK4HZ New Kensington, PA |
Re: HP 8753D Connection Adapters?
Hi James, The APC-7 connectors are genderless. They can interconvert to male or female. Tony On Sun, 6 Feb 2022 at 22:26, James Sawle (MD0MDI) <james@...> wrote: Thanks for the info, I have searched for this but can only find what I think is called the Male APC-7 to SMA or N-Type, I cannot even find any photos for a Female version of the Adapter, Basically all of the adapters that I find seem to have the very same fitting that is on the front of the Analyzer. --
|