¿ªÔÆÌåÓý

Date

Re: Looking for Agilent PNA old FW repo?

 

Have you checked with Keysight?? I recently requested FW for my unit and they provided me with A.07.50.67.? Not what you need, but I assume that they will help you out if you put in a ticket with Keysightcare.

good luck!
Scott


Re: 5086-7337 *Yig Oscillator issue*

 

Reading the schematics, I bring you some ideas, correct me if i'm wrong:

- I must measure the voltage drop across R1 (one of the power resistors in the plugin) and that will reveal the current sent to the YO coil, isn't it?

- The black heatsink on A9 board it's always hot. I think it is hotter than it should be, should check Q1 too? Maybe the transistor is shorted and it doesn't send sufficient current to drive the YO. What you think?

Regards,
Ruben


Re: 8566B attenuator fix

 

Thank you very much for answer. I guess your board is an extended version of mine : you have 4 solenoids instead of 3 for me (10+20+30db =60) but
do you know what is the 3 legs chip near the 8L05 regulator ? It is stamped C9Z and I have no idea...


Re: 8350B *restoration adventure?*

 

Ruben -

WELL DONE - Amazing work and obviously a lot of work. The 83590 series have some of the highest current draw, so if it works satisfactorily with these it should be good for most if not all plugs. One possible exception might be the plug that drives the mmWave sources, but you can deal with that when (and if) it occurs.

EXCELLENT WORK

Cheers!

Bruce

Quoting RubenRubio <rubenrb2019@...>:

*Great news!

* My 8350B sweeper is now working! Closed with a standard power cord. You can see some of the montage steps in the pics attached to this post.

Finally, I decided to swap the old burnt main transformer with switching power supplies and a little transformer on a tight budget. A real challenge of wiring, space management, mechanical assembly and proper cooling. After a several hours of work, today I've done the different PS adjustments with sucess. The changes include new 230VAC fan (with more CFM), restructured mains wiring, removal of the old big caps to maximize room for the switchers and new RAM battery pack (w NiMH cells).

Without the external feeding, now it's a more comfortable unit to continue the repairing attempt of the 83592A plugin...

Regards,
Ruben

PD: special thanks to Yves Tardif , without your PS data?I could never have done this repair ;)



Re: 8350B *restoration adventure?*

 

Great news!

My 8350B sweeper is now working! Closed with a standard power cord. You can see some of the montage steps in the pics attached to this post.

Finally, I decided to swap the old burnt main transformer with switching power supplies and a little transformer on a tight budget. A real challenge of wiring, space management, mechanical assembly and proper cooling. After a several hours of work, today I've done the different PS adjustments with sucess. The changes include new 230VAC fan (with more CFM), restructured mains wiring, removal of the old big caps to maximize room for the switchers and new RAM battery pack (w NiMH cells).?

Without the external feeding, now it's a more comfortable unit to continue the repairing attempt of the 83592A plugin...

Regards,
Ruben

PD: special thanks to Yves Tardif, without your PS data?I could never have done this repair ;)


Re: 50 Ghz step attenuator driver board

 

It is very far from being official, but looking at your picture it looks very similar to the one I'm working on right now.

/g/HP-Agilent-Keysight-equipment/message/133239


Re: 8566B attenuator fix

 

WELL DONE !

It would have been nice to get better closeup of recovering the contacts after pulling them through, but overall excellent !

Did you note any deformation of the leaves due to pulling the pins through? Did you have much trouble re-engaging the pins?

Thanks!

Cheers!

Bruce

Quoting bagojfalvibagoj@...:

If you guys want to watch me unscrew screws for 30 or so minutes feel free:


(Ignore any awkward background noises, I didn't check the audio. I only recorded this so I know how to put it back together.)

The last clip is debugging with a DMM. Static voltages are ok, but it definitely gets stuck in some abnormal state. The coil gets hot.

I also added a scribble about what I believe the schematic of the control looks like.



50 Ghz step attenuator driver board

 

Hello,?
Does anybody have the schematic diagram --and eventually the parts layout --of the latest 24V driver board? fitted in the 60db 50 Ghz step attenuators (the older ones did have the traditional double V mechanical contacts). Thank you.


HP 35670A display questions

 

What sort of display is used on a 35670A, and did this change to an LCD
at some point? Was a CRT ever used?

From this post:

/g/HP-Agilent-Keysight-equipment/message/78558

it seems to have been an ELD display at some point.

I ask because having looked at pictures of various units, there seems to
be two different displays: one with an orangish text and another that
looks more yellow, typically on newer Agilent units. The yellow displays
look nicer to my eye.

Unfortunately, I couldn't find any information about this in the various
manuals.

Thanks
Matt


Re: My 6655 thinks it is a 6653

 

I don't really know enough about running HP BASIC programs to say for sure, but HT BASIC does look like a promising option to me.

Another option would be to figure out what the program is doing and run the GPIB commands manually.
From a quick glance over it, it first defines arrays of data addresses & lengths and different arrays of values for the different models. When copying the array of values to be used, it replaces positions 21 and 32 with the model# selected (e.g: it replaces 6555 with 6655).
For the "initialization" mode it turns on cal mode with "CAL:STATE ON,<passcode>", writes each value to the eeprom with "DIAG:EEPR <address>,<length>,<value>", then turns off the cal mode again with "CAL:STATE OFF".


On Fri, 10 Mar 2023 at 15:12, Brendan Murphy <ranger89ish@...> wrote:
Hmm, ok. So if I bought something like HT BASIC to run the listed program from pages 66 to 71 that should theoretically fix it then? Just asking for thoughts and input before I go out and just buy that.


Re: 8566B attenuator fix

 

If you guys want to watch me unscrew screws for 30 or so minutes feel free:


(Ignore any awkward background noises, I didn't check the audio. I only recorded this so I know how to put it back together.)

The last clip is debugging with a DMM. Static voltages are ok, but it definitely gets stuck in some abnormal state. The coil gets hot.

I also added a scribble about what I believe the schematic of the control looks like.

?


Re: 8566B attenuator fix

 

Well good luck - keep us informed. For more information on proper drive, you could look at the schematic of the DP attenuator driver - or masure what the 8566 does - I believe it is a pulse from the CPU but not certain.

Cheers!

Bruce


Cheers!

Bruce

Quoting bagojfalvibagoj@...:

I ordered a breadboard overnight. Never have one when you need it.
I have a frightening suspicion, which is that this attenuator works perfectly I just couldn't drive it properly. I hope at least I didn't make it worse.
Now I kinda understand what the circuit is doing. Hopefully tomorrow I can check properly.

The original symptom was about 50dB loss on the input in both frequency modes, so that makes it either the connector, cables, the attenuator or the switch.
I took the attenuator out first as I read that is likely to have issues and since I couldn't turn on 40dB attn on it I concluded this is the culprit.
But maybe not...

Rabbit holes rabbit holes everywhere.



Re: 8566B attenuator fix

 

I ordered a breadboard overnight. Never have one when you need it.
I have a frightening suspicion, which is that this attenuator works perfectly I just couldn't drive it properly. I hope at least I didn't make it worse.
Now I kinda understand what the circuit is doing. Hopefully tomorrow I can check properly.

The original symptom was about 50dB loss on the input in both frequency modes, so that makes it either the connector, cables, the attenuator or the switch.
I took the attenuator out first as I read that is likely to have issues and since I couldn't turn on 40dB attn on it I concluded this is the culprit.
But maybe not...

Rabbit holes rabbit holes everywhere.


Re: 8566B attenuator fix

 

I think that is correct.

I don't have the problem you do - I bought one of the HP relay drivers and built a small board with a DIP socket, that lets me plug in the attenuator and drive it using the switchwes. The drivers are cheap ($25) but the cables with matching plugs can be a bit harder to find.

Cheers!

Bruce

Quoting bagojfalvibagoj@...:

I'm looking at the board now and other than the ULN2003A drivers, there are 4 D flipflops which makes me wonder if this needs continuous supply in order to retain state correctly.

Since I was turning the psu on and off it might've caused the improper behavior.



Re: 8566B attenuator fix

 

I'm looking at the board now and other than the ULN2003A drivers, there are 4 D flipflops which makes me wonder if this needs continuous supply in order to retain state correctly.

Since I was turning the psu on and off it might've caused the improper behavior.


Re: 8566B attenuator fix

 

I already managed to free the leaves. I think I was lucky. They seem ok to me.

I would guess that the magnet is necessary to make the system bi-stable. The shaft sticks to it on one side and stays there until the coil shoots it over to the other side to stick there. The solenoid still should work as a solenoid I believe, but it doesn't look like doing a very good job at that.


Re: 8566B attenuator fix

 

One thing I forgot to mention is that the magnets are necessary for proper operation of the plungers - I would not try to operate the device with them removed.

Again, I will defer to Don Bitters on this as he has much more experience.

Cheers!


Quoting bagojfalvibagoj@...:

Anyway, the plunger thing is going to be a tomorrow problem.

I noticed two things upon disassembly. This thing looks brand new inside. I have never ever in my life saw such healthy looking o-rings. Second, there's a lot more electronics than I anticipated.

I took the magnet off one of the solenoids and tried to make it toggle. It seems to be very very weak and sometime it doesn't even move. In such a case the current draw stays high which I don't really understand. The shaft seems to be moving reasonably freely.

Here's a video I made. I'm turning the psu on and off like 3 or 4 times. You can see it slightly move at 2 sec, until for the last cycle it actuates properly.
Video ( )

Ideas?



Re: 8566B attenuator fix

 

Need to see more pictures. Generally, you can manipulate the plunger w/o removing the magnet - but I have not worked on many of the attenuators with the PCB inside - most the simpler variety.

You can recover the leaves by GENTLY pulling them up using tweezers (to lift, not pinch) or (as I do) a dental pick of the correct shape to slide under and lift.

I do not see the switch (fingers that are interrupted by the plunger) in this picture - could you show more ?

Cheers!

Bruce



Quoting bagojfalvibagoj@...:

Anyway, the plunger thing is going to be a tomorrow problem.

I noticed two things upon disassembly. This thing looks brand new inside. I have never ever in my life saw such healthy looking o-rings. Second, there's a lot more electronics than I anticipated.

I took the magnet off one of the solenoids and tried to make it toggle. It seems to be very very weak and sometime it doesn't even move. In such a case the current draw stays high which I don't really understand. The shaft seems to be moving reasonably freely.

Here's a video I made. I'm turning the psu on and off like 3 or 4 times. You can see it slightly move at 2 sec, until for the last cycle it actuates properly.
Video ( )

Ideas?



Re: 8566B attenuator fix

 

Anyway, the plunger thing is going to be a tomorrow problem.

I noticed two things upon disassembly. This thing looks brand new inside. I have never ever in my life saw such healthy looking o-rings. Second, there's a lot more electronics than I anticipated.

I took the magnet off one of the solenoids and tried to make it toggle. It seems to be very very weak and sometime it doesn't even move. In such a case the current draw stays high which I don't really understand. The shaft seems to be moving reasonably freely.

Here's a video I made. I'm turning the psu on and off like 3 or 4 times. You can see it slightly move at 2 sec, until for the last cycle it actuates properly.

Ideas?


Re: 8566B attenuator fix

 

Waaaay too late. On the other hand I didn't need a video for it. I managed to screw that up totally on my own.

I noticed now how one of the little flappy flaps is stuck. :/