Re: 8510C questions and cabling
Yes you can't daisy chain the but you need option 001 in one of the test sets...I do believe. Pete wa2odo [email protected]> wrote:
By
Pete Manfre <pmanfre@...>
·
#104544
·
|
Re: 8510C questions and cabling
None of the cabling is obvious from the Preflight Checkout diagram or the entire document. At this point my concern is identifying specialty HP cables I need to get. I'd like to be able to switch the
By
Reginald Beardsley
·
#104543
·
|
Re: HP 3455a DVM repair
Went *up* and overheated? That doesn't sound right. I'd bet that it overheated and went up in value as a result. Are there tantalum capacitors in a 3455A? (I had one for years, long ago, but never had
By
Dave McGuire
·
#104542
·
|
Re: Fixing a sick 4332A LCR meter
Joel, I have been following the troubleshooting tr
By
Sean Turner
·
#104541
·
|
HP58503A GPSDO changing firmware to remove 1PP2S (Option H14)
I found an HP58503A GPSDO locally, with a faulty psu, but very cheap (?50). The PSU is a quick repair, using a Meanwell PT-65S (5V/1A, +/-15V 0.5A) Unfortunately my version had RevC "Option H14",
By
Jeremy Owen
·
#104540
·
|
Re: Fixing a sick 4332A LCR meter
Ok, then I'd suggest we focus on getting the oscillator running again ! Since it used to be OK, you may want to start by looking for any broken wires, etc ... Maybe it's just a broken wire. The
By
Joel Setton
·
#104539
·
|
Re: Configuring an 85301C test range using an 8530A and 83631B
Yes, I've done exactly this, minus the PC. I threw together a simple stepper controller, got a cheap stepper-driven rotary table, and tied it into my 8530A. It worked fine, though I'd really like to
By
Dave McGuire
·
#104538
·
|
Re: HP 3455a DVM repair
After replacing the resistor all seems okay in that area.? No other failures.? Just a bad resistor that went up in value and then overheated. The latest is that a new failure has surfaced.? Now the
By
Bob Albert <bob91343@...>
·
#104537
·
|
Re: 58503A GPSDO PSU details/availability
Hi Andy, I asssume you sorted this out after all this time, but if not, I can send you the wiring details. And it's just a very ordinary PSU.? 5V 1A and +/-15V 0.5A. Look on EEVBLOG and you will find
By
Jeremy Owen
·
#104536
·
|
Configuring an 85301C test range using an 8530A and 83631B
I've been studying the Agilent 85301B/C configuration guide, 5967-6042E, with the aim of being able to set up an automated anechoic test range for the 2.4 & 5 GHz ISM bands. My 8530A does not have
By
Reginald Beardsley
·
#104535
·
|
Re: Fixing a sick 4332A LCR meter
OK, this is getting ridiculous. Now it's "working" again. Almost makes me think there's something mechanical going on too. I hate finding mechanical issues...I have a Tek 7104 that had the resistor in
By
Sean Turner
·
#104534
·
|
Re: Fixing a sick 4332A LCR meter
I think that something else has broken...the C function was somewhat functional before, and now it is not. Both C and L, the needle remains below zero scale and there is no oscillator activity.
By
Sean Turner
·
#104533
·
|
Re: Fixing a sick 4332A LCR meter
Couple of further notes: - The signal at A1TP1/A1Q2 emitter disappears entirely in L or C - The frequency does not change no matter what range we're on - The signal actually is correct at 70 mVrms...I
By
Sean Turner
·
#104532
·
|
Re: Fixing a sick 4332A LCR meter
Joel, 1) no worries 2) OK, I'm not sure how I didn't put 2+2 together on that but it makes perfect sense now. The 427A has a log scaled resistance scale! 3) A1TP1/A1Q2 emitter (in R mode, with a 1K
By
Sean Turner
·
#104531
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
by the way, later you can download the V4.3 and V4.5 from Tony G depository (above) and update the software first to 4.3 and then to 4.5 for that you simply need to connect that CD drive to the IDE
By
amirb
·
#104530
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
by the way, when you say the scope suddenly died, what exact
By
amirb
·
#104529
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
+1 for Amirb's suggestion of a CF IDE adaptor. I've done this on a number of instruments. Robert G8RPI.
By
Robert G8RPI
·
#104528
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
Hi Rik, That is because the format that they use for storing the likely compressed data is their own proprietary format. dd can backup/copy a full disk, or a partition, or a file, it doesn't care. It
By
Chuck Harris
·
#104527
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
A ghost image is a backup (full disk or only the partition) made by Norton ghost a DOS later Windows application (Norton later Symantic), it uses a proprietary technology to back up the disk. You
By
Rik Bos
·
#104526
·
|
Re: First Gen HP 54810A with dead HD, Got ISO 3.5, Any way to write it to a drive?
dd simply copies the disk exactly as the data appears on the drive. It is used in forensics to clone drives from crime scenes as it makes an exact copy down to the partition tables, and bad block