Re: E4401B No 50MHZ Cal signal


 

The only advice I can offer is not to adjust anything until you fix the problem. I don't have this instrument.
Hopefully you have a good block diagram and a description to figure it out. You are lucky to have the schematics.
Peter


On Sat, Mar 30, 2019 at 1:06 PM Michael Williams <mike@...> wrote:

Hi Peter,

?

I am aware of the differences between a 75 ohm input and 50 ohm input.? I have a 50 to 75 ohm impedance matching pad that I will be using on the input when measuring signals with a 50 ohm source.? I know this is not ideal, but will work for the things I plan to do with this gear.

?

The issue I am having at this time is that the *internal* 50 mhz calibration signal is not getting generated, causing issues during the alignment procedure.? Unfortunately, on this model, the input is switched internally to the calibration signal, so I can¡¯t even use an external signal to get through the alignment.

?

Thanks,

?

Mike

?

From: <[email protected]> on behalf of "peter bunge via Groups.Io" <bunge.pjp=[email protected]>
Reply-To: "[email protected]" <[email protected]>
Date: Saturday, March 30, 2019 at 11:59 AM
To: "[email protected]" <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] E4401B No 50MHZ Cal signal

?

Are you using 75 ohm components to calibrate it?

You do realize that it will not work with 50 ohm components?

?

On Sat, Mar 30, 2019 at 12:50 PM Michael Williams <mike@...> wrote:

Greetings,

?

I have an Agilent E4401B with option 1DP (75 ohm input) and option A4H (GPIB interface) that I recently found at a swap meet.

?

It powers up, displays a trace, and when I apply a test signal to the input, I can see it on the display ¨C so far so good.? The only downside at this point is a persistent message that an alignment is necessary.

?

The problem is that the alignment procedure fails.? After running an ¡°align now¡± the noise floor is off the top of the screen.? I have also noticed that when selecting the internal 50 MHZ source, I do not see a corresponding signal on the screen.? This leads me to believe that the internal 50 mhz signal is not getting generated, causing the alignment procedure to turn the internal gain up way too high.?

?

I have the component level schematics for the device, and have confirmed that on the RF assembly, the 2nd VCO is not receiving power because the ¡°TEST MUX¡± ¨C U56 is not getting enabled ¨C All of the MUX lines (MUX_EN, MUX_0-MUX_2) are all at 0 v.?

?

This is where I¡¯m stuck ¨C I see that U55 (Fractional Divider) has connections the MUX pins, but have no idea what tells it to place the MUX into a given state.? I can¡¯t tell if this is a logic/software issue, cabling, or a bad U55.

?

I¡¯m hoping that someone here with more familiarity with these instruments has some insight as to what might be causing this, and next steps I should take to figure out the issue.

?

Thanks!

?

Mike Williams