开云体育

Re: Synthesizer module in ESG series E4435B


 

So just out of curiosity where didi you find the schematics?? I thought there no schematics avaible anywhere for this device?


Il giorno mer 14 nov 2018 alle ore 23:41 Milan Vasic <vasa1958@...> ha scritto:
Someone ask for ESG Fractional N divider.
Not my and don't ask for better resolution.?
Milan

On Sun, 11 Nov 2018, 14:35 Tom B <tbryan@... wrote:
Hi Leo,

Which synthesizer board do you have?

There is a service note (E4436B-01F) that says: "Serial Numbers:
[0000A00000 / 9999Z99999]" "The ESG-B AP/DP Family of Signal Generators
has been experiencing failures in the Frac-N board assemblies. This is
due to a failure in the divide-by-eight circuit of the Divider chip A24
U313 which causes the ESG to display Error 511 "Output unleveled". All
ESG-B AP/DP units with the serial numbers mentioned above may experience
Error 511 along with UNSTABLE or NO OUPUT below 1GHz. If this occurs,
the A24 Frac-N/Divider Assembly (E4423- 60019) needs to be replaced.
This assembly should be ordered using part number E4423-69019 and should
be treated as a warranty repair. The defective assembly should then be
returned with documentation referencing this service note as "Error 511"."

Tom Bryan



On 11/11/2018 4:39 AM, Leo Bodnar wrote:
> I have E4435B signal gen from ESG series that loses output in certain
> parts of its output range 250..350MHz, 500..700MHz, etc.
> The problem seems to be in the VCO or its PLL loop.
> Its output is supposed to be 500..1000MHz before further
> multiplication/division but lower part of the range is unstable. As
> you go down in frequency from 1000MHz to 500MHz, somewhere around
> 690MHz it becomes noisy and at 680MHz it loses signal altogether.
> These siggens sometimes have prescaler U313 1GC1-4210 failure, but
> here it does not seem the case as it is further down the signal path -
> after YIG multiplier that converts 0.5-1GHz VCO output into 4-8GHz signal.
>
>





--

Alberto Vaudagna

IDRA-XAM Electronics Firmware and Software division manager- Team H2politO

H2politO – Politecnico di Torino

Mob:?+39?347?3906977

?

This e-mail and any attachments may contain confidential material for the sole use of the intended recipient. If you are not the intended recipient,?please be aware that any disclosure, copying, distribution or use of this e-mail or any attachment is prohibited. If you have received this e-mail in error,?please contact the sender and delete all copies. H2politO kindly reminds you to consider the environment before printing.?Thank you for your cooperation

Join HP-Agilent-Keysight-equipment@groups.io to automatically receive all group messages.