Electrolytics strike again...
Got 8751A VNA failing power-on test some time ago and finally got some time
to fix it.
It had ALL THREE receivers failing self-test. Shuffling them didn't change
anything (it is three IDENTICAL boards). Pulled them from the unit, removed
covers and visually inspected them. On the first glance everything looked
perfectly neat -- no obvious signs of something blown, very good assembly
job, high quality components -- so I prepared for a difficult fight getting
in the very guts.
However, on the second, more thorough visual inspection something caught my
eye. There was a barely visible kinda like oily film and tiny droplets on
part of the board around those AK9201A-VP Digital Filter chips. Barely
visible, very easy to miss, as if boards were not washed thorougly enough.
Closer look revealed some slight solder joints corrosion in that area.
Nothing major, almost natural but anyway...
All alarms went off immediately, reached for my desoldering station gun and
pulled all electrolytics from one board (there are 9 x 10uF50V, 2 x 47uf35V,
7 x 100uF35V, and one 470uF16V, all genuine KME type, looked perfectly OK)
and found that all those 10uF50V around DF chip failed miserable with 100+
Ohm ESR and their internal liquid totally gone. That was that oily film
around then covering quite a big area. All others also leaked some but not
that bad so they were still within specs (on the margin of failure) and they
also produced that stinky odor when desoldered and removed.
Luckily enough those boards were of way better quality than infamous
Tektronix 2465B A5 board. Also the genuine through-hole Japanese Nichicon
KME seem to be filled with less aggressive blood than those SMD caps on
2465B's A5 so after thorough washing with all electrolytics removed boards
looked pretty fine, no need for restoration job.
New Panasonic FR capacitors were installed then, boards assembled, put back
in the 8751A and bingo! -- No Errors Found, works like a charm...
Will probably replace all remaining electrolytics before calibrating and
putting it for sale, then repeat for my primary one sitting on the workbench
and used from time to time...
Resume: one should _ALWAYS_ replace _ALL_ aluminum electrolytics in _ALL_
instruments that mature before calling it refurbished and taking on
calibration. No matter how nice those electrolytics look, where they were
made, where the entire instrument was assembled (8751A are fully Japanese so
one would expect high quality parts and assembly) -- _REPLACE THEM ALL_ It
is not worth a bother checking them one-by-one and only replace ones that
already failed -- they will _ALL_ fail soon. And it is actually worse with
high end instruments -- those failures are _GRADUAL_ so it is not like
something goes BOOM! and instrument goes dead. They start drifting out of
specs instead, becoming flakey and unreliable way before they finally fail.
In my early years, 40+ years ago, we were taught that electronics is all
about contacts -- if it didn't work some contact was missing where it's
needed or present where it is NOT needed.
These days that sacral knowledge should be probably replaced by new one,
about aluminum electrolytics. 99.99% failures I see for last 10 years or so
were caused by bad capacitors...
---
*