The 8553B does not have a LO output jack, only the 8553L does.
To use a non-option 007 8601A with the 8553B would require
8601A modifications and a special cable to connect to the
141T.
Regards,
Dave Wise
toggle quoted message
Show quoted text
-----Original Message-----
From: hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]On Behalf Of iw3sgg
Sent: Wednesday, November 21, 2007 12:14 PM
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: Hp141+8553B+8552B and HP8601A
--- In hp_agilent_equipment@..., "jaking"
<jaking@...> wrote:
Look in the 8601a manual (read Option 7).
There is provision to take control signal from the 8601,
and use that to
control the spectrum analyzer.
The frequencies between the analyser and the 8601a are the same.
Yours sincerely
ZL4AI
Hi ZL4AI,
Thanks for repply.
I'm looking the manual and I'm looking the HP8601.....I havent the
option 7.
I have only a VTO OUTPUT connector,the LO INPUT connector is not
mounted on the rear pannel.
If I read right I must connect the first LO OUTPUT op spectrum
analyzer (HP8553B) to the LO INPUT of HP8601A.
It is correct??
For you is possible to modify my HP8601A to ADD the 7 option??
Now I'm reading the manual from bama site.
Thanks for repply again.
Best regards.
iw3sgg