Re: 8566B SA all PLLs unlocked


 

¿ªÔÆÌåÓý

Finally got a chance to work on it and found the culprit. It¡¯s the SCR for the -40V crowbar.

The SCR (CR25) conducts, or leaks while Vgc is zero. This is a false over-voltage condition tripped the over-current protection circuits(Q6) therefore shuts down the -40V supply but current leakage on the SCR is not strong enough to blow up the fuse for the -40V. Removing the SCR then the -40V is well and alive but of course I lose the over-voltage protection which is a no-no as two of the consumers of the -40V is the YTO, YTX main coil drivers, expensive ones.

Now that CR25, 2N4186 is a MOTOROLA silicon control rectifier which is long obsolete, what would be a good substitute?

Thanks,
Calvin


On Jul 20, 2018, at 10:16 PM, Calvin Guan <guancalvin@...> wrote:

Hello,

I turned on my 8655 yesterday and found out that I don¡¯t see the staircase noise floor. PL1,PL2 M/N, YTO, REF all shown as unlocked. Instrument responds to first key press, and then display freeze. It does always respond to the green instrument preset key thought. No good, last time I used it is 3 month ago and it was working.

First suspected the CPU and front panel boards but nothing seemed wrong there. After some troubleshooting, the culprit was found to be the negative regulator where -40 is dead. I pulled a neg regulator A18 from a spare unit and swapped, now it¡¯s great again ¡ª whew.

I¡¯m trying to figure out why the -40 is dead as the A18 is quite expensive on ebay. I don¡¯t see anything obvious wrong. The fuse is ok suggesting the -40 crowbar is probably ok. Did diode test on Q5, Q6 and the diodes around the -40 regulator circuit, seems ok to me. Have anyone seen this?

Thanks,
Calvin