Re: HP-8443A Tracking Generator Problems


David C. Hallam
 

I don't have the test equipment you outlined. My counter is a HP-5245L and
limited to 50 MHz. I have been looking for additional plug-ins to extend
its range but at the moment I don't have any. My scope is a TEK 465 which
has a bandwidth of 100 MHz. Since my interest is in the HF (3-30 MHz)
portion of the ham radio spectrum, I just never saw the need to have
extended range test equipment.

I did what I will call a quick and dirty test. I connected the 30 MHz
calibration signal from the 8552B IF section to the RF input of the 8553B RF
section. At a 30 MHz setting on the 8553B frequency dial, I get the signal
peak on the 141T screen along with the marker from the 8443A. I can turn
the 8443A on and off and the marker pip also goes on and off. The 8433A
display reads all 0's.

If I set the 8443A to the external (counter) mode, it will correctly display
the frequency of a signal from my signal generator.

Rightly or wrongly, what do I conclude from all of this?

1 The counter portion of the 8443A is working correctly
2 The tracking generator working correctly
3 The no display problem in the tracking mode is related to counting of
the signal output of the tracking generator rather than a signal generation
problem in the tracking generator.

David

-----Original Message-----
From: hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]On Behalf Of
microwaveengineer1968
Sent: Thursday, November 15, 2007 9:31 PM
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: HP-8443A Tracking Generator Problems



This is the link for a manual for the 8443A

Now down to the bone my friend, i have looked at the manual and
compiled some preliminary checks for you to do, this assumes that you
have acess to a counter to 500MHz and maybe a powermeter or scope
with appropriate bandwidth.
First off you need to verify the operation of the tracking generator
portion of your unit, dont worry about the counter for now, basically
the 8443 receives three signals from the 8553 and 8552 units, those
signals are mixed together sequentially to form a 0-110MHz signal
which is then amplified and leveled, the manual has a nice block
diagram in it btw !
The first signal is a 200-310MHz which is basically the 1st LO of the
8553, the amplitude is supposed to be -6dBm, the next signal is a
150MHz fixed LO @ -2dBm and the third Signal is a 47MHz fixed lo @ -
7dBm
First connect your counter to the RF output of your 8443A, set the
attenuator to 0dB and then set the analyzer to zero span and tune the
frequency and see if the counter indicates a frequency reading in
accordance to your analyzer scale, if you dont get any indication you
will have to verify the presence of your LO signals, you may just
have a bad interconnect cable to your analyzer, measure the 150MHz
signal on A11 J1 (you have to remove the cable and then measure the
signal coming out of the cable) , measure the 47MHz signal on A13 J1
(once again measure whats coming out of the cable) and the 3MHz fixed
LO on J3 (cable) then move on to A9 J1 on this cable you should be
able to measure a signal from 200-310MHz the frequency should be
tuning when you turn the tuning knob on your RF Section.
Do those tests and let me know the outcome.


--- In hp_agilent_equipment@..., "David C. Hallam"
<dhallam@...> wrote:
>
> I have some problems with my 8443A and need help locating the
issues. The
> nixie tube display shows all 0. I believe the tracking generator
is still
> working because the marker is on the screen of the 141T. I can
connect the
> RF output of the 8443A to my HP-5245L counter which will display
the marker
> frequency. I can move the maker with 8533B frequency control and
the
> display on the 5245L will change with the movement of the marker.
>
> The 8443A will function as a counter because I can set it in the
external
> mode and use it as a frequency counter.
>
> I followed the counter trouble shooting tree in the manual and
reach the end
> point where it says to repair A5. However in the service section
for A5, I
> get oscilloscope displays from each of the 6 test points on A5 that
match
> the photo in the manual which leaves me wondering. I don't have the
> extender to remove the module from the 8443A and measure the
voltages, etc.
> at the points indicated on the repair table.
>
> Does anyone have any suggestions about what to do next? Is my
conclusion
> that the tracking generator portion functioning properly logical?
My skill
> level as an electronics technician is not that great with
instruments like
> this.
>
> David
> KC2JD/4
>

Join [email protected] to automatically receive all group messages.