Hi Adri,

In the 8594E, I can check the power supply voltages on the A16 boards, there
is 4 test points for 5v, +12v, +15v and -15v. May be it is the same with
your SA?

Roland F8CHK


________________________________

De : hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]
Envoy¨¦ : mardi 11 juillet 2017 22:00
? : hp_agilent_equipment@...
Objet : [hp_agilent_equipment] Re: 8595E suddenly faulty




thanks Scott and thanks Hermann for your advice
yes I got the service manual and I'm trying to collect the info for the
troubleshooting

Searching the Web I found a very limited number of faulty YTO and
considering that it was mounted on the whole 858xfamily I was expecting much
more occurrences (there must be a huge number of them out there)

Nevertheless I agree it is one of the possible culprits and I'm thinking
about how to check its output as my only other S.A. goes only to 1.8Ghz
(this YTO , part nr5086-7903, starts from 3.07GHz)


Anyway I'm also hoping for a hint about a way to check the power supply
voltages without pulling it out

thanks
adri