Re: 8754D vs 8753ES


Dr. David Kirkby (Kirkby Microwave Ltd)
 

On 20 October 2015 at 15:36, W2HX w2hx@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:


Other than the LCD display, is the 8753ES any ¡°lesser¡± of a VNA than a similarly outfitted ¡°D¡±? I ask because it seems that many ES¡¯s go for less money than the ¡°D¡±s. is this just a fluke of a lot of supply of the ESs? Any thoughts on this?

Thanks

Eugene W2HX


Not that I am aware of. The ES has the advantage that there are 4 buttons for switching channels. It also takes later firmware, and I suspect that means it has a later (faster) CPU.

Now I don't know what difference there is between an 8753E and an 8753ES. That might be worth investigating more, but i would certainly not buy a D unless it was a lot cheaper than an E or ES.

Despite the extra bulk, I'd buy one with an external test set, as it gives you a lot more flexibility. Note the 6 GHz option on the D/E/ES are all hardware, so you can't easily upgrade them.

Dave

Join [email protected] to automatically receive all group messages.