Re: Hello, and 16500A memory sizes


 

On 19/01/2013 16:05, Glen Slick wrote:
On Sat, Jan 19, 2013 at 7:50 AM, Chris Jones chris@...
<mailto:chris%40stumpie.com>> wrote:
>
> Now I've installed ROMs version 00.02, and the 16500A has developed a
> new fault: the self-tests 'HIL Controller', 'Front Panel' and 'Touch
Screen'
> all fail. I thought it was a software problem but putting the old 00.00
> ROMs back in the problem is still there. I'm hoping it's just a bent pin
> on the backplane connector or something but it's very annoying to get
this
> far and have a new fault to deal with, apparently unrelated to the RAM.
>
> Has anyone got experience of what's likely to cause these three self-test
> failures?
>
> Chris

Don't all of those things connect through the same ribbon cable
between the front panel and the CPU board? Are you sure that cable
was reconnected correctly at both ends when you reinstalled the CPU
board? Or does that cable connect between the front panel and the
backplane board? It's been a while since I opened up a 16500.
I've sorted out that problem:


It was a broken ceramic resonator providing the clock to the custom HP chip which handles the interface to the front panel. Having fixed that, it passes the self tests...but doesn't see the discs in the floppy drives. This time, putting software version 00.00 back in makes the floppy drives work again, but something about the floppy disc hardware seems incompatible with software 00.02. This is annoying now...

Chris