开云体育

Re: 3457a on the way


Peter Gottlieb
 

Like this one:?



I shouldn't imagine there's much special about these except the physical size.

On 12/31/2012 7:25 PM, J. L. Trantham wrote:

Steve, Jeff, and all,

My 2703Axxxxx 3457A also has the SAFT LX 1634, Lithium 3.0V, battery and it
measures 3.032 VDC. Likewise, I can not see a date code on the battery,
even after removing the A1 Board and looking as far under the battery as
possible.

My A1 Board is 03457-665xx, REV A, 2703. According to the manual, this is
the 'New Main Controller'. However, on my board, A11R644 is a 17.4 K
resistor and A11R645 is a 13 K resistor which represents the 'old' resistors
and matches the schematic. The 'new' resistors would be 13 K and 12.7 K
respectively, according to the parts list and 'Changes'. The 'old' battery
is listed as a 'Battery 2.9V .9A-HR Li/S-Diox W-Flex', according to the
Agilent website. Likewise, the Agilent website lists the 'new' battery as a
'Battery 3V 1.2A-HR Lithium Poly Carbon'. Only the manual parts list lists
the 'new' battery as 3.4 V.

In addition, I don't think I have ever seen a 3.4 V Lithium battery.

So, the question is, when it comes time to replace the battery, what should
be used?

Joe

-----Original Message-----
From: hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>
[mailto:hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>] On Behalf Of Steve Krull
Sent: Monday, December 31, 2012 11:20 AM
To: hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>
Subject: Re: [hp_agilent_equipment] 3457a on the way

Joe and all,

I just had a quick look inside my 3457A again. Mine has the 3.0 volt
lithium battery, SAFT LX-1634. Obsolete at Agilent, as is the newer
battery. Google was no help either. Mine measures 3.03 volts and there's
no evidence of corrosion so that's good. I couldn't see a date code on
it; probably on the underneath side. I've replaced batteries by
paralleling the existing connections with an appropriate power supply
and then unsoldering the old battery with an isolated-tip iron and
soldering in the new battery. I've also carefully soldered a new battery
in parallel with the old and then clipped out the old one. No problems
with lost data so far.

I'm not sure how the cal numbers increment. I'll have to experiment with
that sometime. Right now I need to repair the 1349D display in my 8757A
so I can get on with a sweeper plug in project, so the volt-nuttery is
on hold for awhile.

Happy New Year to all!

Steve

On 12/30/2012 5:34 PM, J. L. Trantham wrote:

Steve,

Thanks for the data. Mine is also 2703A prefix with REV?:6,0 and
CALNUM?:98. Not a multiple of 34. Would be interesting to see what the
CALNUM increments by after an Agilent CAL.

I, too, need to look at the battery condition. I have not looked at the
manual regarding replacing the battery. Has anyone done that without
losing
the CAL Constants?

Joe

-----Original Message-----
From: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
[mailto:hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>] On Behalf Of Steve
Sent: Sunday, December 30, 2012 4:14 PM
To: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
Subject: Re: [hp_agilent_equipment] 3457a on the way

It would be interesting to see if a newer rev is out there. My 3457A
is s.n.
prefix 2703, with rev 6,0 and option 0, CALNUM=34. I wonder if that's a
default number for anything less than a full cal at Agilent? The last
calibration was at least 6 years ago and performed by what was then Boeing
Military Airplane Company's metrology lab. I need to open it up and check
the battery condition.

Steve

On Dec 29, 2012, at 8:08 PM, "J. L. Trantham" <jltran@...
<mailto:jltran%40att.net>
<mailto:jltran%40att.net>
<mailto:jltran%40att.net>
<mailto:jltran%40att.net> > wrote:

If the 'SELF TEST OK' message appears, there is no need to make any
'adjustments'. Just do the 'front panel CAL' if needed.

As I said, I would check it out, assume it is the best instrument in
your
collection, send it to Agilent for CAL and see what you get.

I would appreciate knowing what 'REV?' and 'OPT?' says when you get a
chance. 'CALNUM?' would be interesting as well.

In the 3458A, the firmware is in an EPROM (6 EPROM's for the older
units)
and can be removed, a socket placed, and easily upgraded by
purchasing the
latest pre-programmed EPROM (or EPROM's for the older units) from
Agilent.
The only problem is they have a $50 minimum for this $18 part for the
later
units.

My wife thinks I am going to appear on an episode of 'Hoarders'.

Joe

-----Original Message-----
From: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
[mailto:hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com> ] On Behalf Of Jeff
Machesky
Sent: Saturday, December 29, 2012 8:01 PM
To: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
Subject: Re: [hp_agilent_equipment] 3457a on the way

Wow, lots of replies all of a sudden. I'm already prepping the wife for
the $200 + price tag of calibration. Funny how when I was single I
would
have about $800 in cash in my wallet at all times and now I beg for 20
bucks, Hmm. Sad part is I make about 4 times the money. In any event
I've not received to much feedback on the "Self Test OK" message the
seller had posted. Any comments? I'm too much of a skeptic when it
comes
to eBay purchases. It's just a convenient place to purchase such goods.
Any feedback would be appreciated as to possible pitfalls regarding
this
device. I like to prep for issues rather then build myself up for
failure. What do they say...it's better to be pleasantly surprised then
let down.

Thanks,

Jeff

On 12/29/2012 5:40 PM, J. L. Trantham wrote:

According to the manual, there are only two 'adjustments' that can be
made
on a 3457A, Input Offset Amplifier adjustment and AC Converter
Frequency
Response, both needed only if there is a 'HARDWARE ERR' failure
message
after 'TEST' is selected and then only if it is a specific
'AUXERR' or
16 or
256 is seen. Otherwise, all the calibrations are done from the front
panel
with specific inputs from the front panel.

The CALNUM? is incremented by 'several digits' with a 'complete
calibration', one for each calibration point entered, per the manual.
Interestingly, when I sent my two 3458A's to Agilent for calibration,
the
CALNUM incremented by only 1. However, when I calibrated one of them
before
sending it to Agilent, (since I lost the data in the DALLAS CALRAM
chip that
I was removing) the CALNUM went from 1 to something like 34 or
something. I
don't recall. It seems that if you have the appropriate software to
run the
complete calibration protocol, it only increments by '1' instead of by
all
the data points entered. Such software exists for the 3457A but I have
never seen it available 'on theBay'. I suspect Agilent would have that
software and equipment to do that calibration and, thus, an Agilent
calibration may only increment the CALNUM? by 1.

When getting an Agilent calibration of the 3458A, you get 'As
Received' and
'As Completed' data. Very helpful to me in that the only two points my
'House CAL' of the one 3458A failed were the two 'midrange' AC Voltage
values. All else 'PASSED'.

I agree with Dave. If it's HP/Agilent, I prefer Agilent to do the CAL.
If
it's Solartron, I prefer AMETEK (Solartron), etc.

I believe that having some 'basic' professionally calibrated
instruments
(DMM's, Noise Sources, Power Sensors, Frequency Standards (unless you
have a
GPSDO, CS Standard, etc.) etc.), that you can then use as 'transfer
standards' to do your own 'in house' calibration of other instruments,
is
very important if you want to set up a reliable workshop.

Of course, you will also need a 'stable' source of the various
signals
that
you will use to be 'measured' by the various 'DUT's', such as
resistance,
voltage, current, frequency, etc. The 3458A is relatively easy to
calibrate, requiring only 10.000000 VDC, 10000.000 ohms, and some AC
voltage
at various frequencies, IIRC. I have never CAL'd a 3457A but the
3478A
is a
multi-step process.

This whole thing can become very 'addictive'. Be careful.

Joe

-----Original Message-----
From: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
[mailto:hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>] On Behalf Of David
Kirkby
Sent: Saturday, December 29, 2012 5:53 PM
To: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
Subject: Re: [hp_agilent_equipment] 3457a on the way

On 29 December 2012 20:01, Jeff Machesky <jeff@...
<mailto:jeff%40codebest.com>
<mailto:jeff%40codebest.com>
<mailto:jeff%40codebest.com>
<mailto:jeff%40codebest.com>
<mailto:jeff%40codebest.com>
<mailto:jeff%40codebest.com> > wrote:
Thanks Dave, I actually have watched those videos. Bit drawn out
like
most of his videos..but still good. Too much detail is not always a
bad
thing.
I thought the bit showing the noise on the DVM was a bit silly when it
was connected to a DC power supply.

As for the 3457A, if it works I plan on getting it calibrated by
Agilent
within the year. From what I understand it's about a $200 US
investment.
The meter was last calibrated in '98, so I'll be curious to see how
accurate it is when I get it.
I think the calibration service you chose might dictate whether you
get data about the condition when sent.

When I send mine in for cal, I'd like to know what was out and by how
much. But I'm not going to pay extra for a calibration service that
provides that. As far as I'm concerned, if Agilent calibrate it, then
it is OK. For me personally, it makes no difference whatsoever if it
has ISO, NIST or whatever calibration. But I'd prefer Agilent to
someone else.

I have calibration certificate here for an Agilent VNA calibration
kit. It was done by a calibration house in the USA. But from what I
can gather from reading the documentation, the equipment to calibrate
them is not available commerically. So it makes me wonder how a lab
can calibrate a cal kit, when the equipment to do it can't be bought.

I suspect there is a fairly cosy realationship between some test
equipment dealers and calibration facilities.

I plan on purchasing some
voltage references from the well known site as a basic test of the
3457A. I may even calibrate it based on those references if it's way
out
and later getting it NIST traceable calibrated.
I don't know if there are pots in there you can adjust with a trimmer,
or if it is all done electronically. You might find it is impossible
to calibrate yourself.

I never had any reason to look inside mine.

Dave


------------------------------------

Yahoo! Groups Links





--------------------------------------------------------------------------------

No virus found in this message.
Checked by AVG - www.avg.com <>
Version: 10.0.1430 / Virus Database: 2637/5500 - Release Date: 12/31/12

Join HP-Agilent-Keysight-equipment@groups.io to automatically receive all group messages.