Steve,
Thanks for the data. Mine is also 2703A prefix with REV?:6,0 and CALNUM?:98. Not a multiple of 34. Would be interesting to see what the CALNUM increments by after an Agilent CAL.
I, too, need to look at the battery condition. I have not looked at the manual regarding replacing the battery. Has anyone done that without losing the CAL Constants?
Joe
toggle quoted message
Show quoted text
-----Original Message----- From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...] On Behalf Of Steve Sent: Sunday, December 30, 2012 4:14 PM To: hp_agilent_equipment@... Subject: Re: [hp_agilent_equipment] 3457a on the way It would be interesting to see if a newer rev is out there. My 3457A is s.n. prefix 2703, with rev 6,0 and option 0, CALNUM=34. I wonder if that's a default number for anything less than a full cal at Agilent? The last calibration was at least 6 years ago and performed by what was then Boeing Military Airplane Company's metrology lab. I need to open it up and check the battery condition. Steve On Dec 29, 2012, at 8:08 PM, "J. L. Trantham" <jltran@... <mailto:jltran%40att.net> > wrote: If the 'SELF TEST OK' message appears, there is no need to make any 'adjustments'. Just do the 'front panel CAL' if needed.
As I said, I would check it out, assume it is the best instrument in your collection, send it to Agilent for CAL and see what you get.
I would appreciate knowing what 'REV?' and 'OPT?' says when you get a chance. 'CALNUM?' would be interesting as well.
In the 3458A, the firmware is in an EPROM (6 EPROM's for the older units) and can be removed, a socket placed, and easily upgraded by purchasing the latest pre-programmed EPROM (or EPROM's for the older units) from Agilent. The only problem is they have a $50 minimum for this $18 part for the later units.
My wife thinks I am going to appear on an episode of 'Hoarders'.
Joe
-----Original Message----- From: hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com> [mailto:hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com> ] On Behalf Of Jeff Machesky Sent: Saturday, December 29, 2012 8:01 PM To: hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com> Subject: Re: [hp_agilent_equipment] 3457a on the way
Wow, lots of replies all of a sudden. I'm already prepping the wife for the $200 + price tag of calibration. Funny how when I was single I would have about $800 in cash in my wallet at all times and now I beg for 20 bucks, Hmm. Sad part is I make about 4 times the money. In any event I've not received to much feedback on the "Self Test OK" message the seller had posted. Any comments? I'm too much of a skeptic when it comes to eBay purchases. It's just a convenient place to purchase such goods. Any feedback would be appreciated as to possible pitfalls regarding this device. I like to prep for issues rather then build myself up for failure. What do they say...it's better to be pleasantly surprised then let down.
Thanks,
Jeff
On 12/29/2012 5:40 PM, J. L. Trantham wrote:
According to the manual, there are only two 'adjustments' that can be
made on a 3457A, Input Offset Amplifier adjustment and AC Converter Frequency Response, both needed only if there is a 'HARDWARE ERR' failure message after 'TEST' is selected and then only if it is a specific 'AUXERR' or 16 or 256 is seen. Otherwise, all the calibrations are done from the front
panel with specific inputs from the front panel.
The CALNUM? is incremented by 'several digits' with a 'complete calibration', one for each calibration point entered, per the manual. Interestingly, when I sent my two 3458A's to Agilent for calibration,
the CALNUM incremented by only 1. However, when I calibrated one of them before sending it to Agilent, (since I lost the data in the DALLAS CALRAM chip that I was removing) the CALNUM went from 1 to something like 34 or something. I don't recall. It seems that if you have the appropriate software to run the complete calibration protocol, it only increments by '1' instead of by
all the data points entered. Such software exists for the 3457A but I have never seen it available 'on theBay'. I suspect Agilent would have that software and equipment to do that calibration and, thus, an Agilent calibration may only increment the CALNUM? by 1.
When getting an Agilent calibration of the 3458A, you get 'As Received' and 'As Completed' data. Very helpful to me in that the only two points my 'House CAL' of the one 3458A failed were the two 'midrange' AC Voltage values. All else 'PASSED'.
I agree with Dave. If it's HP/Agilent, I prefer Agilent to do the CAL.
If it's Solartron, I prefer AMETEK (Solartron), etc.
I believe that having some 'basic' professionally calibrated instruments (DMM's, Noise Sources, Power Sensors, Frequency Standards (unless you have a GPSDO, CS Standard, etc.) etc.), that you can then use as 'transfer standards' to do your own 'in house' calibration of other instruments,
is very important if you want to set up a reliable workshop.
Of course, you will also need a 'stable' source of the various signals that you will use to be 'measured' by the various 'DUT's', such as
resistance, voltage, current, frequency, etc. The 3458A is relatively easy to calibrate, requiring only 10.000000 VDC, 10000.000 ohms, and some AC voltage at various frequencies, IIRC. I have never CAL'd a 3457A but the 3478A is a multi-step process.
This whole thing can become very 'addictive'. Be careful.
Joe
-----Original Message----- From: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com> <mailto:hp_agilent_equipment%40yahoogroups.com> [mailto:hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com> <mailto:hp_agilent_equipment%40yahoogroups.com>] On Behalf Of David
Kirkby Sent: Saturday, December 29, 2012 5:53 PM To: hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com> <mailto:hp_agilent_equipment%40yahoogroups.com> Subject: Re: [hp_agilent_equipment] 3457a on the way
On 29 December 2012 20:01, Jeff Machesky <jeff@...
<mailto:jeff%40codebest.com> <mailto:jeff%40codebest.com> <mailto:jeff%40codebest.com> > wrote:
Thanks Dave, I actually have watched those videos. Bit drawn out like most of his videos..but still good. Too much detail is not always a
bad thing. I thought the bit showing the noise on the DVM was a bit silly when it was connected to a DC power supply.
As for the 3457A, if it works I plan on getting it calibrated by
Agilent within the year. From what I understand it's about a $200 US
investment. The meter was last calibrated in '98, so I'll be curious to see how accurate it is when I get it. I think the calibration service you chose might dictate whether you get data about the condition when sent.
When I send mine in for cal, I'd like to know what was out and by how much. But I'm not going to pay extra for a calibration service that provides that. As far as I'm concerned, if Agilent calibrate it, then it is OK. For me personally, it makes no difference whatsoever if it has ISO, NIST or whatever calibration. But I'd prefer Agilent to someone else.
I have calibration certificate here for an Agilent VNA calibration kit. It was done by a calibration house in the USA. But from what I can gather from reading the documentation, the equipment to calibrate them is not available commerically. So it makes me wonder how a lab can calibrate a cal kit, when the equipment to do it can't be bought.
I suspect there is a fairly cosy realationship between some test equipment dealers and calibration facilities.
I plan on purchasing some voltage references from the well known site as a basic test of the 3457A. I may even calibrate it based on those references if it's way
out and later getting it NIST traceable calibrated. I don't know if there are pots in there you can adjust with a trimmer, or if it is all done electronically. You might find it is impossible to calibrate yourself.
I never had any reason to look inside mine.
Dave
------------------------------------
Yahoo! Groups Links
|