开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

Re: HP 3403C True RMS Voltmeter display fault


 

Ok, tgat's good to know.

DaveD
KC0WJN


On Tue, Mar 4, 2025 at 13:12 Ed Marciniak via <ed=[email protected]> wrote:
The OSCON type capacitors aren’t available in higher voltage ratings.

From: [email protected] <[email protected]> on behalf of Dave Daniel via <kc0wjn=[email protected]>
Sent: Tuesday, March 4, 2025 11:30:05 AM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] HP 3403C True RMS Voltmeter display fault
?
I'll add that since, in general, modern caps are smaller than vintage caps, make sure the the voltage rating of the new cap provides sufficient margin over the voltage rail. For instance, Tektronix often used caps rated for +15v on +12v rails. In all the cases where I have recapped equipment, a modern +25v cap was smaller in physical size than the original +15v caps.

DaveD
KC0WJN


On Tue, Mar 4, 2025 at 12:14 ed breya via <eb=[email protected]> wrote:
Arnaud, you can replace Ta caps with alternatives in most applications, so there's no need to spend too much for an original type. Usually ones that burn out are in supply bypassing spots, which are not critical. The main issue then is space available for new parts, since the common fix is to go with a much larger capacitance (and physical size) aluminum electrolytic type to get ESR comparable to the Ta ones. I'd say put in an Al cap of at least 2-10 times as much C as the original Ta, with enough V rating for the supply voltage. Or, use the biggest C that can physically fit into the space available. You can look at various brands and families of caps to choose the lowest ESR you can find, but they won't be dramatically different.
?
Another option is to go with an OSCON organic dielectric type Al cap. These are built for low ESR and can beat Ta in this respect, so fitting is easier. The problem here is that (I think) OSCONs are only available in radial mount style (can with leads out the bottom), so trickier to fit in axial-lead spots. If axials are available, then no problem.
?
Remember this all only applies to supply bypassing spots. Sometimes Ta is needed for particularly low DC leakage and better stability like in timing circuits. Ones in these kinds of spots hardly ever fail because they're not stressed. Supply bypassing ones fail a lot, but value stability and leakage hardly matter here, so bigger Al types or OSCONs (which have the worst leakage) work just fine to get decent ESR.
?
Ed
?
?

Join [email protected] to automatically receive all group messages.