开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

Re: HP 54512B (pair) repair


 

Peter,
Fortunately, the PS troubleshooting section (p.6-21) has some good directions on how to check the PS once detached from the modules it supplies. It involves loading only the 5.2V output (with a 2ohms/25W resistor). That should provide a determination on whether the -12V rail is loaded down by a module (figuring out what exactly loads it is another issue, given the lack of schematics, and per block diagram I can only tell the -12V goes to the main board), or there's a fault inside the PS itself.?
There's also a fair section on assemblies removal. Will come in handy, given the mainboard needs to come off. Still not the best SM I've seen by HP standards.?
Radu.?


On Thu, Feb 27, 2025 at 4:34?PM Peter Gottlieb via <hpnpilot=[email protected]> wrote:
You are seeing the tabs which are spot welded to the cell, it's not a socket,
the cell is soldered to the board.? Well you might be able to pry but risk
damage to the board or traces.? At this point replace the thing.? Best is to
desolder it from the bottom of the board.? Definitely figure out what's up with
that power supply though.

Peter


On 2/27/2025 2:32 PM, Radu Bogdan Dicher via wrote:
> I lifted today the CRT card, just enough to have some access to the battery,
> but I'm still unsure how it's supposed to come off. I tried prying the top
> spring "leaf," but it seems attached (possibly soldered) to the battery - or
> is that a top lid over the battery? And then there's a piece of foam
> underneath, possibly double-sided sticking. I'd rather not remove the whole
> mainboard just to replace this, but I don't think I can try many more prying
> and pulling before?I risk something more serious due to the lack of access.
>
> Hoping someone having done this before?recognizes the type of battery and its
> socket from the enclosed pictures?and has some good pointers. I am not seeing
> any other battery on this board - I was expecting?a bigger, cylindrical?type,
> but I think this must be it for the 54512B.
>
> Thanks a bunch in advance.
> Radu.
>
> On Wed, Feb 26, 2025 at 7:08?AM Radu Bogdan Dicher via
> <> <vondicher=[email protected]> wrote:
>
>? ? ?Thanks, Jim. I still have to perform that test where I play The Cure from
>? ? ?my garage with the door open, which?I'm sure you'd hear from your bench!!
>? ? ?;))))
>
>? ? ?I tried?to pry off the battery this morning without further disassembly,
>? ? ?but it doesn't want to come off that easily. There seems to be a blob of
>? ? ?solder - has anyone seen that before? - on top of the holder, and some
>? ? ?sort of?foam?tape below it. Maybe I need to apply force a little bit
>? ? ?differently.
>
>? ? ?Unless there's some secret?regarding how to remove the battery, I'll have
>? ? ?to at least remove at least the CRT module to gain?better access to it. I
>? ? ?really think access to a battery - even one guaranteed?a decade - should
>? ? ?be easier than this. So this should be possible later in the week when I
>? ? ?have more time.
>? ? ?Radu.
>
>? ? ?On Wed, Feb 26, 2025 at 5:38?AM Jim Ford via <>
>? ? ?<james.ford=[email protected]> wrote:
>
>? ? ? ? ?Speaking of CR2032 coin cells, don’t spend $$ at a local store; order
>? ? ? ? ?from Digital-Key for about $0.50 in quantity.? I usually buy 50 pieces
>? ? ? ? ?at a time.? But then I have a keychain LED flashlight (torch for the
>? ? ? ? ?Brits) that my kids bought me some years ago (best gift ever!) that
>? ? ? ? ?takes 2 cells and a 3-LED baseball cap that I wear when walking my
>? ? ? ? ?dogs at night that takes 4.? So I go through a lot of them.? Mouser
>? ? ? ? ?might have them as well.? I’ve not ordered from D-K for a while
>? ? ? ? ?because they buried the big product table, and it seems like something
>? ? ? ? ?else they did got me PO’d at them.? HTH.
>
>? ? ? ? ?Jim Ford
>? ? ? ? ?Laguna Hills, California, USA
>? ? ? ? ?> On Feb 25, 2025, at 7:46?PM, Peter Gottlieb via
>? ? ? ? ?<> <hpnpilot=[email protected]> wrote:
>? ? ? ? ?>
>? ? ? ? ?> ?CR followed by numbers indicates the physical size, CR2032 etc.? I
>? ? ? ? ?would say if you're already in there, replace them as even if they are
>? ? ? ? ?good they won't last much longer.? I always replace memory hold up
>? ? ? ? ?cells in anything that crosses my bench.? Low hanging fruit.
>? ? ? ? ?>
>? ? ? ? ?> The power supply might be low because of an overload caused by a
>? ? ? ? ?shorted capacitor.? Perhaps try unplugging the supply connections. I
>? ? ? ? ?can't remember if that model's supply will run correctly without
>? ? ? ? ?load.? But if you do go inside it, check the capacitors for
>? ? ? ? ?capacitance, leaks and ESR.
>? ? ? ? ?>
>? ? ? ? ?> Peter
>? ? ? ? ?>
>? ? ? ? ?>
>? ? ? ? ?>> On 2/25/2025 9:35 PM, Radu Bogdan Dicher via
>? ? ? ? ?<> wrote:
>? ? ? ? ?>> I've gotten a bit of a chance tonight to check a couple of things
>? ? ? ? ?with one of the units:
>? ? ? ? ?>>
>? ? ? ? ?>> *Unit #2 *(S/N: 3214A00704: consecutive units!)
>? ? ? ? ?>>
>? ? ? ? ?>>? * Voltages on rails:
>? ? ? ? ?>>? ? ? o +11.924V
>? ? ? ? ?>>? ? ? o +12.050V
>? ? ? ? ?>>? ? ? o -3.27V(fluctuating somewhat; no significant ripple on it)
>? ? ? ? ?>>? ? ? o -5.2289V
>? ? ? ? ?>>? ? ? o 3.5285V
>? ? ? ? ?>>? ? ? o 5.2067V
>? ? ? ? ?>>? * There's a coin battery on the main board, which I didn't remove
>? ? ? ? ?yet (I'd
>? ? ? ? ?>>? ? rather measure its voltage before I do, in case it's still good
>? ? ? ? ?and holds
>? ? ? ? ?>>? ? the memory). It's a Panasonic and I can only read "CR..." from
>? ? ? ? ?it. Would
>? ? ? ? ?>>? ? there potentially be more harm done by removing it, than simply
>? ? ? ? ?losing the
>? ? ? ? ?>>? ? cal constants, in case it's still good for some reason? I'm
>? ? ? ? ?thinking of
>? ? ? ? ?>>? ? biting the bullet and removing it, replacing it, and then seeing
>? ? ? ? ?where
>? ? ? ? ?>>? ? that takes me.
>? ? ? ? ?>>
>? ? ? ? ?>> So this second unit has more issues than just a dead battery (in
>? ? ? ? ?case the latter checks out). I'll have to remove the power supply
>? ? ? ? ?module and see what could be wrong with the -12V rail. I'm
>? ? ? ? ?provisionally blaming the unruly -12V rail for the additional error
>? ? ? ? ?this second unit has, which the first doesn't. So my money's on that
>? ? ? ? ?the first one has a good -12V rail.
>? ? ? ? ?>> Radu.
>? ? ? ? ?>>
>? ? ? ? ?>> On Tue, Feb 25, 2025 at 7:14?AM Radu Bogdan Dicher via
>? ? ? ? ?<> <> <vondicher=[email protected]>
>? ? ? ? ?wrote:
>? ? ? ? ?>>
>? ? ? ? ?>>? ? Awesome, thank you, Peter.
>? ? ? ? ?>>
>? ? ? ? ?>>? ? I wonder if anyone has a fresh recollection of what battery or maybe
>? ? ? ? ?>>? ? battery-backed RAM would be suspect and possibly needed so I'd start
>? ? ? ? ?>>? ? working on an order to have the parts ready by the time I carve
>? ? ? ? ?out some
>? ? ? ? ?>>? ? time to open these. From what I know about these generation scopes,
>? ? ? ? ?>>? ? replacing the battery is probably a good idea anyway. The cal
>? ? ? ? ?constants
>? ? ? ? ?>>? ? seem to be gone, so nothing to lose there.
>? ? ? ? ?>>
>? ? ? ? ?>>? ? I wasn't able to find this information (particularly info on the
>? ? ? ? ?battery)
>? ? ? ? ?>>? ? in the manuals, which I find odd.
>? ? ? ? ?>>? ? Radu.
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>? ? On Tue, Feb 25, 2025 at 5:26?AM Peter Gottlieb via
>? ? ? ? ?<>
>? ? ? ? ?>>? ? <> <hpnpilot=[email protected]> wrote:
>? ? ? ? ?>>
>? ? ? ? ?>>? ? ? ? Seems like you are on the right track. Many HP scopes of
>? ? ? ? ?that era
>? ? ? ? ?>>? ? ? ? lose their
>? ? ? ? ?>>? ? ? ? minds if the memory battery goes dead.
>? ? ? ? ?>>
>? ? ? ? ?>>? ? ? ? Peter
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>? ? ? ? On 2/24/2025 11:09 PM, Radu Bogdan Dicher via
>? ? ? ? ?<>
>? ? ? ? ?>>? ? ? ? <> wrote:
>? ? ? ? ?>>? ? ? ? > Hi all,
>? ? ? ? ?>>? ? ? ? > Over the weekend I've got a couple of 54512Bs at my desk
>? ? ? ? ?to seek to
>? ? ? ? ?>>? ? ? ? bring back
>? ? ? ? ?>>? ? ? ? > to full functionality. My main scope is a 54522A, and I
>? ? ? ? ?assume they
>? ? ? ? ?>>? ? ? ? have some
>? ? ? ? ?>>? ? ? ? > commonalities, but I've yet to open mine (it came to me fully
>? ? ? ? ?>>? ? ? ? refurbished and
>? ? ? ? ?>>? ? ? ? > never needed to open the case). I'm not very familiar with
>? ? ? ? ?this
>? ? ? ? ?>>? ? ? ? other series,
>? ? ? ? ?>>? ? ? ? > and hope on some good input and recommendations.
>? ? ? ? ?>>? ? ? ? > The documentation is not as good as others I've seen from HP.
>? ? ? ? ?>>? ? ? ? Specifically for
>? ? ? ? ?>>? ? ? ? > this unit (54512B), Keysight only seems to have a very
>? ? ? ? ?skinny (by HP
>? ? ? ? ?>>? ? ? ? > standards) Service Manual which can't honestly be called
>? ? ? ? ?>>? ? ? ? "module-level Service
>? ? ? ? ?>>? ? ? ? > Manual," as it doesn't really into much detail on the
>? ? ? ? ?different
>? ? ? ? ?>>? ? ? ? modules. Then
>? ? ? ? ?>>? ? ? ? > there's a "Front Panel Reference," which is an odd other
>? ? ? ? ?way of
>? ? ? ? ?>>? ? ? ? naming what I
>? ? ? ? ?>>? ? ? ? > think is the Operating Manual. A couple other manuals have
>? ? ? ? ?to do with
>? ? ? ? ?>>? ? ? ? > programming. That's it.
>? ? ? ? ?>>? ? ? ? > Sometimes in a series you have to look for the main unit,
>? ? ? ? ?but in
>? ? ? ? ?>>? ? ? ? this case I'm
>? ? ? ? ?>>? ? ? ? > not sure that's the 54501A, as the catalog spec sheet seems to
>? ? ? ? ?>>? ? ? ? suggest, or the
>? ? ? ? ?>>? ? ? ? > 54505B, as per the SM. Maybe someone very familiar with
>? ? ? ? ?the series
>? ? ? ? ?>>? ? ? ? would know.
>? ? ? ? ?>>? ? ? ? > Artek doesn't have manuals for the 54512B at all, but they
>? ? ? ? ?do for
>? ? ? ? ?>>? ? ? ? the 54501A,
>? ? ? ? ?>>? ? ? ? > and if applicable, I'd get the series of six manuals they
>? ? ? ? ?sell as a
>? ? ? ? ?>>? ? ? ? bundle. If
>? ? ? ? ?>>? ? ? ? > no go, I'll be a bit in a bind with fixing this given what
>? ? ? ? ?I have at
>? ? ? ? ?>>? ? ? ? hand. I
>? ? ? ? ?>>? ? ? ? > am aware HP was not longer offering component SMs by this
>? ? ? ? ?point,
>? ? ? ? ?>>? ? ? ? though I
>? ? ? ? ?>>? ? ? ? > don't know for a fact on this specific series and scope
>? ? ? ? ?(but I do
>? ? ? ? ?>>? ? ? ? regarding
>? ? ? ? ?>>? ? ? ? > the better?/later? 54522A).
>? ? ? ? ?>>? ? ? ? > Is this really capable of FFT to 500MHz, even if the scope
>? ? ? ? ?is rated
>? ? ? ? ?>>? ? ? ? to 300MHz?
>? ? ? ? ?>>? ? ? ? > I can't make sense of that.
>? ? ? ? ?>>? ? ? ? > Now, on to the specific issues with the units.
>? ? ? ? ?>>? ? ? ? > *Unit #1 *(S/N: 3214A00705)
>? ? ? ? ?>>? ? ? ? >
>? ? ? ? ?>>? ? ? ? >? ?* Upon starting it, it reports something like "TRYING TO
>? ? ? ? ?START
>? ? ? ? ?>>? ? ? ? REALTIME
>? ? ? ? ?>>? ? ? ? >? ? ?CLOCK," which is unsuccessful after a time countdown
>? ? ? ? ?from I
>? ? ? ? ?>>? ? ? ? think 60:
>? ? ? ? ?>>? ? ? ? >? ? ?"REALTIME CLOCK FAILED TO START - DISABLED."
>? ? ? ? ?>>? ? ? ? >? ?* Once I get it to display the default screen and output
>? ? ? ? ?a trace,
>? ? ? ? ?>>? ? ? ? it stated
>? ? ? ? ?>>? ? ? ? >? ? ?"Default cal loaded. Re-cal instrument."
>? ? ? ? ?>>? ? ? ? >? ?* Self Test passes all.
>? ? ? ? ?>>? ? ? ? >? ?* I'm also seeing a solid and apparently accurate 5V
>? ? ? ? ?from the DC
>? ? ? ? ?>>? ? ? ? Cal output
>? ? ? ? ?>>? ? ? ? >? ? ?in the back. I haven't tried the AC Cal output.
>? ? ? ? ?>>? ? ? ? >? ?* Note this one had the Cal switch in the back in the
>? ? ? ? ?UNPROTECTED
>? ? ? ? ?>>? ? ? ? position
>? ? ? ? ?>>? ? ? ? >? ? ?when I got it.
>? ? ? ? ?>>? ? ? ? >? ?* I'm able to display the square wave test signal from
>? ? ? ? ?the front
>? ? ? ? ?>>? ? ? ? panel.
>? ? ? ? ?>>? ? ? ? >
>? ? ? ? ?>>? ? ? ? > *Unit #2 *(S/N: 3214A00704: consecutive units!)
>? ? ? ? ?>>? ? ? ? >
>? ? ? ? ?>>? ? ? ? >? ?* Upon staring it, it stated "CALIBRATION FACTOR
>? ? ? ? ?CHECKSUM ERROR.
>? ? ? ? ?>>? ? ? ? MOVE REAR
>? ? ? ? ?>>? ? ? ? >? ? ?PANEL CAL SWITCH TO UNPROTECTED. POWER INSTRUMENT OFF
>? ? ? ? ?AND THEN
>? ? ? ? ?>>? ? ? ? ON AGAIN.
>? ? ? ? ?>>? ? ? ? >? ? ?RE-CAL INSTRUMENT."
>? ? ? ? ?>>? ? ? ? >? ?* It also says "Trying to start realtime clock (09290)."
>? ? ? ? ?Maybe
>? ? ? ? ?>>? ? ? ? that's an
>? ? ? ? ?>>? ? ? ? >? ? ?error code?
>? ? ? ? ?>>? ? ? ? >? ?* Upon switching the rear Cal switch to UNPROTECTED, it
>? ? ? ? ?boots
>? ? ? ? ?>>? ? ? ? successfully,
>? ? ? ? ?>>? ? ? ? >? ? ?goes to a default measurement screen, but gives me a
>? ? ? ? ?warning "Failed
>? ? ? ? ?>>? ? ? ? >? ? ?interpolar cal." The trace seems to jump around on Y a
>? ? ? ? ?bit.
>? ? ? ? ?>>? ? ? ? >? ?* Self Test passes all /but/ Timebase. It also displays
>? ? ? ? ?the same
>? ? ? ? ?>>? ? ? ? error as
>? ? ? ? ?>>? ? ? ? >? ? ?default screen ("Failed interpolar cal").
>? ? ? ? ?>>? ? ? ? >? ?* 5V out at DC Cal is good.
>? ? ? ? ?>>? ? ? ? >? ?* This one had the Cal switch in UNPROTECTED position,
>? ? ? ? ?so maybe
>? ? ? ? ?>>? ? ? ? that's why
>? ? ? ? ?>>? ? ? ? >? ? ?it booted successfully.
>? ? ? ? ?>>? ? ? ? >? ?* I'm able to display the square wave test signal from
>? ? ? ? ?the front
>? ? ? ? ?>>? ? ? ? panel,
>? ? ? ? ?>>? ? ? ? >? ? ?though it seems to be offset vertically a bit.
>? ? ? ? ?>>? ? ? ? >
>? ? ? ? ?>>? ? ? ? > To me, this seems to tell me the battery holding the cal
>? ? ? ? ?constants
>? ? ? ? ?>>? ? ? ? is done -
>? ? ? ? ?>>? ? ? ? > going by what I know about my 54522A. It's also possible
>? ? ? ? ?some Dallas
>? ? ? ? ?>>? ? ? ? RAMs are
>? ? ? ? ?>>? ? ? ? > out of battery - though I don't know if this uses those for
>? ? ? ? ?>>? ? ? ? non-volatile
>? ? ? ? ?>>? ? ? ? > memory. I have not yet opened the units, so I've no idea
>? ? ? ? ?where the
>? ? ? ? ?>>? ? ? ? rails are,
>? ? ? ? ?>>? ? ? ? > if any caps have leaked, etc. But maybe the information
>? ? ? ? ?above gives
>? ? ? ? ?>>? ? ? ? some clues
>? ? ? ? ?>>? ? ? ? > on what's going on, enough to start thinking on it. It
>? ? ? ? ?also seems
>? ? ? ? ?>>? ? ? ? the issues
>? ? ? ? ?>>? ? ? ? > are shared very similarly, so maybe that's making the
>? ? ? ? ?repair more
>? ? ? ? ?>>? ? ? ? streamlined.
>? ? ? ? ?>>? ? ? ? > Looking forward to your input.
>? ? ? ? ?>>? ? ? ? > Thank you,
>? ? ? ? ?>>? ? ? ? > Radu.
>? ? ? ? ?>>? ? ? ? >
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>>
>? ? ? ? ?>
>? ? ? ? ?>
>? ? ? ? ?>
>? ? ? ? ?>
>? ? ? ? ?>
>? ? ? ? ?>
>
>
>
>
>
>
>






Join [email protected] to automatically receive all group messages.