¿ªÔÆÌåÓý

Re: Agilent 34461A schenmatic


 

¿ªÔÆÌåÓý

Do you have the full schematic for the 34410a?


From: [email protected] <[email protected]> on behalf of Yves Tardif <yves_tardif@...>
Sent: Monday, October 7, 2024 12:37 PM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] Agilent 34461A schenmatic
?

?

As specified in the last email, I don't have the part numbers unfortunately, only the marking codes or the Keysight part numbers. Also, I can't find these part numbers on the Keysight website.

?

If I were in your situation, I would first look for a 1.3MHz 8 pin switcher with the same package.

And from the circuit and the coil value, I would try to see if the pin out seems compatible, and if the installed coil would be compliant with a 1.3MHz operating frequency.

?

A friend of mine probably has a 34461A that he could lend me, but don't expect to have more details on the part numbers, but I could take the waveforms at the switcher pins.

I am not absolutely sure that the switcher of the 34461A and 34465A are identical because I do not have a detailed photo of this section for the 34461A, but the coil value is the same.

?

?

?

De?: [email protected] <[email protected]> De la part de Keith Harvey
·¡²Ô±¹´Ç²â¨¦?: 7 octobre 2024 14:06
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Agilent 34461A schenmatic

?

Hi Yves,

?

Thank you for the insight.

?

If you have the Agilent part numbers for the components I listed, that might prove helpful.

?

Regards,

Keith Harvey


From: [email protected] <[email protected]> on behalf of Yves Tardif <yves_tardif@...>
Sent: Monday, October 7, 2024 11:58 AM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] Agilent 34461A schenmatic

?

Hi Keith,

?

I checked my 34410A and my 34465A, and unfortunately, I don't have a 34461A.

However, they all have the same function to perform, that is to say from +10V, provide a voltage of +3.3V, +1.2V and possibly +2.5V.

A switcher lowers a DC voltage from about +10V to +3.3V, the frequency of the switcher is about 1.3MHz for both instruments.

The 34461A has the same part numbers as the 34465A, there is a very good chance that the 34465A and 34461A have the same circuit for the section concerned.

?

I don't know if this can help you, but you have an idea of ??the damaged circuit.

I don't have the part numbers unfortunately, only the marking codes or the Keysight part numbers.

?

I repaired a 34461A a few months ago, the +3.3V switch was not damaged, but there was a short circuit on the +3.3V line caused by a faulty capacitor.

?

Waveform measured at output switcher connected to coil.

?

Yves

Join [email protected] to automatically receive all group messages.