Re: HP 141T Power Supply Help Needed


 

¿ªÔÆÌåÓý

That display is normal for an 8553B in FULL SCAN mode.? The dip is the marker that indicates the center frequency.

Dave Wise


From: [email protected] <[email protected]> on behalf of paulswed <paulswedb@...>
Sent: Thursday, September 19, 2024 4:47 PM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed
?
Look at that. The only issue I see is that the display is upside down.
I guess you need to flip it over. Humor.
By the looks of it you have some signal. But your IF BW may be narrow.
Typically with a wider bandwidth you would see noise in the floor.
Or its telling you gain is set low something silly like that. Front panel stuff.
The 141T is a nice spectrum analyzer. It was the first one I used in the Navy circa 1975. Liked it a lot. But never owned one. Have newer units.
Paul
WB8TSL