¿ªÔÆÌåÓý

Re: HP 141T Power Supply Help Needed


 

¿ªÔÆÌåÓý

Bill,

?

What is the voltage at the collector of Q4?? Also if you have a 10 A ammeter remove F4 and measure the current.? I sounds like there is a short somewhere.? Keep the fuse in the circuit to protect the ammeter if at all possible. ?Also is R49 getting hot?? Maybe a shorted C9?

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Monday, September 2, 2024 10:14 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Yep all purple/white disconnected on the left.? Currently almost no voltage at output of q4 emitter, -0.02.? ? It's also getting hot.? ?Wonder if it's toast bc I lost my -2.5v since.? ??

?

On Mon, Sep 2, 2024, 11:10?AM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill.

?

You are making progress!? Now we have to find out where the -12.6 current is going.? Especially since the driver is getting hot.? It is funny that the current limiter is showing 0V.? I think some more ohm meter checks are? in order.? I assume the -12.6 to the A5 module is disconnected.? In fact I suggest pulling all of the white / violet wires from the A2 regulator board.? What voltage is there at the emitter of the pass device Q4?? Let me know¡­

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Sunday, September 1, 2024 6:45 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Spent some time today checking through the rest of the line - basic checks on passives look ok. Transistors look ok.? Noticed the driver getting hot though.? ?Not making any progress, despite trying to figure it out.? ?

Current measurements:?

Q9..?

C = 16.27v (was 22v)

B = 2.6v? (was 23.5v)
E = 1.9v (was -17v)

Q10..
C = 2.6v? (was 23.5v)
b = 0v again

E = 0v

Q11..?
C=? ~2.2v (was 23.5v)
B = -0.2v, fluctuates? (was -.64)
E = -1v, fluctuates (was 0)



In other news, I believe I have located the short on the A5 board.? ?Stray lead clipping in a very tough to see location...? ?:\? ? ??

?

On Sun, Sep 1, 2024 at 4:55?PM Bill Berzinskas via <bberzinskas=[email protected]> wrote:

tiny update to the above, replaced?+100v driver and that helped immensely!
+100, -100,?+248 all work, adjust and appear happy!? ?

I gained .5v on my -12.6 line, but its still sitting at -3v.? ?

?

On Sun, Sep 1, 2024 at 4:20?PM Bill Berzinskas <bberzinskas@...> wrote:

Getting closer, seeing some signs of life finally.? ? Not all the way but progress is progress.??

I've found that CR16 and CR17 were not good, and the pass trans was also not happy anymore.? ?I've replaced all of that and am now getting a whopping -2.5v!? ?10 more to go.? :-).? Will check through resistors this evening.? ? ?I'm also seeing some off voltages in my other rails, will have to go back through the?+100 and -100 lines as well.? ? ?

Current status:?
+79
-82
+217
-2.5

The purple/white wire presenting the short was easy to find, its the one going to the board just in front of the power?supply. Looks like thats A5, so I'll start mulling over the schem?for that as well.? ?

? ?

?

On Sat, Aug 31, 2024 at 5:00?PM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

It sounds like you are on the right track!? So find out were the .04Ohm wire goes¡­

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Saturday, August 31, 2024 2:29 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Neat trick checking for shorts.?? As you expected probably, 0.04ohm.???? I have found one purple/white wire which when removed brings that up to 384 ohm.???

I'll be out of pocket for a day but will look at the supply one more time in isolated state and then trace that wire out.???

Thank you!

?

On Sat, Aug 31, 2024, 12:11?PM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

So you replaced the pass transistor and Q9 then it popped the 3A fuse F4.? And at that point the regulator worked but there must be a short on the -12.6V buss.? Please run an Ohm meter check on the -12.6V buss which is / are the white / violet wires from the A2 assembly.? Check R42, CR16 and Q10.? ?From my calculations the current limiter Q10 should start taking drive away from the driver Q9 at about 2.8A.? ?I take it that no plugins are installed.? The only place in the 141T that the -12.6 goes is / the? plugin connectors, the flood gun filament, and the trace align control on the front panel.? If there is a short (<4 Ohms) then you have to find which wire has that low impedance.? Let me know what you find.

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Saturday, August 31, 2024 8:07 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Those numbers just don't make any sense.??? I pulled all the transistors AGAIN, desoldered and cleaned.? Tested them all with a meter, repopulate.??? Back to 23.5v....???

?

On Fri, Aug 30, 2024, 6:22?PM Bill Berzinskas via <bberzinskas=[email protected]> wrote:

Replaced Q9, powered on and blew the fuse.? ?Replaced Fuse?+ Q9 again?+ pass transistor and no fuse pop.? ?
Not sure if better, worse or just different :-)?

Q9..?

C = 22v

B = 54v?!? (was 23.5v)
E = -17v.?

Q10..
C = 54v..? (was 23.5v)
b = 0.58v (up from 0)?

E = 0v

Q11..?
C= 54.7 (was 23.5v)
B = -.64
E = 0

?

On Fri, Aug 30, 2024 at 5:48?PM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

OK it seems that Q 9 has a base-emitter voltage of 38.5V which says that the base-emitter junction is open because a good NPN silicon transistor can only have a forward base-emitter voltage of 1V under extreme conditions.? Normally the Vbe is in the order of 0.7V. If the emitter or base connection of Q9 were open the emitter voltage will be dragged negative by R41.? That is not good for the pass transistor Q4 to see that much negative voltage because that will exceed the Vebo rating of the pass device which may degrade the Beta of the pass device.? Basically it will Zener the base emitter junction with that much reverse bias.? Typical Vebo is in the order of <10V which is in the absolute maximum ratings of the part.? A 2N3053 should work OK as the driver Q9.? Let me know how it goes.

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Friday, August 30, 2024 2:56 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

That does make sense,?thank you for that explanation..?

I do believe I've stated it backwards.? ? 22v appears on a 2n3053 that I've replaced at pin 3, which is actually the collector..? :facepalm:..?

?

Remeasure / restate Q9..?

C = 22v

B = 23.5v
E = -15v.?


? ??

?

On Fri, Aug 30, 2024 at 3:02?PM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

Lets back up a minute¡­ If there is +22V at the emitter of the driver Q9 then there must be +22V at the base of the series regulator Q4 and you say that there is less than 1V ?at the emitter of the pass transistor Q4 and also no voltage at the top of R42 (the current sense resistor) then there is 22V across the base emitter junction of? Q4 and that can only be a failed device or some bad wiring or even a cracked runner on the board. A good NPN transistor has no more than 1V Vbe.? ??Does that make sense to you?? Or is the voltage at the emitter of the driver Q9 a negative 22V?? If the voltage at the emitter of Q9 is negative then what is the voltage at the base of Q9?

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Friday, August 30, 2024 1:21 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Ok..? I had a few bad transistors in the mix, but the one that was in the unit checks out ok.? ? ?I tossed a small handful to alleviate future issues.? ?

I spent a little time simulating this and from what I'm seeing this almost has to be a problem w/ the voltage divider / potentiometer.? ?I can do a lot of "shady" things in the sim and none of them give me issues like I'm seeing.? ?Removing R43 in the sim, or the pullup R45 seem to skew the voltages like I'm seeing.? ? ?The sim shows that the trimpot is kind of what "kicks things off", and the fact that I'm only getting about -1v here seems to further hint towards R43.? ? I don't have a 22.1k or even a 22k on hand, so I'll probably get creative to see what happens.? ?Resistance measurements on the pot itself look ok, i'm really hoping to not have to yank that thing.??

?

On Fri, Aug 30, 2024 at 11:58?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:

I'll check these over lunch.? ?Interestingly, two of the original transistors in that spot cause the 3a fuse to blow.? ?Now sure if thats indicative of something else going on w/ a good transistor in the seat..? ? tbd.??

?

On Fri, Aug 30, 2024 at 11:43?AM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

22V on the emitter of the driver is very strange.? It seems that the pass transistor Q2 has an open base ¨C emitter junction and a shorted collector base junction to come up with those voltages.? That would explain the ¡°on output¡± situation.?? An Ohm meter should be sufficient to check the status of Q2.? If your meter has a diode scale, that would give you a quick indication of the status of Q2.?

?

I am thinking ?of designing an over voltage crowbar circuit to protect the -12.6 V supply from damaging sensitive circuits down stream of this supply if the regulator goes to an over voltage condition.? Some of the plugins appear to have a crowbar in just such a condition.? The filament of the flood gun is run from the -12.6V and in my case it went over voltage and it burned out the flood gun filament. ?

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Friday, August 30, 2024 10:17 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Thanks Michael,?
here's a recap what i'm seeing around the -12.6v line currently:?

?.65v at base of q11 sensor amp, .275v on collector.??

Driver has 22v on emitter, .27v on base and -.27v collector.?????

Not much happening on the current limiter.?? 0v base.?

?

On Fri, Aug 30, 2024 at 11:15?AM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

OK so you have the early version of the supply.? If you give me the voltages at the base of the driver transistors I can perhaps give you some more pointers.

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Thursday, August 29, 2024 4:34 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

correction, it did have the LAMP..? ??

?

On Thu, Aug 29, 2024 at 5:30?PM Bill Berzinskas <bberzinskas@...> wrote:

My unit is 1615a prefix, it did have the zener and it was measuring closer to 100v instead of the expected 83 or whatever.? It has since been replaced by a pair of diodes that bring it to 89v but I have had to replace that string at least once in my travels so far, so I will eventually look for a more permanent?solution as I've read about.? ?I can check that again this evening!? ?I stopped thinking about it after I got the?+100v supply up!

?

On Thu, Aug 29, 2024 at 4:50?PM Michael Bafaro via <m.bafaro=[email protected]> wrote:

Bill,

?

There are 2 versions of this supply.? There is an early version that has a neon tube as the reference for the +100V supply and a later version that uses a 9V temp compensated Zener that is floating.? That is the reference Zener is connected from the +100V rail and the base of Q4.? It may help to know which version you have.? From my schematic the +100 V is the master and the other supplies are referenced from it.?? The one I have has a serial number prefix 2101A and it appears to be made in the late 1981 or early 1982 time frame. This one has the floating 9V Zener.

?

Mike

?

From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Thursday, August 29, 2024 10:13 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed

?

Finished that up.? ?On the -100 I can adjust up to 96v, a little low but it is adjustable so I think I should come back to that later.??

On 248v I'm currently seeing 290v, and -12v I'm seeing -1.5v.? ? ?I reckon I could go after either of them given the dependencies.? ?

Will grab some measurements in between or after work today.? ?

?

On Thu, Aug 29, 2024 at 10:52?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:

Well, this is awkward.? ? I had replaced both Diff Amp transistors with the same part number.??
I just found that this package is actually reverse from what i removed!? ?using a dorky little?

heathkit transistor tester, i found the pinout to be CBE when looking at it from the flat side.??
The center leg was bent towards the rear of the part, while the original part was bent toward the front.??

I reversed both transistors, and got -106 which i was able to adjust appropriately!? ? ??

Strange, but makes sense.? ?I have a few more of these buggers in the supply, so it sounds like I'll be sorting through that today.? ?

Will report back in a bit.??

Thanks for the help ya'll!?

?

On Thu, Aug 29, 2024 at 10:45?AM Frank Mashockie via <fmashockie=[email protected]> wrote:

Also, don't forget about Gianni's HP e-book.? If I remember correctly, he goes into great detail about the 141T.? Here's the link.? They can be slow to download and I forget which section includes the 141T. ?

Join [email protected] to automatically receive all group messages.