¿ªÔÆÌåÓý

Re: HP 8341A how to install attenuator


 

¿ªÔÆÌåÓý

Hello Guy,

I¡¯m traveling at the moment so I don¡¯t have references in hand ?but yes, you need to modify the firmware so the attenuator will be recognized. There is a procedure in the service manual. There should be a piece of paper in a small cage in the left rear side of the instrument and it will show the address that needs to be changed. You can do that with front panel commands and store the changes.?

73,

Steve
WB0DBS
EM17iq09ao


On Mar 9, 2023, at 5:28 AM, Guy DAIPRA via groups.io <on1ev@...> wrote:

?Hello, I do have an HP 8341A without attenuator? and I have installed a new attenuator inside. the attenuator do not function .
  • Do I have to change the eprom of the processor? board? to have access to the ctrl of the attenuator ??
  • Is there another solution: e.g. a bridge to be placed inside the sweeper ?
  • Is there an option to be activated inside the firmware of the sweeper ?
thanks for any information?
ON1EV

Join [email protected] to automatically receive all group messages.