¿ªÔÆÌåÓý

Re: Mating APC 3.5 with APC 2.92 connectors


 

¿ªÔÆÌåÓý

So the HUBER+SUHNER? Torque wrench 74 Z-0-0-21 is set to 1 Nm? = 8.85 in lbs

?

OK for SMA but a bit high for For 3.5mm, 2.92mm, and 1.85mm as the recommended torque is 8 in-lbs

Perhaps HUBER+SUHNER? should standardise there 8 mm Torque wrench to below 1 Nm at 8 in lbs making it compliant ?with

??? ??? SMA , 3.5mm, 2.92mm, and 1.85mm made out of stainless steel

My HUBER+SUHNER? Torque wrench 74 Z-0-0-21 came in with a lode of ex aircraft spares still in its original sealed rapping a good find for me

?

Paul

?

From: [email protected] [mailto:[email protected]] On Behalf Of Paul Bicknell
Sent: 04 March 2023 23:15
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Mating APC 3.5 with APC 2.92 connectors

?

Hi Tom

Thank you for the information and from memory that looks familiar as I thought it was strange that the SMA was 4 in-lbs and the

3.5mm, 2.92mm, and 1.85mm recommended torque is higher at 8 in-lbs but the SMA value I was given must have been for Brass

It all makes sense now. my Suhner Torque ranch is set to 1Nm I must check what this = in-lbs

Regards Paul

?

From: [email protected] [mailto:[email protected]] On Behalf Of Tom Lee
Sent: 04 March 2023 22:55
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Mating APC 3.5 with APC 2.92 connectors

?

From Microwaves101.com:

??????? SMA (stainless steel) 7-10 in-lbs
??????? SMA (brass) 3-5 in-lbs
??????? SSMA (stainless steel) 7-8 in-lbs
??????? SMM (stainless steel) 2-3 in-lbs
??????? SMC (brass) 3-5 in-lbs
??????? Type N 12-15 in-lbs
??????? TNC 12-15 in-lbs
??????? SC 12-15 in-lbs
??? ??? For 3.5mm, 2.92mm, and 1.85mm recommended torque is 8 in-lbs
??? ??? To connect 3.5 or 2.92 to SMA, 5 in-lbs
??? ??? For 1 mm connectors, 4 in.-lbs.

Different sources give slightly different values, so use whatever the manufacturer recommends. But in the absence of specific information, the above values should be fine.


--Cheers,
Tom

-- 
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
420 Via Palou Mall
Stanford University
Stanford, CA 94305-4070

On 3/4/2023 14:45, Paul Bicknell wrote:

Hi Steve it has been a long time since we have spoken

?

Regarding maximum torque I recently read about the torque requirements but unfortunately I cannot find the article? it defined the

Torque settings for SMA , 3.5 , 2.92,alao 2.4 , 1.85 and they were not all the same?

?

Do you or anyone have the Torque settings for the above connectors

?

Join [email protected] to automatically receive all group messages.