Hi,

Just checked my HP 410C, and only Common is regular wire,? both Ohms/Amps and DCV are coax.

Regards,
Askild


On Tue, Mar 1, 2022 at 6:49 PM Bob Albert via <bob91343=[email protected]> wrote:
Don't forget the resistor in the DC probe.

On Tuesday, March 1, 2022, 05:00:07 AM PST, n4buq <n4buq@...> wrote:


COMMON and OHMS/AMPS is regular probe wire.? DCV is shielded (coaxial).? ACV is shielded with two conductors (triaxial).

Thanks,
Barry - N4BUQ


From: "Askild" <megafluffy@...>
To: "HP-Agilent-Keysight-equipment" <[email protected]>
Sent: Tuesday, March 1, 2022 6:46:18 AM
Subject: Re: [HP-Agilent-Keysight-equipment] HP 410B probe wire
Hi,

Some of these are screened? (caoaxial)??wires, but the screen is only connected inside the instrument, and not connected on the probe.
So please check how many of them are coaxial.? Or maybe someone else has the correct answer.

Regards,
Askild



On Tue, Mar 1, 2022 at 3:39 AM saipan59 (Pete) <saipan1959@...> wrote:
I believe the most important feature of "probe wire" is that it have very fine strands, so that it doesn't break internally with a lot of flexing.
And of course, the insulation needs to be appropriate to the peak voltage levels.

Pete