Keyboard Shortcuts
Likes
Search
TinySA ULTRA LEVEL CAL -> Signal Level incorrect
#ultra
zeenko store (aliexpress) Order date: Dec 18, 2022 I have tinysa ULTRA with HW V0.4.5.1. Apply a new firmware today tinySA4_v1.4-33-ga8bffcf.bin to my TinuSa ULTRA CONFIG -> MORE -> CLEAR CONFIG CONFIG -> LEVEL CAL -> RESET CALIBLATION & CALIBLATE 100kHz to 5.34GHz an error was displayed. >1T 30.00026MHz -60,9dBm/T Signal Level incorrect Check cable between connection connectors Touch screen to continue I have reverted to the firmware around the date of purchase. Can I install the latest firmware? tinySA4_v1.3-556-g9825888.bin is CALIBLATION and SELF TEST OK tinySA4_v1.3-563-g1da836e.bin is Signal Level incorrect Masahiro Kusunoki/JP3SRS |
开云体育Yes ? Calibration was performed using the RG223 cable. This cable is also used in VNA and there are no broken wires. ? Thank you in advance. ? From: [email protected] <[email protected]> On Behalf Of Erik Kaashoek
Sent: Monday, January 9, 2023 11:29 PM To: [email protected] Subject: Re: [tinysa] TinySA ULTRA LEVEL CAL -> Signal Level incorrect #ultra ? Can test again with a different coax cable? For more info on the tinySA go to |
开云体育Erik-san ? No, I changed to RG223 cable and performed the test. tinySA4_v1.3-556-g9825888.bin is CALIBLATION and SELF TEST OK firmware after tinySA4_v1.3-563-g1da836e.bin is Signal Level incorrect. The firmware after tinySA4_v1.4-33-ga8bffcf.bin is also Signal Level incorrect. ? Masahiro Kusunoki/JP3SRS ? From: [email protected] <[email protected]> On Behalf Of Erik Kaashoek
Sent: Monday, January 9, 2023 11:42 PM To: [email protected] Subject: Re: [tinysa] TinySA ULTRA LEVEL CAL -> Signal Level incorrect #ultra ? So all OK now? For more info on the tinySA go to |
开云体育Eric-sanWhat code did you use with the clear config? ? 1234 Enter Did the tinySA reboot after entering the code ? Yes Did the ref level indicator turn red after the clear config ? Yes Masahiro Kusunoki/JP3SRS 2023/01/10 1:16、Erik Kaashoek <erik@...>のメール:
|
This is very confusing.
Can you use the serial over usb command to dump the low correction table: correction low
This will dump the correction table.
Also give the command? leveloffset And post the respone Please post the data using the latest fw as I do not understand why one fw gives a success selftest and others do not -- For more info on the tinySA go to https://tinysa.org/wiki/
|
开云体育
tinySA Shell ch> (null)? ch> correction low index frequency value correction low 0 10000 14.2 correction low 1 40000 9.5 correction low 2 100000 7.6 correction low 3 300000 3.8 correction low 4 500000 1.5 correction low 5 900000 0.1 correction low 6 4000000 0.3 correction low 7 6000000 0.0 correction low 8 30000000 0.0 correction low 9 90000000 -0.8 correction low 10 210000000 0.7 correction low 11 300000000 0.4 correction low 12 380000000 1.4 correction low 13 510000000 1.0 correction low 14 600000000 0.8 correction low 15 690000000 0.7 correction low 16 780000000 2.6 correction low 17 810000000 5.2 correction low 18 820000000 6.9 correction low 19 830000000 9.5 ch> leveloffset leveloffset low 0.0 leveloffset low output 0.0 leveloffset switch 0.0 leveloffset receive_switch 0.0 leveloffset out_switch 0.0 leveloffset lna 0.0 leveloffset harmonic 10.5 leveloffset shift1 0.5 leveloffset shift2 3.0 leveloffset shift3 0.0 leveloffset drive1 0.0 leveloffset drive2 -1.5 leveloffset drive3 -0.5 leveloffset direct 30.0 leveloffset direct_lna 0.0 leveloffset ultra 0.0 leveloffset ultra_lna 0.0 leveloffset adf 0.0 leveloffset direct output 0.0 ch> ?
? Display Error message on LCD screen >1T 29.999MHz -61.6dBm ? Test 12: Pass Test 13: signal level Fail Touch screen to continue ?
? Display Error message on LCD screen >1T 30.00028MHz -60.7dBm/T ? Signal level incorrect Check cable between connectors Touch screen to continue ?
? before????????????????????????????????? after tinySA4_v1.3-556-g9825888.bin?????????? firmware tinySA_v1.4-33-ga8bffcf.bin index frequency value?????????????????? index frequency value correction low 0 10000 14.2???????????? correction low 0 10000 14.2 correction low 1 40000 9.5????????????? correction low 1 40000 9.5 correction low 2 100000 7.6???????????? correction low 2 100000 7.6 correction low 3 300000 3.8???????????? correction low 3 300000 3.8 correction low 4 500000 1.5???????????? correction low 4 500000 1.5 correction low 5 900000 0.1???????????? correction low 5 900000 0.1 correction low 6 4000000 0.3??????????? correction low 6 4000000 0.3 correction low 7 6000000 0.0????????? ??correction low 7 6000000 0.0 correction low 8 30000000 0.0?????????? correction low 8 30000000 0.0 correction low 9 90000000 -0.8????????? correction low 9 90000000 -0.8 correction low 10 210000000 0.7???????? correction low 10 210000000 0.7 correction low 11 300000000 0.4???????? correction low 11 300000000 0.4 correction low 12 380000000 1.4???????? correction low 12 380000000 1.4 correction low 13 510000000 1.0???????? correction low 13 510000000 1.0 correction low 14 600000000 0.8???????? correction low 14 600000000 0.8 correction low 15 690000000 0.7???????? correction low 15 690000000 0.7 correction low 16 780000000 2.6???????? correction low 16 780000000 2.6 correction low 17 810000000 5.2???????? correction low 17 810000000 5.2 correction low 18 820000000 6.9???????? correction low 18 820000000 6.9 correction low 19 830000000 9.5???????? correction low 19 830000000 9.5 ch> leveloffset???????????????????????? ch> leveloffset leveloffset low 0.0???????????????????? leveloffset low -1.3 leveloffset low output 0.0????????????? leveloffset low output 0.0 leveloffset switch 0.0????????????????? leveloffset switch 0.0 leveloffset receive_switch 0.0????????? leveloffset receive_switch 0.0 leveloffset out_switch 0.0????????????? leveloffset out_switch 0.0 leveloffset lna 0.0???????????????????? leveloffset lna 0.0 leveloffset harmonic 10.5?????????????? leveloffset harmonic 10.5 leveloffset shift1 0.5????????????????? leveloffset shift1 -0.1 leveloffset shift2 3.0????????????????? leveloffset shift2 2.9 leveloffset shift3 0.0????????????????? leveloffset shift3 -0.0 leveloffset drive1 0.0????????????????? leveloffset drive1 0.0 leveloffset drive2 -1.5???????????????? leveloffset drive2 -1.0 leveloffset drive3 -0.5???????????????? leveloffset drive3 -0.2 leveloffset direct 30.0???????????????? leveloffset direct 30.0 leveloffset direct_lna 0.0????????????? leveloffset direct_lna 0.0 leveloffset ultra 0.0?????????????????? leveloffset ultra 0.0 leveloffset ultra_lna 0.0?????????????? leveloffset ultra_lna 0.0 leveloffset adf 0.0???????????????????? leveloffset adf 0.0 leveloffset direct output 0.0?????????? leveloffset direct output 0.0 ? There was a change in the value of leveloffset. ? Masahiro Kusunoko/JP3SRS ? From: [email protected] <[email protected]> On Behalf Of Erik Kaashoek
Sent: Tuesday, January 10, 2023 5:58 AM To: [email protected] Subject: Re: [tinysa] TinySA ULTRA LEVEL CAL -> Signal Level incorrect #ultra ? This is very confusing. correction low This will dump the correction table. And post the respone For more info on the tinySA go to |
The calibration did partially complete, the LNA level calibration seems to have failed.
And selftest 13 fails, this is the LNA test. It seems the LNA is not working correctly The reason the older FW did not show this error is because the check of the signal level during LNA calibration was only introduced in newer FW version. You can try yo replace the LNA, this has previously been done by someone so it seems to be possible. The LNA is U14 and it is a?BGA2817, available from many suppliers for acceptable price. Or you can contact the supplier for replacement. -- For more info on the tinySA go to https://tinysa.org/wiki/ |
开云体育Erik-san ? Thank you for taking a look at the caliblation results. I will replace the parts for U14 BGA2817 as advised. It will take some time for the parts to arrive so I will work on it next week. ? Thanks. Masahiro Kusunoki/JP3SRS ? From: [email protected] <[email protected]> On Behalf Of Erik Kaashoek
Sent: Tuesday, January 10, 2023 7:42 PM To: [email protected] Subject: Re: [tinysa] TinySA ULTRA LEVEL CAL -> Signal Level incorrect #ultra ? The calibration did partially complete, the LNA level calibration seems to have failed. For more info on the tinySA go to |
开云体育Erik-san ?
CLEAR CONFIG -> CALIBLATE 100kHz to 3.54GHZ PASS SELF TEST PASS ? There was also a change in the value of lna in Level offset. Before LNA Replace????????????????????? After LNA Replace ch> correction low????????????????????? ch> correction low index frequency value?????????????????? index frequency value correction low 0 10000 12.2???????????? correction low 0 10000 12.2 correction low 1 50000 7.6????????????? correction low 1 50000 7.6 correction low 2 200000 4.5???????????? correction low 2 200000 4.5 correction low 3 400000 2.2???????????? correction low 3 400000 2.2 correction low 4 900000 0.4???????????? correction low 4 900000 0.4 correction low 5 20000000 -0.4????? ????correction low 5 20000000 -0.4 correction low 6 30000000 0.0?????????? correction low 6 30000000 0.0 correction low 7 100000000 -0.8???????? correction low 7 100000000 -0.8 correction low 8 160000000 -0.4???????? correction low 8 160000000 -0.4 correction low 9 230000000 0.5????????? correction low 9 230000000 0.5 correction low 10 290000000 0.3???????? correction low 10 290000000 0.3 correction low 11 400000000 1.0???????? correction low 11 400000000 1.0 correction low 12 520000000 0.1???????? correction low 12 520000000 0.1 correction low 13 600000000 0.5???????? correction low 13 600000000 0.5 correction low 14 660000000 0.4???????? correction low 14 660000000 0.4 correction low 15 740000000 1.5???????? correction low 15 740000000 1.5 correction low 16 790000000 3.0???????? correction low 16 790000000 3.0 correction low 17 810000000 4.7???????? correction low 17 810000000 4.7 correction low 18 820000000 6.3???????? correction low 18 820000000 6.3 correction low 19 830000000 8.7???????? correction low 19 830000000 8.7 ? ch> leveloffset???????????????????????? ch> leveloffset leveloffset low 0.0???????????????????? leveloffset low -1.2 leveloffset low output 0.0??????? ??????leveloffset low output 0.0 leveloffset switch 0.0????????????????? leveloffset switch 0.0 leveloffset receive_switch 0.0????????? leveloffset receive_switch 0.1 leveloffset out_switch 0.0????????????? leveloffset out_switch 0.0 leveloffset lna 0.0?? ??????????????????leveloffset lna -0.9 leveloffset harmonic 10.5?????????????? leveloffset harmonic 10.5 leveloffset shift1 -0.1???????????????? leveloffset shift1 -0.1 leveloffset shift2 2.4????????????????? leveloffset shift2 2.4 leveloffset shift3 -0.4 ????????????????leveloffset shift3 -0.5 leveloffset drive1 0.0????????????????? leveloffset drive1 0.0 leveloffset drive2 -1.0???????????????? leveloffset drive2 -0.9 leveloffset drive3 -0.1???????????????? leveloffset drive3 -0.1 leveloffset direct 30.0? ???????????????leveloffset direct 29.4 leveloffset direct_lna 0.0????????????? leveloffset direct_lna 30.9 leveloffset ultra 0.0?????????????????? leveloffset ultra -1.6 leveloffset ultra_lna 0.0?????????????? leveloffset ultra_lna -1.0 leveloffset adf 0.0???????????????????? leveloffset adf 0.0 leveloffset direct output 0.0?????????? leveloffset direct output 0.0 ? ? The problem of passing with the old firmware and failing with the new v1.4 has been resolved with LNA replacement. I will close this issue. ? Thanks. ? Masahiro Kusunoki/JP3SRS ? From: [email protected] <[email protected]> On Behalf Of Masahiro Kusunoki
Sent: Tuesday, January 10, 2023 8:06 PM To: [email protected] Subject: Re: [tinysa] TinySA ULTRA LEVEL CAL -> Signal Level incorrect #ultra ? Erik-san ? Thank you for taking a look at the caliblation results. I will replace the parts for U14 BGA2817 as advised. It will take some time for the parts to arrive so I will work on it next week. ? Thanks. Masahiro Kusunoki/JP3SRS ? From: [email protected] <[email protected]> On Behalf Of Erik Kaashoek ? The calibration did partially complete, the LNA level calibration seems to have failed. For more info on the tinySA go to |
Masahiro-san
Thanks for the excellent feedback. I will send this info to manufacturing for future improvements. The non zero leveloffset values are either in the firmware of measured during calibration As the LNA calibration previously did fail the leveloffset LNA stayed zero. Now the LNA is working the calibration process updates it to the measured value. -- For more info on the tinySA go to https://tinysa.org/wiki/ |