Hello,
I've got a tinySA Ultra in December 2022 and all was good (tests & calibration at least). I also have a tinySA non-ultra. Yesterday while playing around, I've noticed that the signal level indicated by the Ultra is lower by about 10dB compared to the other. I tried to do a self-test and level calibration and both failed (pictures attached). I've cleared configuration and disabled "save config" in "presets" turned off and on (after a while as instructed by the seller) and measured the 30MHz CAL output signal which stays at ~-45dBm compared to the expected -35dBm. Measured the CAL signal with the other tinySA and it is around -35dBm so the output is good. Third picture is the CAL signal as measured by Ultra. Is there anything else that can be done in terms of software/configuration or something is broken in hardware? The system itself kinda works, but all signals are lower. ![]()
WhatsApp Image 2023-01-18 at 23.02.02 _1_.jpeg
![]()
WhatsApp Image 2023-01-18 at 23.02.02.jpeg
![]()
WhatsApp Image 2023-01-19 at 11.25.58.jpeg
|
Do you have a scope?
If yes, can you input a 10MHz signal between -20dBm and 0dBm (such that it is well visible on your scope), you can use the tinySA to generate such a signal? Remove carefully the cover close to the RF connector Switch on the tinySA Ultra in default settings (0-800MHz) and measure the signal level at C58 and C24. (either side of the capacitor is OK) The levels should be fairly close. If not, the input switch is not working correctly -- For more info on the tinySA go to https://tinysa.org/wiki/ |
Hello, thank you for the response.
At 10MHz@-7dBm?output from tinySA, I get the following data: - level indicated by Ultra: -17dBm - input pin ~120mVpp - C58 ~120mVpp - C24 ~100mVpp |
I think (I may be wrong, as a beginner) that the switch is for IN/OUT. If that is the case, I can replicate the 10dB error by doing the reverse: generate a signal with the Ultra and read it on tinySA. At [email protected],?tinySA reads -27dBm. Using the CAL output @10MHz I see -34dBm on tinySA.
|