On Mon, Aug 10, 2020 at 01:46 PM, David Lee wrote:
I noticed that the wiki shows two different values for correction table slot 4. Under firmware update it shows a value of 50000000 0.0? Under USB command: correction it shows slot 4 as?30000000 0.0 (the setting I am using now)
If I reset the level cal, then do the selftest, only Test 10 fails with "signal level critical"
David,
? Erik used precision components in the production level tinySa and some of the Wiki content may not reflect current correction table values.
? Before you updated to the new firmware version you leveloffset values should have been close to the following:
? leveloffset low 0
? leveloffset high 100
? and your correction table values should have been as follows:
0 10000 6.0
1 100000 2.8
2 200000 1.6
3 500000 -0.4
4 50000000 0.0
5 140000000 -0.4
6 200000000 0.4
7 300000000 3.0
8 330000000 4.0
9 350000000 8.1
With those values performing a selftest should run without error unless your calibrator signal level is a lot off from -25dBm.
Try the following from a serial terminal:
1. enter <clearconfig 1234>
2. enter <reset>
3. re-establish your serial connection
4. enter <leveloffset low 0>
5. enter <saveconfig>
6. run a selftest? [Config->Self Test on the tinySA or <selftest 0> from the serial terminal]
The self test should run without error.? If you entered a leveloffset low value of 3 when it should have been 0, all your values will be 3dB too high which would explain why your selftest failed.? If the above doesn't solve your issue then hopefully Erik can offer another solution.
If the above works its not necessary to run a Config->Level Cal as it sets the leveloffset low value as you did manually.? Also, as a final check there should be no red values displayed in the side status panel.
- Herb?