On Wed, Dec 16, 2020 at 04:45 AM, Erik Kaashoek wrote:
Set the start frequency to 1MHz to avoid the zero Hz peak that keeps the auto attenuator at 30dB
Or disable auto attenuation in the startup settings and set attenuation to zero
? In addition to the attenuation setting, the resolution bandwidth (RBW) of any spectrum analyzer also determines its noise floor (and how small a signal you can measure).? So if you need to measure signals in the -90dBm range then in addition to using 0dB attenuation you should also lower your RBW to at least 100kHz (Frequency->RBW->100kHz).
? Lower bandwidths increase your sweep time so unless you really need to scan the entire 350MHz span you should reduce your frequency range to the area you are interested in.? Examples follow using my tinySA with the same capacitors installed.
With 30dB attenuation and 600kHz RBW the? ? ? ? ? ? ?With 0dB attenuation and 600KHz RBW the noise? ? ? With 0dB attenuation and 10kHz RBW the noise
noise floor is ~-60dBm.? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ?floor is ~-89dBm. No 48kHz signals detected.? ? ? ? ? ?floor is ~-100dBm.? Reduced 48KHz related signal detected.

??

??

Note in the above screenshots how the scan time for a 350MHz span increases from 405ms to 115.8s for a RBW change of 600kHz to 10kHz.? Professional grade spectrum analyzers also have increased sweep times with reduction in RBW, but are orders faster, and orders more expensive than the tinySA.
- Herb