开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

Good battery life

 

Hooray for the tinyVA battery.? I had not used my unit in over a year, turned it on, and the battery was at almost full charge.? To me, that is an outstanding shelf life from a quality rechargeable battery.? Thank you.


Re: Measuring resonant coil outputs

 

Hi Dave,
?
Lots of useful info there. Thanks. I will also write separately to you.
--
Julian


Re: TinySA ULTRA 4" - No longer works

 

On Fri, Apr 4, 2025 at 04:32 PM, Gerd wrote:
Erik, did you design the printed circuit board ?
?
Yes
?
?
Before I started to disassemble the device, I decided to check if the signal generator was working. A picture of this menu is attached to the letter, the problem also affected this part of the control circuit. When I selected LOW OUTPUT - ON from the menu, a signal appeared at the RF tinySAoutput, but the frequency barely ran 50MHz. From that frequency onwards, the device could no longer hold the signal carrier. I have not tested the signal generator of this device before, so I have no comparison of how it should work.
?
You have to switch the signal generator LOW OUTPUT to ON to get the correct output signal
?
?
I opened the tinySA ULTRA at Erik's suggestion and a visual test did not detect any poor connections.
Since I have had problems with similar cable connections in other devices before, I cleaned the cable terminals just in case (even though there was no visible problem).
?
After reconnecting the ribbon cable, almost all problems disappeared. Self Test was successful.
?
Happy to hear
?
?
Even after a successful Self Test, the signal generator is unable to transmit a signal in normal mode, I have to activate LOW OUTPUT - ON in the menu. Is this a normal approach or is there another problem here?
?
This is normal behavior
I also took a picture of the last measurement.
?
Erik, based on the information you provided, could you tell me which terminal in the ribbon cable had a poor connection? I'll look at it later with a powerful microscope. (It seems that some types of connectors are not worth the money.)
?
Don't know, sometimes we see this problem, hence the suggestion to reseat the cable.
--
Designer of the tinySA
For more info go to


Re: TinySA ULTRA 4" - No longer works

 

Could it be alingment or some solid particle? I had issues like those with expensive connectors.


On Sat, Apr 5, 2025, 01:32 Gerd via <gerd.vipre=[email protected]> wrote:
Hi.
?
> Can you open the case and try to reseat the LCD cable ? Maybe that will solve the problem.
Erik, did you design the printed circuit board ?
?
Before I started to disassemble the device, I decided to check if the signal generator was working. A picture of this menu is attached to the letter, the problem also affected this part of the control circuit. When I selected LOW OUTPUT - ON from the menu, a signal appeared at the RF tinySAoutput, but the frequency barely ran 50MHz. From that frequency onwards, the device could no longer hold the signal carrier. I have not tested the signal generator of this device before, so I have no comparison of how it should work.
?
I opened the tinySA ULTRA at Erik's suggestion and a visual test did not detect any poor connections.
Since I have had problems with similar cable connections in other devices before, I cleaned the cable terminals just in case (even though there was no visible problem).
?
After reconnecting the ribbon cable, almost all problems disappeared. Self Test was successful.
?
Even after a successful Self Test, the signal generator is unable to transmit a signal in normal mode, I have to activate LOW OUTPUT - ON in the menu. Is this a normal approach or is there another problem here? I also took a picture of the last measurement.
?
Erik, based on the information you provided, could you tell me which terminal in the ribbon cable had a poor connection? I'll look at it later with a powerful microscope. (It seems that some types of connectors are not worth the money.)
?
Sincerely,
g
?
?


Re: TinySA ULTRA 4" - No longer works

 

Hi.
?
> Can you open the case and try to reseat the LCD cable ? Maybe that will solve the problem.
Erik, did you design the printed circuit board ?
?
Before I started to disassemble the device, I decided to check if the signal generator was working. A picture of this menu is attached to the letter, the problem also affected this part of the control circuit. When I selected LOW OUTPUT - ON from the menu, a signal appeared at the RF tinySAoutput, but the frequency barely ran 50MHz. From that frequency onwards, the device could no longer hold the signal carrier. I have not tested the signal generator of this device before, so I have no comparison of how it should work.
?
I opened the tinySA ULTRA at Erik's suggestion and a visual test did not detect any poor connections.
Since I have had problems with similar cable connections in other devices before, I cleaned the cable terminals just in case (even though there was no visible problem).
?
After reconnecting the ribbon cable, almost all problems disappeared. Self Test was successful.
?
Even after a successful Self Test, the signal generator is unable to transmit a signal in normal mode, I have to activate LOW OUTPUT - ON in the menu. Is this a normal approach or is there another problem here? I also took a picture of the last measurement.
?
Erik, based on the information you provided, could you tell me which terminal in the ribbon cable had a poor connection? I'll look at it later with a powerful microscope. (It seems that some types of connectors are not worth the money.)
?
Sincerely,
g
?
?


Re: Measuring resonant coil outputs

 

1)? For those who didn't bother to look at the diagram, it shows specifically 35 kHz.? Enough said.....

THE REST:

I as well come from a physics background.?

Be extremely careful with those power levels.? At 35 kHz and 1 kV feeding a spark gap, you are evidently using shock excitation to look for ring-down and harmonic generation from your resonant LC network.? As a precaution, I would strongly recommend installation of the following from MiniCircuits installed at the input of the TINYSA.? Leave it installed.?

? ?? VLM-33W-2W-S+?

It is presently under redesign by MiniCircuits, but doing what you do, I'd almost "require" you install it or equivalent to protect the TINYSA - or any SA, for that matter.

Considering your application, that should be installed in addition to attenuators? It would be best to have a selection to match the levels of harmonics and fundamentals excited by the shock exitation.? These are available from ebay, or, better quality, Pasternack.? Values might include the following:? 20 dB, 10 dB, 6 dB, and 3 dB.? Small SMA style packages at 1 watt or less are adequate for your application. ?

Again, considering your power and voltage levels, I'd start with the short included antenna.? As such, it is an extremely short portion of a wavelength.? Therefore, you likely won't experience any frequency bias from the antenna up to where it becomes a 0.1-wavelength and more. ?

Also start at some distance from your spark gap and the LC network.? Approach the operating system with caution to, again, be "kind" to the SA. ?

You will also require multiple discharges of the spark gap to capture the ring-down and harmonic structure.? It would be advised to use the MAX HOLD function to capture the sum of a number of gap discharges.? Otherwise, the sweep is too slow (andy SA will exhibit this) to faithfully capture what you're after.??

I've done just what you are looking for before retirement for Uncle at KAFB in Albuquerque.? All I'll write based on the experience with Uncle is: "It's amazing what can be done with wide-band RF pulses".? I've also experimented as you here in my small lab.? If you want to email me offline of the group, please - I'm open.? Emails are:

? ?? DaveArea51A at gmail dot com
? ?? W zero LEV at arrl dot net

Glad to help where I can.

Dave - W?LEV? ? ??


On Fri, Apr 4, 2025 at 9:53?AM Kerrowman via <979jules979=[email protected]> wrote:

Hi all,

I’m new to this group, a physicist by background but also new to using a spectrum analyser. Before I go and actually buy the Aursinc V0.3.1 model, I wanted to ask about how I might go about making measurements relevant to my research so I can be sure this is the best and most affordable approach.

To summarise, I am investigating resonance in coils driven at radio frequencies and high voltages and looking at the standing waves that result from certain coil configurations and wire lengths. To make it easier to show what I’m doing, I attach some graphics showing the circuit (without an output load), a typical coil configuration and an example of standing waves that show on a scope using a low-powered setup of 18V to the primary coil.

In particular, the queries I have are:

  • Can I detect what frequencies are emanating from the coils by using the supplied aerial placed within say 50cm of the coils?
  • What sort of attenuator would I need to cope with the high voltages and power in the secondary in particular?
  • Are there any others known about how are doing this type of measurements?
  • Will the tinySA-App work with Windows 10?

Thank you in advance for any help.

?

?
Julian Perry



--
Dave - W?LEV



Re: TinySA ULTRA 4" - No longer works

 

Indeed, the display test failed also. This indicates bad communication over the SPI bus.
Can you open the case and try to reseat the LCD cable? Maybe that will solve the problem
--
Designer of the tinySA
For more info go to


Re: Measuring resonant coil outputs

 

Thank you. Are the attenuators specific to your design of tinySA or are they are standard fitting device available from various outlets?
?
One other point; what feature of a single loop (of 2mm tinned copper wire) determines its optimum detection frequency and is there an optimum diameter or configuration for my frequency range (e.g. square vs circular)?
?
--
Julian


Re: TinySA ULTRA 4" - No longer works

 

It looks like the MCU can't read data through the SPI interface. if it still doesn't work after removing the SD card, the MUC's SPI port may be faulty. the SD card, LCD, and RF chip are all mounted on the same SPI bus. Maybe the SPI bus was damaged when the SD card was inserted or removed.


Re: Measuring resonant coil outputs

 

When you are probing with a loop antenna you can start at large distance (two meters) and reduce the distance if there are no strong signals.
Unless you are exiting the coil with a LOT of power, then you may have to start at larger distance.
To be absolutely sure, you can put a 20 or 30 dB attenuator directly on the RF connector of the tinySA and connect the antenna to the attenuator.
Make sure you never ever touch anything with the telescopic antenna.
--
Designer of the tinySA
For more info go to


Re: Measuring resonant coil outputs

 

On Fri, Apr 4, 2025 at 03:57 AM, Kerrowman wrote:
I know the tinySA is not accurate below 100kHz but I have heard that by making some Config adjustments then it will detect them.
The tinySA Basic (0.3.1 model) will actually work well down to 10 kHz but you lose some dynamic range
--
Designer of the tinySA
For more info go to


Re: Measuring resonant coil outputs

 

What about the issue of potentially damaging the SA? How can I work out the dB of the fields surrounding the coils to best protect it? I do have one of those cheap field meters for elec and mag fields, which measures in V/m and ?T.
--
Julian Perry
Kerrowenergetics.org.uk


Re: Measuring resonant coil outputs

 

:)
--
Julian Perry
Kerrowenergetics.org.uk


Re: Measuring resonant coil outputs

 

Yes, an important question. 10kHz - 20MHz I think is reasonable. I know the tinySA is not accurate below 100kHz but I have heard that by making some Config adjustments then it will detect them.
--
Julian Perry
Kerrowenergetics.org.uk


Re: Measuring resonant coil outputs

 

On 04/04/2025 09:54, Kerrowman via groups.io wrote:
Will the tinySA-App work with Windows 10?
Yes.

David
--
SatSignal Software - Quality software for you
Web:
Email: davidtaylor@...
BlueSky: @gm8arv.bsky.social, Twitter: @gm8arv


Re: Measuring resonant coil outputs

 

What frequencies are you targeting?
The supplied antenna is good for 50 MHz and higher but is a lot less sensitive for 10 MHz and lower
For low frequencies you better use a (home made) loop antenna
--
Designer of the tinySA
For more info go to


Measuring resonant coil outputs

 

Hi all,

I’m new to this group, a physicist by background but also new to using a spectrum analyser. Before I go and actually buy the Aursinc V0.3.1 model, I wanted to ask about how I might go about making measurements relevant to my research so I can be sure this is the best and most affordable approach.

To summarise, I am investigating resonance in coils driven at radio frequencies and high voltages and looking at the standing waves that result from certain coil configurations and wire lengths. To make it easier to show what I’m doing, I attach some graphics showing the circuit (without an output load), a typical coil configuration and an example of standing waves that show on a scope using a low-powered setup of 18V to the primary coil.

In particular, the queries I have are:

  • Can I detect what frequencies are emanating from the coils by using the supplied aerial placed within say 50cm of the coils?
  • What sort of attenuator would I need to cope with the high voltages and power in the secondary in particular?
  • Are there any others known about how are doing this type of measurements?
  • Will the tinySA-App work with Windows 10?

Thank you in advance for any help.

?

?
Julian Perry
Kerrowenergetics.org.uk


Re: TinySA ULTRA 4" - No longer works

 

Try:

Remove your SD card.

Clear configuration
Load defaults
Level calibration


On Thu, 03 Apr 2025 13:21:44 -0700
"Gerd via groups.io" <gerd.vipre@...> wrote:

Hi.

With nothing connected to the inputs and all default settings (after clear config)
I did a "clear config" again and was able to set the time and date correctly.
Last time I put a point between the dates and it broke the recording.

The remaining errors are the same.

Sincerely,
g.
--

73

-Jim
NU0C


Re: TinySA ULTRA 4" - No longer works

 

I wonder if there is a possibility that the firmware is corrupted in some way. Perhaps re-installing the same version of firmware might produce a clue?

Just my 2 cents!

Sam Reaves
ARS W3OHM
Owner / Moderator of:
LeCroy Owners Group on 开云体育
Sencore Owners Group on 开云体育
Sprint Layout Group on 开云体育??
Pulsonix EDA Group on 开云体育
LPKF Owners Group on 开云体育
Electronics and Mechanical Hardware Design Engineering Manager
Staff Scientist Andritz Rolls Global Research Center (RETIRED)



On Thu, Apr 3, 2025 at 4:21?PM Gerd via <gerd.vipre=[email protected]> wrote:
Hi.
?
> With nothing connected to the inputs and all default settings (after clear config)?
?
I did a "clear config" again and was able to set the time and date correctly.
Last time I put a point between the dates and it broke the recording.
?
The remaining errors are the same.
?
Sincerely,
g.
?


Re: Auto Save Issues

 

Basically I want to save continuous traces as fast as the TinySA ultra allows until the SD card is filled. With the autosave function the fastest it goes is 1 save every 30 seconds. When I manually press trace to SD I can create a new save about every 3 seconds. My scan time for the desired data is 270 ms.? What settings can be used to make the autosave function run as fast as possible? I am not worried about using up the entire SD card rather quickly. I actually expect this.??
Best to you and thanks,
James