¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io

cal isolation


 

Hi,what do i do during the cal when it gets too the isolation?,ive seen a few vids but its not clear,do i connect the 50 ohm load to ch 0 or leave nothing connected to that port,or put the load on ch 1 ,its not clear what i need to do.


 

tekguy,
You should connect 50 ohm loads to both connectors. If you have an attenuator of 10 dB or more, that can be used as a reasonable 50 ohm load.
--John Gord

On Sat, Feb 18, 2023 at 12:43 PM, tekguy la wrote:


Hi,what do i do during the cal when it gets too the isolation?,ive seen a few
vids but its not clear,do i connect the 50 ohm load to ch 0 or leave nothing
connected to that port,or put the load on ch 1 ,its not clear what i need to
do.


 

Many times, it's appropriate to skip the isolation cal. Even
professionally, HP (the 8753C) gives one the choice to specifically skip
isolation. If you do conduct the isolation cal, terminate both ports with
a good 50-Ohm load.

Dave - W?LEV

On Sat, Feb 18, 2023 at 8:43 PM tekguy la <uktekguy@...> wrote:

Hi,what do i do during the cal when it gets too the isolation?,ive seen a
few vids but its not clear,do i connect the 50 ohm load to ch 0 or leave
nothing connected to that port,or put the load on ch 1 ,its not clear what
i need to do.





--
*Dave - W?LEV*
--
Dave - W?LEV


 

does that mean i just leave both ports open and skip to the next step?.


 

Yes, just declare the cal complete without doing the isolation.

Dave - W?LEV

On Sat, Feb 18, 2023 at 10:28 PM tekguy la <uktekguy@...> wrote:

does that mean i just leave both ports open and skip to the next step?.





--
*Dave - W?LEV*
--
Dave - W?LEV


 

On professional VNAs like the 8753, the internal isolation of the hardware is so good that the isolation calibration step is only useful if you are using a fixture that has isolation problems.
For the NanoVNAs (of all flavors, I think), the internal isolation is not so great, and the added calibration step can help improve the dynamic range of S21.
--John Gord

On Sat, Feb 18, 2023 at 02:19 PM, W0LEV wrote:


Many times, it's appropriate to skip the isolation cal. Even
professionally, HP (the 8753C) gives one the choice to specifically skip
isolation. If you do conduct the isolation cal, terminate both ports with
a good 50-Ohm load.

Dave - W?LEV

On Sat, Feb 18, 2023 at 8:43 PM tekguy la <uktekguy@...> wrote:

Hi,what do i do during the cal when it gets too the isolation?,ive seen a
few vids but its not clear,do i connect the 50 ohm load to ch 0 or leave
nothing connected to that port,or put the load on ch 1 ,its not clear what
i need to do.





--
*Dave - W?LEV*


--
Dave - W?LEV


 

On Sun, Feb 19, 2023 at 02:27 AM, John Gord wrote:


the added calibration step can help improve the dynamic range of S21.
Which is of little consequence unless you are measuring high-isolation devices like filters or diplexers, and none at all for those doing one-port reflection measurements (antennas), which I suspect is quite common.
On my commercial VNAs I choose "Omit Isolation" unless it is specifically needed for the measurement at hand.
73, Don N2VGU