help with calibration
The first thing to check is if something is amiss in the calibration
process you are using. One common mistake is that you need to reset the cal before doing a new one. You can also verify the cal before you look at SWR or other measurements. See the detailed description of calibration and verification in "The Absolute Beginners Guide" in the files section of this group, at the following link: /g/nanovna-users/files/Absolute%20Beginner%20Guide%20to%20The%20NanoVNA/Absolute_Beginner_Guide_NanoVNA_v1_5.pdf On Fri, Mar 10, 2023, 5:31 PM AJC via groups.io <AGCHURCHJR= [email protected]> wrote: I have a VNA SAA-2N. It will not calibrate (it says it does) but with a 50 |