No Horizontal Indication


 

I am trying to use my NANOVNA for the first time. I have gone through the calibration process several times. I believe the unit is calibrated, but when I turn it on I do not get the typical horizontal lines. I can turn on the four different traces and see the information about the traces at the top, but there isn¡¯t anything ever displayed. I may have one of the menu items set up incorrectly. How do you get it to sweep? Any suggestions?

Mike - K8AGY
Wires-X
Node 53380
Room 63380


 

Post a picture or take a screenshot so we can see what is going on.

Roger


 

Go into the Scale menu and change SCALE/DIV and/or REFERENCE POSITION for each TRACE


 

What should I change the scale/div or the reference position to?


I appreciate your help.

Thanks,
Mike - K8AGY
Wires-X
Node 53380
Room 63380

On Aug 15, 2020, at 3:45 PM, Oristo <ormpoa@...> wrote:

Go into the Scale menu and change SCALE/DIV and/or REFERENCE POSITION for each TRACE
;data=02%7C01%7C%7Cd60f2e19e930499197d208d84153c916%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C637331175391140360&amp;sdata=psN70IK0QQeiQJ8R6lvDQPFVujZKwfoZwaePqfxIXuI%3D&amp;reserved=0



 

Here is a screen shot.


Mike - K8AGY[cid:AD96CA15-CDED-4B72-8F0D-E002C943C15D]
Wires-X
Node 53380
Room 63380

On Aug 15, 2020, at 3:44 PM, Roger Need via groups.io<> <sailtamarack@...<mailto:sailtamarack@...>> wrote:

Post a picture or take a screenshot so we can see what is going on.

Roger


 

I vaguely recall a post here about a week or two ago
regarding how to enable and disable scanning through the stand-alone user interface.
Somewhere in that menu tree.
Anybody remember?

Jerry, KE7ER


 

What should I change the scale/div or the reference position to?
Depends on FORMAT settings, but numbers on screen shot look pretty huge;
perhaps cal failed (did you first select RESET?)
or "Clearconfig 1234' from a terminal program is wanted
/g/nanovna-users/message/15041


 

Here is a screen shot.
Perhaps CALIBRATION is toggled off?


 

Oops, replace "CALIBRATION" with "CORRECTION".
Depending on FORMAT and CHANNEL settings for each TRACE,
scans can be meaningless..

On Sat, Aug 15, 2020 at 7:11 PM Oristo via groups.io <ormpoa=
[email protected]> wrote:

Here is a screen shot.
Perhaps CALIBRATION is toggled off?




 

Oristo,

That was it! I toggled the ¡°Correction¡± and it is working!

Thank you!

Mike

On Aug 15, 2020, at 7:15 PM, Oristo <ormpoa@...<mailto:ormpoa@...>> wrote:

Oops, replace "CALIBRATION" with "CORRECTION".
Depending on FORMAT and CHANNEL settings for each TRACE,
scans can be meaningless..

On Sat, Aug 15, 2020 at 7:11 PM Oristo via groups.io<> <ormpoa=
[email protected]<mailto:[email protected]>> wrote:

Here is a screen shot.

Perhaps CALIBRATION is toggled off?


 

That was it!
Thanks for feedback; I was also lost there before; CORRECTION option was added by some firmware "update".