Can't save calibration with NanoVNA saver in NanoVNA device,


 

Hello All,
When calibrate the device (nanoVNA-H) all is 100% OK.
When I do calibrate with nanoVNA saver (several versions incl.0.2.2-1), all is
also OK 100%, exept lhe last step - saving the calibrtion in nanoVNA-H. No way to
save this calibration in nanoVNA-H.I can save it to the Disk only.
Seems somewhere I make mistake,,,, Where ?
The connection between the Saver and device is established and active.
The VNA-H is not prepared to get data ?
I have not found the procedure for recording in the device via NanoVNA Saver anywhere. Many things are described how to calibrate, but this is not exactly what it is.
Any Help will be appreciated.Thanks in advance,
Peter


 

Hi Peter,

You can't save to your device with the nanoVNA saver.
The backup is done to your computer.
It is advisable to write detailed information in the note.

Gyula HA3HZ


 

Hi Gyula,
Many thanks for the answer.Will see what I can do wiy saved calibrations on
the Hard disk.
Best Regards !


 

Hi Peter,
just a quick note: On the NanoVNA, you calibrate the device itself, and
what it outputs on its screen, and via the USB cable. In NanoVNA-Saver,
you calibrate exclusively within the software - it relies on the
calibration on the NanoVNA itself remaining constant. This also allows
calibrating with more data points than the NanoVNA itself allows. Using the
"segments" setting, you can get many more points than the 101 of the device
itself. There is no way to save calibration to the device. Neither the
firmware, nor NanoVNA-Saver, supports this.

Since we are on the subject of calibration: NanoVNA-Saver has a calibration
assistant, and 5 manual buttons for calibration. Use the assistant. The
manual buttons take whatever is currently stored (and being displayed) by
the application, and uses that as the calibration data. It does *not* make
a separate sweep for the new standard - it assumes the user knows to do
this themselves. The latest version has a warning popup if a user tries to
use the manual buttons.

I hope you enjoy using the application :-)

--
Rune / 5Q5R

On Sat, 18 Jan 2020 at 18:58, <s_peter@...> wrote:

Hello All,
When calibrate the device (nanoVNA-H) all is 100% OK.
When I do calibrate with nanoVNA saver (several versions incl.0.2.2-1),
all is
also OK 100%, exept lhe last step - saving the calibrtion in nanoVNA-H. No
way to
save this calibration in nanoVNA-H.I can save it to the Disk only.
Seems somewhere I make mistake,,,, Where ?
The connection between the Saver and device is established and active.
The VNA-H is not prepared to get data ?
I have not found the procedure for recording in the device via NanoVNA
Saver anywhere. Many things are described how to calibrate, but this is not
exactly what it is.
Any Help will be appreciated.Thanks in advance,
Peter




 

Hi Rune,
Thank You for the explanation.I know that and use all features of the nanoVNA Saver.The Calibration Assistant is very helpfull ! Very good software for nanoVNA users.Thank You once again.Maybe in future we can get and "Save to device" button.
Best Regards de Peter/Lz2pg


Brian D
 

Sorry to resurrect a dead thread, but I am still a little confused. Here is what confuses me.

If the device is not calibrated, but I do a calibration on NVNA Saver, is the device then calibrated? So if I connect an antenna will I get the correct results? Or...

Does the device need to be calibrated first for the frequency range I want to use? Then will I need to calibrate NVNA Saver?

See where I am confused?

Again, sorry for the lame question.


 

Need made calibration or in NanoVNA Saver or in NanoVNA
If you made or use calibration in NanoVNA Saver need disable calibration in NanoVNA


Brian D
 

Got it. One of the other but not both. Thanks


 

I think DiSlord is incorrect - You can save cal in both environments.
In nanovnaSaver, if there are no cals saved, it uses the devices calibration.
I believe there was a forum thread regarding this back at the start of the year.
Search out Rune's posts as he is the nanovnaSaver creator.

On Monday, June 29, 2020, 12:13:08 p.m. EDT, Brian D via groups.io <sv.takara@...> wrote:

Got it. One of the other but not both. Thanks


 

yes, so far i am using nanovna-h without device calibration. Loading
calibration file (From PC) in nanovasver is necessary if device is not
calibrated.

On Mon, Jun 29, 2020 at 9:50 PM Larry Rothman <nlroth@...> wrote:

I think DiSlord is incorrect - You can save cal in both environments.
In nanovnaSaver, if there are no cals saved, it uses the devices
calibration.
I believe there was a forum thread regarding this back at the start of the
year.
Search out Rune's posts as he is the nanovnaSaver creator.

On Monday, June 29, 2020, 12:13:08 p.m. EDT, Brian D via groups.io
<sv.takara@...> wrote:

Got it. One of the other but not both. Thanks






--

*D.Prabakaran *
*VU3DXR*


 

I believe Rune said that the NanoVNA should be self-calibrated at least over It¡¯s complete frequency span in order to prevent bad data causing NanoVNA-saver to crash. Then you should cal in NanoVNA-saver. In NanoVNA-saver if you set up multiple segments you really need to calibrate in NanoVNA-saver.
Gary
W9TD


 

On Mon, Jun 29, 2020 at 09:20 AM, Larry Rothman wrote:


I think DiSlord is incorrect - You can save cal in both environments.
In nanovnaSaver, if there are no cals saved, it uses the devices calibration.
I believe there was a forum thread regarding this back at the start of the
year.
Search out Rune's posts as he is the nanovnaSaver creator.

On Monday, June 29, 2020, 12:13:08 p.m. EDT, Brian D via groups.io
<sv.takara@...> wrote:

Got it. One of the other but not both. Thanks
If you want use calibration in NanoVNA Saver, then Need reset calibration on NanoVNA, made calibration on NanoVNA Saver, and use this calibration and always disable calibration on NanoVNA then run nanoVNA saver mode.

Or made calibration on NanoVNA and not do it on NanoVNA saver calibration.

NOT need made calibration together. it possible, but not need