Re: Help with LPF Measurement
One thing you could also try is running the filter in a program like Elsie (which is free, for Windows only, though) and look at the S11 plot from Elsie and see if it looks like what you're measuring for S11. A huge mismatch out of band isn't unusual, but it should be fairly decent below the filter cutoff.
toggle quoted message
Show quoted text
Other things to check - self resonance on the inductors? Or capacitive coupling around the filter. The S21 shows the rolloff, but then bounces back up. That's sort of suspicious. It almost looks like not all the sections are active (or that there's some other resonances, like in an Elliptic, you see a bounce back up) -----Original Message-----
From: <[email protected]> Sent: Mar 12, 2025 4:11 PM To: <[email protected]> Subject: [nanovna-users] Help with LPF Measurement I've searched through the message list looking for symptoms similar to mine and didn't find anything. I'm somewhat new to this and would appreciate any help you can offer. I've been trying to get my NanoVNA-F (firmware 1.0.5) to show the results of a 40-meter low pass filter I've built. I know I'm doing something wrong here but can't figure it out. The VNA image is calibrated (including pass-thru), set for the range 6.5 to 29 MHz and logmag S21. Shouldn't this give a flat area leading up to the design frequency followed by a significant roll off through the remainder of the frequency range? My setup is very similar to W2AEW's YouTube demonstration of the NanoVNA with a LPF but gives entirely different results. The image is entirely different, with very large insertion loss and a steep area of suppression peaking at 11 MHz followed by a steep recovery. I worried that my filter wasn't what I thought it was so I tested it with a 7 MHz square wave, looking at input and output with the FFT function of my 'scope. The before and after images show several things: (1) the square wave (yellow trace) is hardly a square wave. It wasn't much of one before attaching to the input of the filter but it's definitely loaded and distorted by the filter. (I can live with the distortion in the test, since the filter is supposed to reduce the high freq distortion anyway, right?) (2) The purple trace is the filter output--pretty nice sine wave. (3) The FFT shows a good 40+ db suppression of the harmonics, especially the odd numbers. So the filter seems to be doing its job. There may be some things not right about my testing of the filter using signal generator and oscilloscope -- I'm new at this. For example, the shoulder on the filter output at about 10 MHz is troublesome, but still the logmag S21 output shouldn't be affected, should it? Any help in understanding why my results look as they do would be greatly appreciated. Best & 73, Mitch NK3H |
Re: Help with LPF Measurement
Hi Mitch. Dave can assist and we have confirmed that you filter in indeed
set up for a 50 ohm source and load termination. The response is validation from a simple AC simulation. I looked at your photo's of your test measurement system and my first comment is: OUCH! Neatness in construction of this RF stuff cannot be over emphasized. Assume that your toroid L's are spot on or at least close to the desired values (you can measure them on your VNA) and the C's all look reasonable... That said, try to build as though it were the final part. Get some RF connectors, SMA or BNC or UHF and apply them to the in and output ports. And then using the appropriate adapter, coax into the VNA test set. Dress those toroid L's up, short the leads and although it does not matter as much with a toroid form, dress then at right angles to each other. Try to get the flow of your construction so that it follows the flow of the schematic as posted. |
Re: Help with LPF Measurement
Hi Dave,
S21 is the gray, I think (the not yellow one -- I'm color blind) print at the top right of the screen in the last two images I posted. It's line is flat at 0 dB across the frequency range with the test rig shorted and gives the unusual curve I showed when the filter is used. Connecting the Ports 0 & 1 of the VNA together (without the test rig) also show 0 dB flat across the frequency range. Is that what you meant? Many thanks. Mitch NK3H |
Re: S-Parameter Plotter
These plots compare reference impedance renormalization using S11, S21, S12, and S22 (yellow) with using just S11 and S21 (cyan). Shown are results for two different 10.7 MHz Murata ceramic filters renormalized from 50 to 330 ohms. The signal passband is 10.6 to 10.8 MHz. The plots demonstrate that when using a NanoVNA that can measure only S11 and S21, you can benefit by reversing the filter and measuring S22 and S12 as well.
Download the S-parameter plotter with renormalization and Y21 method: Brian |
Re: Help with LPF Measurement
Please measure S21 instead of S11 using the NANOVNA! All S11 tells us is
the match (which is pretty bad!). Dave - W?LEV On Thu, Mar 13, 2025 at 12:14?AM Mitch NK3H via groups.io <mitch= [email protected]> wrote: Thank you Alan. I've attached an image of the whole setup including the-- *Dave - W?LEV* -- Dave - W?LEV |
Re: Help with LPF Measurement
Thank you Alan. I've attached an image of the whole setup including the filter, the test rig and the VNA. Please note that I'm just using hookup wire to connect the filter to the testing rig--no SMA or BNC connectors. It made no visible difference when testing on the scope using the wires or direct to a leg of the input and output capacitors. The other image is with a wire connecting the input to the output of the test right. The result is 0db flat across the range for S21. The signal generator has 50 ohm output impedance, and the scope has a very high impedance input. Not sure about the VNA. I tried the setup with a 50 ohm resistor to ground on the filter output and found no difference in the VNA behavior. Thanks for your help. 73, Mitch NK3H
|
Re: Help with LPF Measurement
1) Please post a schematic (already requested).
2) Did you measure the individual components before constructing the filter with them? 3) We assume the filter is designed for 50-ohms in and out? (already assumed) 4) Is the dielectric of the capacitors appropriate for the frequency range? 5) Instead of using the o'scope and FFT, please give us a plot of S21 using the NANOVNA. Please use SAVER (or equivalent) to accomplish your sweep with more points than are available on the native instrument. 6) If your signal generator can sweep, you can also measure the filter response with that. A flat output noise generator can also be used instead of the swept sig. gen. Dave - W?LEV On Wed, Mar 12, 2025 at 11:11?PM Mitch NK3H via groups.io <mitch= [email protected]> wrote: I've searched through the message list looking for symptoms similar to-- *Dave - W?LEV* -- Dave - W?LEV |
Re: Help with LPF Measurement
One last item, on your scope pulse gen measurement. Dial your pulse gen
back to 100 kHz, you should see the fidelity of the square wave as measured at the output load termination R. As you increase frequency, the fidelity of the square wave will suffer, of course, and for the most part you should see a sine wave attenuated at the pulse input frequency. |
Re: Help with LPF Measurement
Confirmed and looks ok... Mitch.
Assume that your layout is such that the inductors are ideally non coupled to each other, ideal each at right angles. The connectors are panel mount? Soldered their body to ground and center conductor to the input pads. A photo of you filter card would be useful. In any case, re calibrate and then check your calibration is ok by checking your through cable fiving a S21 of nearly 0 dB and a S11 of at least 20 dB or better. |
Help with LPF Measurement
I've searched through the message list looking for symptoms similar to mine and didn't find anything. I'm somewhat new to this and would appreciate any help you can offer.
I've been trying to get my NanoVNA-F (firmware 1.0.5) to show the results of a 40-meter low pass filter I've built. I know I'm doing something wrong here but can't figure it out. The VNA image is calibrated (including pass-thru), set for the range 6.5 to 29 MHz and logmag S21. Shouldn't this give a flat area leading up to the design frequency followed by a significant roll off through the remainder of the frequency range? My setup is very similar to W2AEW's YouTube demonstration of the NanoVNA with a LPF but gives entirely different results. The image is entirely different, with very large insertion loss and a steep area of suppression peaking at 11 MHz followed by a steep recovery. I worried that my filter wasn't what I thought it was so I tested it with a 7 MHz square wave, looking at input and output with the FFT function of my 'scope. The before and after images show several things: (1) the square wave (yellow trace) is hardly a square wave. It wasn't much of one before attaching to the input of the filter but it's definitely loaded and distorted by the filter. (I can live with the distortion in the test, since the filter is supposed to reduce the high freq distortion anyway, right?) (2) The purple trace is the filter output--pretty nice sine wave. (3) The FFT shows a good 40+ db suppression of the harmonics, especially the odd numbers. So the filter seems to be doing its job. There may be some things not right about my testing of the filter using signal generator and oscilloscope -- I'm new at this. For example, the shoulder on the filter output at about 10 MHz is troublesome, but still the logmag S21 output shouldn't be affected, should it? Any help in understanding why my results look as they do would be greatly appreciated. Best & 73, Mitch NK3H |
Re: VNA-F V2 brand new screen black
I would return it. Alternatively check with the manufacturer if your unit
is genuine and that their warranty policy is. On Tue, 11 Mar 2025 at 23:23, alsodanev2 via groups.io <alsodanev2= [email protected]> wrote: I got it from an auction site called nellis auctions. They sell surplus |
Re: #firmware #nanovna-h
#firmware
#nanovna-h
I managed to get it work with v1.2.40 firmware from DiSlord and the DfuSe Demo software.
Initially I was using the wrong path to update the firmware, essentially uploading instead of updating, like in the photo. Rookie mistake.
success nanovna .png
![]()
WhatsApp Image 2025-03-11 at 15.24.49_aa8b1993.jpg
![]()
WhatsApp Image 2025-03-11 at 15.25.32_144b0534.jpg
|
to navigate to use esc to dismiss