Ok, "reflection coefficient" on the Smith Chart??
Bingo!! VNA #2 on the Smith Chart I see a 50 ohm load on Open, Short and of course Load.? Calibrated or not.
It's Broke?
George
toggle quoted message
Show quoted text
On 3/2/2022 10:52 PM, WD0AKZ via groups.io wrote:
I'm sorry, I don't understand "reflection coefficient".
I did now Reset Calibration and looked at the LogMag results if that what you were asking:
#1 VNA O= -3db? S- -1db? L= -40db
#2 VNA O= -86db? S= 85DB? l= -88db
tnx
George
On 3/2/2022 10:29 PM, Jim Lux wrote:
On 3/2/22 8:16 PM, WD0AKZ via groups.io wrote:
Yup, I agree... "Something's wrong here, this looks like you're measuring your load."
Nope... "Something's wrong here, did you measure your load by accident? "
"I'd go back and check the load and short measurements (or the cals) "
I repeated this Calibration/test a number of times today, but I just did it again with the same results.
Lets say that I am Calibrating/testing correctly, so on the Open/Short tests, regarding the LogMag results, VNA #1 or #2 or Both are correct?
what readings do you get for all three terminations *without calibration*? - the NanoVNA isn't perfect, but you should get close to 1, 0, and -1 for the reflection coefficient.
And between two units, you'd expect to see similar values.