capacitor leakage go/no-go test
I have various piles of old caps, many NOS, etc. I'm interested in doing a simple go/no-go type of test for leakage.
Have power supplies that can do 0 to 3kV DC. Is it practical to put the cap in series with a microammeter (mechanical, to not risk damage to a DVM) and a current-limit resistor, then just dial up the voltage up to the rating of the cap (or beyond)? One part of my question is: When (for example) a paper coupling cap goes leaky, just how leaky is it? If a (for example) 20uA full-scale meter would not be sensitive enough, perhaps I set up a triode tube with a rather high-value grid R, and see if a leaky cap is able to change the bias. Thanks for any comments! Pete |
On 2022-11-29 9:29 p.m., saipan59 (Pete) wrote:
I have various piles of old caps, many NOS, etc. I'm interested in doing a simple go/no-go type of test for leakage.Mr Carlson's Lab has published a design for a device for making this test that seems to work well (based on his videos). --T Thanks for any comments! |
Hi Pete & All
I, too, am interested in this and have been thinking about building a leakage tester.? I had one years ago, back in the days of valves, that used a 'magic eye' as the leakage detector, although I often suspected that test equipment might have been leakier than the things it was testing!? I shall follow this with interest. -- All the best Richard M0RBF |
开云体育Unless you want to repair early tube equipment you should hardly
need a leakage tester. If you do deal with early tube equipment
then replacing all paper capacitors with plastic film types is a
sensible start. A regular 20K/V meter on a voltage range will
indicate 0-50uA leakage and have series protection. Most bench
DMM's are 10M ohm Rin with a basic 200mV FSD. If you put a suspect cap in series with a DVM and a power source
you can start on a suitable V range then switch to current if
needed. On a voltage range with 10Mohm input you are also seeing
0.1uA/Volt. Putting 1.1 Mohm across the terminals as a shunt gives
1uA/Volt, 100K gives 10uA/Volt. So I don't think a leakage tester
is very useful apart from mains leakage which is another matter. What is FAR more useful is an ESR meter. Most electronic
equipment made in the last 20 years? fails because of
electrolytics loosing capacity through loosing water, often the
can ruptures or detaches. Many PC motherboards kill CPU's because
of cooked electrolytics. The ESR meter allow very fast testing of
electrolytics and the detection of failing parts before they do
extensive damage. Many CCFL's fail because their reservoir
capacitors fail, many wall warts fail for the same reason, the
result is pulsing the equipment at double line frequency which
will often cause a processor to be reset very rapidly. ESR meters measure impedance with a low voltage but relatively
high current signal around 10-50KHz,? Testing is mostly in-place
i.e. mostly no de-soldering is needed, with a good tester it
should be possible to check a capacitor in a few seconds. Regards, Alan G8LCO |
开云体育
Have you checked out videos on Utube from Mr. Carlsons channel. He does one of these?
-=Bryan=-
From: [email protected] <[email protected]> on behalf of Richard via groups.io <hamandchips@...>
Sent: November 30, 2022 2:06 AM To: [email protected] <[email protected]> Subject: Re: [Test Equipment Design & Construction] capacitor leakage go/no-go test ?
Hi Pete & All
I, too, am interested in this and have been thinking about building a leakage tester.? I had one years ago, back in the days of valves, that used a 'magic eye' as the leakage detector, although I often suspected that test equipment might have been leakier than the things it was testing!? I shall follow this with interest. -- All the best Richard M0RBF |
Thanks for the comments/links so far.
The "Mr. Carlson" stuff: I watched his very 'wordy' video, where he repeatedly says how awesome his tester is, how carefully he has tested his design, and he demonstrates it over and over, and shows us how the old magic-eye-based testers are dangerous and inferior. I went to the Patreon page to find out how it works, but found nothing but buttons to click to send $money$. Annoying. I don't care about seeing his Top-Secret schematic, but would anyone be able to explain how his design works? In particular the "forecast" mode - is he doing something very special/clever? Anyway, I gather that the two methods I proposed in the first post are fine. Alan's suggestion of using a VOM sounds practical and easy. Pete |
Just tried a few. Using an HP power supply that does 0-320V, with the current limit set low, and a Simpson 260 VOM on the 50V or 10V scale.
Seems to work well. Here's a pic - good on the left, bad on the right. Some of the 'bad' ones are perhaps marginal - a very small deflection on the meter. BTW, I hope I didn't offend with my criticism of Carlson. He has every right to protect his I.P., to "sell himself", etc. I just exercise my right to ignore his content. Pete |
LOn Wed, 30 Nov 2022 at 02:29, saipan59 (Pete) <saipan1959@...> wrote: I have various piles of old caps, many NOS, etc. I'm interested in doing a simple go/no-go type of test for leakage. Agilent made a couple of high resistance meters.? The 4349A has 1, 2 or 4 channels, depending upon the options installed, but needs an external power supply up to 5 kV. The 4339B is a capable of measuring higher resistances, and has a built in 1 kV power supply. It would be worth looking at the manuals for them, as they cover capacitor testing. They are capable of measuring very low currents - I think down to aA, which I doubt would be necessary for most capacitors.? I was a bit annoyed that I sent two 4339Bs to Keysight UK. Both were very slightly out of specification on the 10 nA range, but neither could be adjusted, despite all adjustments are electronic. ???? There OPA128 IC is designed for electrometers. There used to be some electometers on a PCB from China on eBay, but I can’t find any now.? A fairly standard way of reforming old electrolytic caps is to gradually increase the voltage using a very high resistance, so low current. Personally I just replace them.?
Dave? Dr. David Kirkby, Kirkby Microwave Ltd, drkirkby@... Telephone 01621-680100./ +44 1621 680100 Registered in England & Wales, company number 08914892. Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United Kingdom |
I found the schematic for Carlson's tester on the "not-so-dark web".
It appears that all it really does is compare the test cap to a 66M resistor (three 22M in series), using a couple of 2N3904's configured to be very high impedance.
It uses 27V (three 9V batts in series), so that may be enough to catch 'most' leakage situations (but note that it exceeds the V-rating of some small caps, which is not good!).
I still don't understand how the "forecast" feature might work, and the schematic doesn't appear to have a switch (etc.) for that feature. 'Predicting' a possible cap leakage failure with such a simple scheme seems like snake-oil to me. In the cases where an old magic-eye machine fails to do the right thing, I believe that can be attributed to poor sensitivity in those old designs, and/or components in the tester that are out of spec. Pete |
开云体育
I used to supervise a service team of 14 guys and other things. Carlson would not have been one.? Servicing is about probability, experience and logical inference but it all needs to be based on sound theory, you also need to take less than a month to fix things, glacial is not an acceptable speed setting. Droning? on and on about outer foil? again and again and again shows a lack of focus on what matters at AF & LF in the crap he deals with. But it must impress someone. Capacitors are usualy very reliable but 60 year old paper
dielectrics are often way beyond their use by date, there's not
much point in NOT replacing such parts unless you have a
stationary exhibit in a museum. Any old electrolytic also
qualifies for retirement. Why would you leave a part known to fail
in place only to have to change it some time soon? For a lot of
Tube gear the big issue is with AF coupling caps, there are not
that many, replacing old paper caps with modern plastic film types
is a no-brainer providing that you can get decent replacements. If
you speak to people used to servicing particular gear they may let
you know particular liable to fail parts. Stock faults are common
to all kinds of electronics, there used to be particular brands
which were more liable to cause problems. Long time back there
were some electrolytics in plastic cases that were very
unreliable. One of the causes of electrolytics failing was sweat,
one trace on a foil or paper would condem the part eventually.
Bead tantalum capacitors are another "Replace on Sight" part, they
fail to a very short circuit. On the other hand mil spec tantalums
in cans are very reliable, the flat disk "acid" types are possably
the best electrolytics ever made, now at ransome price levels. An ESR meter needs to deal with series resistances? under 0.2
ohms for larger caps ( e.g. 220-4700 uF). The failure is mainly
due to ripple current at high operating temps or running cheap
electrolytics where higher rated parts are needed. So a useful ESR
meter needs to read down to 0.1 ohms or lower.? It also needs to
be fast settling in a pro setting, waiting more than a second or
two is as useful as toothache when you want to check every at risk
part. I used an Electronic Design Specialists EDS 88A when I was
doing it for real, that was a very effective tool. ESR meters are
absolutely vital when working on switch mode supplies. Now for a
True Story about ESR. There was one product that failed after 6-18 months, we had about 800 in the field. It was a CRT monitor. It would fail with extensive damage to the Flyback Transformer and shorted Horizontal Output transistor.? If you replaced all of the failed parts it would work again then 1-30 days later you needed to replace all of the parts again. The ESR meter found a small 22uF cap in the switching supply that had high ESR. When I checked the part it turned out that the manufacturer had changed the part, the new part had a smaller can.? Smaller can = lower ripple current rating. After a while the capacitor ESR went higher when the product was warm causing the rail to the Horizontal output to increase and do extensive damage. Turns out that some "Value Study" had saved a few cents on one cap without checking the consequences. That cost the Company it's reputation that had developed in 50 years of manufacturing. I still use the EDS88A, but also have a? Vanwell Chinese
multi-function tester, it's surprisingly good for many
semiconductors and will indicate ESR for electrolytics down to 0.1
0hms. It's? too slow for pro use but much better than toothache,
good for hobby users and very inexpensive.But a lot of dirt cheap
Chinese RCL meters that I have seen are very inaccurate. For RCL
indications for rough work? I use a Peak LCR45, that has a good
wide autorange for AF and RF use and is accurate.
Regards, Alan G8LCO |
开云体育A general rule among those of us who restore old electronics equipment to working order is "Don’t waste your time testing wax coated and bumble bee capacitors.”? A bumble bee capacitor has color bands which give the capacitance value in pF using the same color code as resistors. ? 73 (Regards). ? Max K 4 O D S. ? I've Never Lost the Wonder. ? From: [email protected] [mailto:[email protected]] On Behalf Of saipan59 (Pete) ? Just tried a few. Using an HP power supply that does 0-320V, with the current limit set low, and a Simpson 260 VOM on the 50V or 10V scale. |
开云体育To be clear, don’t waste your time testing these capacitors, just replace them. ? 73 (Regards). ? Max K 4 O D S. ? I've Never Lost the Wonder. ? From: [email protected] [mailto:[email protected]] On Behalf Of MAX
Sent: Saturday, December 03, 2022 11:15 PM To: [email protected] Subject: Re: [Test Equipment Design & Construction] capacitor leakage go/no-go test ? A general rule among those of us who restore old electronics equipment to working order is "Don’t waste your time testing wax coated and bumble bee capacitors.”? A bumble bee capacitor has color bands which give the capacitance value in pF using the same color code as resistors. ? 73 (Regards). ? Max K 4 O D S. ? I've Never Lost the Wonder. ? From: [email protected] [mailto:[email protected]] On Behalf Of saipan59 (Pete) ? Just tried a few. Using an HP power supply that does 0-320V, with the current limit set low, and a Simpson 260 VOM on the 50V or 10V scale. |
开云体育Hello Jeff,I understood "bumble bee" capacitors to be a style of capacitor made by Mullard, certain values have colours which are similar to? those of a bumble bee. I see them on ebay advertised as "tropical fish" capacitors, probably for the same reason.? One advertiser is asking ?250 for 4 of them !!? In the hi fi section of course. If he sells them I stand to be a millionaire as I've got 100s NOS. Regards Ken g8beq On 04/12/2022 13:48, Jeff Green wrote:
|