Re: Making a Q-meter / References etc
¿ªÔÆÌåÓýCould larger resistor values be used for measuring larger values of Rs.? I think the answer is yes.? If the signal source is an RF signal generator the load on it would need to always be 50 ohms.? ? 73 (Regards). ? Max K 4 O D S. ? I've Never Lost the Wonder. ? From: [email protected] [mailto:[email protected]] On Behalf Of Mikek ? Sorry forgot my proposed Layout. |
Re: Making a Q-meter / References etc
Got it, Attached is SimSmith solution with measured V to Q equation.
toggle quoted message
Show quoted text
Equations are: XL=2*Pi*G.MHz*L1.H*1e6; Rs=(R1.V.M*50)/(2-R1.V.M); Rs=1/(1/Rs-1/50); Q=XL/Rs; John KN5L On 8/28/22 3:37 PM, John KN5L wrote:
Added a 50 Ohm load to BNC connector fixture shown in: |
Re: Making a Q-meter / References etc
Added a 50 Ohm load to BNC connector fixture shown in:
Section: "Voltage Ratio Q Measurement" With Generator set to 1Vrms output, Oscilloscope measures = 49.5mV Using SimSmith model to reverse out Q results with Q = 110. Need a mV and L to Q formula such that Q can be computed within SimSmith. John KN5L |
File /Q-factor (Q is the inverse of dissipation factor DF)./Method simplifies testing high-Q devices; Alan Victor, EDN Feb 2002.pdf uploaded
#file-notice
Group Notification
The following files and folders have been uploaded to the Files area of the [email protected] group. By: Ron <roncarlson.email@...> Description: |
Re: Making a Q-meter / References etc
Use the five-sided box to mount everything, not the lid. Pete On Sun, Aug 28, 2022 at 5:52 PM Mikek <amdx@...> wrote: I want to try this, since I now have an HP 3400 RF voltmeter. It will only go to 10 MHz. I'll use my scope after that. |
Re: Making a Q-meter / References etc
On Sun, Aug 28, 2022 at 06:07 AM, Mikek wrote:
My scanner is finicky these days, I took pictures.I converted your photos into a PDF using the free open-source GIMP. I put it here in the files section of the group /g/Test-Equipment-Design-Construction/files/Q-factor%20%28Q%20is%20the%20inverse%20of%20dissipation%20factor%20DF%29./Method%20Simplifies%20high-Q%20devices%20EDN%20Magazine%201st%20Feb%202002%20by%20Alan%20Victor.pdf |
Re: Making a Q-meter / References etc
SimSmith model attached,
toggle quoted message
Show quoted text
John KN5L On 8/28/22 12:22 PM, Mikek wrote:
I'm missing a detail. It looks like we need to set the 1 Volt input |
Re: Making a Q-meter / References etc
I'm missing a detail. It looks like we need to set the 1 Volt input without the DUT connected, then insert the DUT and tune for lowest voltage.
(no need for constant voltage source as I said before) ?Is there an edit function with this software? Or delete? ?Help, with best RF practice. ??????????????????? Mikek ?Going to clean my bench so I can get started. |
Re: Making a Q-meter / References etc
I want to try this, since I now have an HP 3400 RF voltmeter. It will only go to 10 MHz. I'll use my scope after that.
?I'll be using a 4" x 2" x 1" tall cast box.See attached pictures. I'll install Teflon* stand offs to solder DUT to. *(if I find them) I'll put BNC connectors on the box. Some questions about best RF practices. 1) Do I want the 50¦¸ termination internal to my box or at the end of the cable going to the meter. (high input impedance meter) 2)? Can ground to the box or should I put in copper straps. 3) Tell me what I should do different. ?????????????????????? Mikek ? |
Re: Making a Q-meter / References etc
Appended to:
is section "FT114-31 9T SRF Q Measurement" which demonstrates measuring a high inductance Ferrite inductor at Self Resonant Frequency (SRF). This measurement demonstrates the difficulty measuring Ferrite inductors using S11 Real and Imaginary measurements. Inductive reactance, ImagZ, is invalid above about 4MHz. Above 4MHz, ImagZ is sum of inductive and capacitive reactance. John KN5L |
Re: Making a Q-meter / References etc
On Sun, 28 Aug 2022 at 13:26, Mikek <amdx@...> wrote: I found this in my files while looking for something else. I would appreciate a scan if you can do so. I can't get images to load either, so it's essentially impossible to follow the description. |
Re: Making a Q-meter / References etc
I found this in my files while looking for something else.
Method Simplifies Testing High-Q devices, Alan Victor. https://www.edn.com/method-simplifies-testing-high-q-devices/ I can't get the Figure 1, Table 1, and figure 2 to load. If anyone is interested I have a paper copy and can upload it to Dropbox. It's just a series LC circuit driven by 50¦¸ and loaded with 50¦¸. Voltage is read with a voltmeter, and some voltage divider math gets you notch depth in db that converts to R of the inductor. |
Re: Optimal full scale deflection on analog meters
Rodger Bean
¿ªÔÆÌåÓýThe Sangamo Weston S 82 galvanometer pictured has a significant overshoot when given a step increase in current. The coil is 146¦¸ and it is recommended that a 2K¦¸ resistor be put in parallel with it, if the galvo is being used with a high source impedance. For those who revel in the mathematics of galvo damping. There is an interesting paper here: ? ? Rodger Bean ? From: [email protected] <[email protected]> On Behalf Of Pete_G4GJL
Sent: Thursday, 25 August 2022 22:30 To: [email protected] Subject: Re: [Test Equipment Design & Construction] Optimal full scale deflection on analog meters ? No surprise, as the movement would normally be damped by the very low resistance of the thermocouple to which it should be connected. With the data you have, make a 0-1 Amp ammeter and test that with a 1/2 FSD current. You will find it exhibits 'normal' damping. That is unless you have a thermocouple to try it out on! ? Pete G4GJL ? On Wed, Aug 24, 2022 at 9:34 PM Mikek <amdx@...> wrote:
|
Re: Making a Q-meter / References etc
Would it not be possible to come up with a temperature stable (within reasonable room temp) in a 1" x 1" size or however you make it low inductance and make a Kelvin connection to that.
?Have a switch to set the drive current, that also sets the gain of the Q meter amp that is driven by the Kelvin connection. I'm assuming a parallel connection, although I do see one advantage of a series connection and that is less interference from local AM stations with big coils. ?Tell me where I'm wrong on this so I can learn. ??????????????????????????????????? Thanks, Mikek As far as overwhelming response, I can't get enough! |