2465B wobbly display
First things first:
toggle quoted message
Show quoted text
When you get a new piece of test equipment, you should always measure the power supply voltages, and check the ripple. When doing the ripple measurements, you will need a scope, and you must jumper the two scope's ground jacks together. Second, the vertical and horizontal amplifiers can create jitter due to the instantaneous heating of the parts in their respective hybrids. This jitter can be balanced out using the jitter adjustments as described in the calibration section of the manual. The adjustment won't bother the calibration significantly. If the ripple isn't coming from the power supply, nor the jitter, you can determine if it is mains related by placing the trigger in the LINE position. If the ripple stops drifting, it is MAINS related. I don't usually find the big electrolytics that filter the rectified power line to be a problem with the 2465, but some do. I almost always find the filter for the +15V supply to be bad. It won't cause ripple, though. -Chuck Harris On Thu, 16 Jun 2022 11:57:17 -0700 "Richard" <mole42@...> wrote:
I've recently acquired a 2465B. The instrument powers up okay but the |
Chuck, great to see your excellent notes on general troubleshooting...long time, How are you doing?
Agree on the bus and 15 V PSU caps notes. The bus caps are ample for 50 or 60 Hz mains. Could the " wobbly" traces be the readout multiplexing, eg as sweep rate changes, the readouts move slightly to prevent burnin? Bon courage Jon The OP should send some scope photos or post a video to clarify the problem.... |
Unfortunately I'm not able to check voltages right now, the instrument is in my shop and my tools are in my workshop - a few miles away.....
Tests reveal that the wobble is in the vertical amplifier, horizontal is stable. It looks like 50Hz noise. Focus & Intensity controls work as expected, Sweep & Channel voltage selection work, most other controls seem to work properly. I will check the power supply over the weekend and report any findings. Richard |
Hi Jon,
I am well. The 2467's have the feature of moving the readout to prevent burning the rather unique MCP (Micro Channel Plate) CRT, but that feature is disabled when the firmware is used with a 2465B configured motherboard. -Chuck Harris On Thu, 16 Jun 2022 23:01:46 -0700 "Froggie the Gremlin" <jonpaul@...> wrote: Chuck, great to see your excellent notes on general |
Hi Richard,
toggle quoted message
Show quoted text
I cannot think of any way you could have 50Hz mains injected into the vertical amplifier (after the channel mux), and not have it injected into the horizontal amplifier. Check the vertical jitter adjustment. It affects everything vertical. -Chuck Harris On Fri, 17 Jun 2022 03:49:18 -0700 "Richard" <mole42@...> wrote:
Unfortunately I'm not able to check voltages right now, the |
¿ªÔÆÌåÓýAny large Transformers or other AC magnetic sources in the local vicinity of the Oscilloscope? ? You might move the Oscilloscope away from the area and try it to see if the problem goes away. ? Ross ? From: [email protected] <[email protected]> On Behalf Of David
Sent: Friday, June 17, 2022 9:13 AM To: [email protected] Subject: Re: [TekScopes2] 2465B wobbly display ? 50Hz coupling through the supply somehow if it affects everything. |
An update since I've done some investigation. The power supply test point measures 10.0047 volts which is pretty close to spec. The vertical jitter control does not improve the display but it does worsen the vertical jitter in either direction. There is evidence of knob twiddler activity around the vertical jitter pot. The power supply is dust-filled and has not been re-capped.
That's all for now, I'm busy for the next few days so I'll not be able to dig any deeper. Richard |