Search
Re: Ramtron chip in place of Dallas...
Jan J
I'm recent here... Can you provide more info on how-to?
toggle quoted message
Show quoted text
On 12/8/2017 8:26 AM, Mark Wendt wrote:
On 12/07/2017 02:33 PM, Jan J wrote:OK... Now how do we get files to you?? Do you have an FTP site?Upload them here to the group's Files section? |
Re: Tektronix 2225 power supply startup ticking / buzzing
Chris Smith
Hi all,
Just an update on this. I finally got around to doing cal on this 2225 and it's quite clearly still a tiny bit sick in the power supply department. The 8.6v rail has a ridiculous amount of ripple on it which is presenting itself on the display at 5/10mV vertical positions. I triggered my other scope off the power supply inverter and the ripple is in sync at the same frequency and measurable at the vertical amp. Ergo, I have decided to replace all of the remaining Philips 030 electrolytic capacitors in the unit. The other ones I pulled had gone high ESR. Hopefully it will go away. They're dotted all around the unit so I'm going to have to strip the panel and the attenuator board out. Capacitor whack-a-mole worked nicely for my 465B and 475. I'll replace the ones I can get to first before I strip it and then re-test.? Best regards, Chris |
Re: Ramtron chip in place of Dallas...
Jan J
OK... Now how do we get files to you?? Do you have an FTP site?
toggle quoted message
Show quoted text
On 12/7/2017 2:32 PM, Sergey Kubushyn wrote:
On Thu, 7 Dec 2017, Jan J wrote:I will give you the latest software I received if you want when the time comes... Warning!!?? It runs under DOS... (And it Won't work on DOS emulator from within Windows-- I tried!).I would like to get is as soon as possible...I booted off CD of DOS 6.22, and ran program off floppy disk... The system you use will also have to have a serial port, and you'll connect to the scope via the 9 pin serial connection on scope.That is not a problem, I do have a real DOS 6.22 machine.So you'll need DOS, a floppy disk, and Serial port with cable, and a programmer that supports the 1225 or 16W08 chip...? And be able to unsolder old 1225 and put a machine pinned IC Socket (Make sure it is not a thick one, or 1225 on this socket will stick down too low, and it won't go back in Rackmount).That's not a problem either -- I have proper desoldering station as well as |
Re: Ramtron chip in place of Dallas...
On Thu, 7 Dec 2017, Jan J wrote:
I will give you the latest software I received if you want when the time comes... Warning!!?? It runs under DOS... (And it Won't work on DOS emulator from within Windows-- I tried!).I would like to get is as soon as possible... I booted off CD of DOS 6.22, and ran program off floppy disk... The system you use will also have to have a serial port, and you'll connect to the scope via the 9 pin serial connection on scope.That is not a problem, I do have a real DOS 6.22 machine. So you'll need DOS, a floppy disk, and Serial port with cable, and a programmer that supports the 1225 or 16W08 chip...? And be able to unsolder old 1225 and put a machine pinned IC Socket (Make sure it is not a thick one, or 1225 on this socket will stick down too low, and it won't go back in Rackmount).That's not a problem either -- I have proper desoldering station as well as EETools TopMaxII programmer. You'll have to have a good Color Bar test signal generator, and unsolder the 1225 chip, and save it with a programmer to disc file....I do have a good test signal generator, no problems. Find the Service Manual online for your scope NOW... You will be referring to it... for switch settings for use with serial port, and serial port cable pinout.OK, will try to find it later tonight when I'm back home and check what intruments I actually have. You should have a spare 1225 or Ramtron on DIP chip socket... and a programmer,? instead of leaving the 1225 soldered to? main board, (as if you don't get it right with the software, you'll have a dead scope)No problems here... Save your original file with programmer.? Dumb luck was the only thing that saved me from over-writing my stock BBRAM...? When you run the CAL program off disc, after selecting the type of scope you are using, program will ERASE the chip in the scope, and from then on, you'll be faced with walking through all the 80 calibration steps, which have sub-steps...? Only when you finish all the steps will it write to the Scope Ram....? And when you finish that, then and only then will you know if it works.... or not, and if it's NOT, you'll wish you had saved BBRAM to disc. Get a programmer if you don't have one.Actually I'm not sure that BBRAM is still alive in my instruments -- it might be already dead... Will have to check. If it works after you run CAL -- and SCH, and waveform and vector and component and other signals look OK, then your finished.... In my case, I had to Hack it in.....? I powered up scope with NO BBRAM in socket.... and discovered that SCH worked....? SO I programmed in FF in half locations, then 1/4, 1/8, etc... until I found the one location where an "F" allowed SCH operation....--- * * KSI@home KOI8 Net < > The impossible we do immediately. * * Las Vegas NV, USA < > Miracles require 24-hour notice. * * |
Re: Ramtron chip in place of Dallas...
Jan J
I will give you the latest software I received if you want when the time comes... Warning!!?? It runs under DOS... (And it Won't work on DOS emulator from within Windows-- I tried!).
toggle quoted message
Show quoted text
I booted off CD of DOS 6.22, and ran program off floppy disk... The system you use will also have to have a serial port, and you'll connect to the scope via the 9 pin serial connection on scope. So you'll need DOS, a floppy disk, and Serial port with cable, and a programmer that supports the 1225 or 16W08 chip...? And be able to unsolder old 1225 and put a machine pinned IC Socket (Make sure it is not a thick one, or 1225 on this socket will stick down too low, and it won't go back in Rackmount). You'll have to have a good Color Bar test signal generator, and unsolder the 1225 chip, and save it with a programmer to disc file.... Find the Service Manual online for your scope NOW... You will be referring to it... for switch settings for use with serial port, and serial port cable pinout. You should have a spare 1225 or Ramtron on DIP chip socket... and a programmer,? instead of leaving the 1225 soldered to? main board, (as if you don't get it right with the software, you'll have a dead scope) Save your original file with programmer.? Dumb luck was the only thing that saved me from over-writing my stock BBRAM...? When you run the CAL program off disc, after selecting the type of scope you are using, program will ERASE the chip in the scope, and from then on, you'll be faced with walking through all the 80 calibration steps, which have sub-steps...? Only when you finish all the steps will it write to the Scope Ram....? And when you finish that, then and only then will you know if it works.... or not, and if it's NOT, you'll wish you had saved BBRAM to disc. Get a programmer if you don't have one. If it works after you run CAL -- and SCH, and waveform and vector and component and other signals look OK, then your finished.... In my case, I had to Hack it in.....? I powered up scope with NO BBRAM in socket.... and discovered that SCH worked....? SO I programmed in FF in half locations, then 1/4, 1/8, etc... until I found the one location where an "F" allowed SCH operation.... In the case of the 1760 scope, that was address 0F of the BBRAM...? I changed the stock data "O" to "F" and SCH then worked.... Minor gain tweeks and configs were done, though I had to replace R145 on SCH board with 200K pot, and tweeked SCH Gain that way..... On 12/7/2017 1:26 PM, Sergey Kubushyn wrote:
On Thu, 7 Dec 2017, Jan J wrote: |
Re: Ramtron chip in place of Dallas...
On Thu, 7 Dec 2017, Jan J wrote:
I do also have 1761 without SCH and 1750 (or whatever -- don't remember right now) with SCH board so I would like to make a PAL/NTSC 1761 (if my dementia serves me right -- it's been couple of years since I got those and didn't have a chence to work on) with SCH so I might ask some questions when it finally got to it. Don't know what I could use that for if at all but it is always better to have an instrument on hand in case it would be needed and that instrument better have all the options available :) Glad to see someone who understood the hoops I jumped through!!!--- * * KSI@home KOI8 Net < > The impossible we do immediately. * * Las Vegas NV, USA < > Miracles require 24-hour notice. * * |
Re: Ramtron chip in place of Dallas...
Jan J
Glad to see someone who understood the hoops I jumped through!!!
toggle quoted message
Show quoted text
On 12/7/2017 9:08 AM, Burt K6OQK wrote:
Jan, |
Re: Ramtron chip in place of Dallas...
Absolutely!
Burt, K6OQK At 07:36 AM 12/7/2017, Chuck Harris wrote Just to remind folks, this kind of stuff is exactlyBurt I. Weiner Associates Broadcast Technical Services Glendale, California U.S.A. biwa@... K6OQK |
Re: Ramtron chip in place of Dallas...
Just to remind folks, this kind of stuff is exactly
toggle quoted message
Show quoted text
the reason this group exists. We are all about tektronix, and other stuff that interests us. What interests us? You decide. -Chuck Harris Burt K6OQK wrote: Jan, |
Re: Ramtron chip in place of Dallas...
Jan,
toggle quoted message
Show quoted text
I enjoy your "wordy" offerings. It's like sitting down with a good book. Burt, K6OQK :) At 04:48 AM 12/7/2017, Jan J wrote When I found the SCH option board from a 1750 scope, I contacted Tek to see if it was a 'drop in' replacement. They said 'Yes'.. Really it's "NO" I was then given some software (DOS) that would 'calibrate'---- a nearly 70 step program that would Create the BBRAM settings for this scope-If you had to replace the BBRAM chip. With a new 1225 chip, I practiced--repeatedly with the program, and I got pretty good at setting up Waveform and Vector display, but I had no component signals to use, and it didn't cover the SCH Option board I was trying to install during the calibrate program, nor after I was done with calibrate and stored the settings to the BBRAM... SCH Option (When I hit the button) did not work.... By this time I was communicating with a thread at TEK, and I kept revisiting the subject of: How can I ENABLE the SCH Option? The 3 or so people that did reply from that thread, were continually referencing to "1760SC" Which I could not find in the software I received from them.... During this time I was also attempting to upgrade from V2.3 to V2.4 firmware in the scope, as one of them said that would do it... but the Upgrade software sent would not erase the NVRAM.... Finally, a fellow who had never replied to me from work, did so from home, and pointed me to a TEK U.K. site where the same software that I had been sent before was posted in a public download.... I downloaded it and found out that it was a LATER version of the same software sent to me before, but this one had 80 steps in the CAL Program, and included references to a "SCOPE BUILD" Document, and covered all the 1700 scope family, and a reference to a TEK Document "See PRC_LANG.DOC for instructions on building Model, Option, and Title codes from the key...." and the following information: 01 NTSC 02 PAL 04 DUAL 10 MISC STEPS (Main Board) 20 VECTOR BOARD 40 SCH BOARD 80 COMPONENT BOARD and a rem statement of the models: TEK1740A 31 TEK 1741A 32 TEK1745A 34 TEK 1750A 71 TEK 1751A 72 TEK 1755A 74 TEK 1760 B1 TEK1761 B2 TEK 1765 B4 TEK 1760SC F1 TEK1761SC F2 TEK1765SC F4 Armed with this information, I contacted everyone at TEK that I had contacted before, asking for the PRC_LANG.DOC..... At this point, ALL Communications with the folks on that support Link STOPPED!!!! After waiting a week... I did an online chat... and was abruptly told I would NOT Get that document.... and demanded to know how I found out about it? "Well, it's a remark statement within the software you released to the public" The chat link abruptly ended.... !!! I then called Support, and was told that that document: PRC_LANG.DOC was a COMPANY DOCUMENT for Company Personnel only.... and I will never get it... That was the end of my support ---- Support!!! I was officially on my own... By this time, I had a couple spare 1225 chips, and was awaiting the Ramtoron 16W08 on a DIP Header... and started "doing experiments" to see if I could find a way to Enable the SCH board from 1750 scope in my 1760 scope... and I in-advertantly powered up scope WITHOUT the 1225 in the socket, and though the gain calibrations were incorrect---SCH OPTION BOARD WORKED!!!! I then purchased a Programmer, read the contents of "stock" 1225 chip, and then by using 'Powers of 2' I put "F" in the first half of BBRAM contents, and put that in scope... SCH WORKED! So I then walked the "F" back until I found the ONE LOCATION (Address 0F) where an "F" would allow me to use the SCH Option.... With this (by then Ramtron chip on Dip Header arrived), I then touched up some gains, and scope was good except for SCH Gain on CRT..... (For some reason--1750 SCH Option board in 1760 scope with Firmware Hack to enable SCH Option board would not adjust SCH Gain in the updated CAL Program the guy pointed me to in the U.K.) I fixed this by putting a 200K pot in place of R145 (33.2K).... and adjusting SCH Gain.... So, even though TEK Said it would be Plug and Play.... It sure as hell wasn't.... and I had to HACK the BBRAM in order to enable the SCH Option board from a 1750 scope to work in a 1760 scope.... Sorry for being so wordy.... but what a "Long, Strange Trip this has been!!!" Jan On 12/6/2017 7:51 AM, Chuck Harris wrote:
Tektronix was always good about sharing parts and ideas among their > product line. Looks to me like it could share a lot in common with > the 2465 family. Not bandwidth, so much, but certainly CRT drive and > control, CPU, front panel control mechanisms... > > I'm pretty sure that most anything that uses a 1225 would rely upon > its battery, as that is the only way it has of preserving data. The > earlier 2465B, and the 2465A used a separate battery, control, and > CMOS RAM to provide non volatile storage. > > Thanks for enlightening me. > > -Chuck Harris > > Jan J wrote: >> Does Composite, Component Waveform and Vector/Diamond display.. >> Plus, Time Code, Audio, and SCH.. SCH is an option board... >> >> I don't know if this forum supports pictures... if it does, here is >> what a good SCH Display looks like..... >> >> On 12/5/2017 9:21 PM, Chuck Harris wrote: >>> Obviously. >>> >>> I was simply expanding on what you stated to include the early >>> 2465B and 2465A. I can't say that I have ever run into the 1700 >>> series.. Isn't it a vector scope, or something like that? >>> >>> -Chuck Harris >>> >>> Jan J wrote: >>>> Also, the Tek 1700 series relies on battery within 1225 only... >>>> no other battery... > > > Burt I. Weiner Associates Broadcast Technical Services Glendale, California U.S.A. biwa@... K6OQK |
Re: Ramtron chip in place of Dallas
Jan J
I started this project about 4-5 months ago....
toggle quoted message
Show quoted text
The reasons for adding SCH Option are as strange as the trip I took getting it to work... I am a retired TV Maintenance Engineer... (ABC). Also a Video Hobbyist. I was given some old broadcast gear that would improve video capturing, and it could use a Genlock Reference Signal to 'lock' all devices to a common reference. I had a black generator on shelf for years, that I had never used... and when I did use it... Captures started failing... coming out in 'Reverse Field Dominance' in SD and 1080i HD... I got around the problem by purchasing another black generator... That did work....?? But I wanted to know WHY did the first Black Generator cause the failures in Field Dominance????? I'm Nosy that way!!! :)! The 1760 scope I had did not 'see' anything wrong, and I suspected SCH error might be the cause, but I had no way of measuring that.... (SCH is the phase of Color Burst compared to Leading Edge of Horizontal Sync in Composite signals)..... I contacted TEK Support and asked if a 1750SCH Option board would work in a 1760 Scope, and they said "YES".? Armed with this, I started looking for an SCH Option board... Since TEK told me it would work.... (Ha-Ha!) I eventually came across a 1750 scope that had a bad CRT, but the fellow was willing to trade me the SCH Option board out of it, if I captured HD Video off his HD Tape machine and? made a BluRay Disk of it for him... Armed with the Option board, I put it in the 1760 scope, and it would not enable from the SCH Button on the front panel...... And That's where this project began!!!!! OH YES!!!? SCH measurements on that Black Generator revealed there was Great JITTER on the SCH measurements... Jitter that was not 'seen' on 1760 Waveform or Vector display.... Jitter that only could be seen by an SCH Measurement!!! I took that Black Generator with bad SCH to the local Dump, and just happened to be there when they were shredding TV's in a great grinder.... I told the operator why I was there, and asked if I could personally pitch this generator in the shredder, and he said "Go For It!!!" ..... That Generator is now in hundreds of thousands of pieces, and will never cause 'Reverse Field Dominance'? Ever again!!!! Jan On 12/7/2017 8:03 AM, Chuck Harris wrote:
Hi Jan, |
Re: Ramtron chip in place of Dallas
Hi Jan,
That is truly strange. Are the later models 176xSC still under Tek support? And I suppose it is possible that there was some customer proprietary information in that setup program that may have triggered their sudden reluctance to share further information.... NSA comes to mind. Anyway, it sounds like your hard work paid off. -Chuck Harris Jan J wrote: When I found the SCH option board from a 1750 scope, I contacted Tek to see if it was... |
Re: Ramtron chip in place of Dallas
Jan J
When I found the SCH option board from a 1750 scope, I contacted Tek to see if it was a 'drop in' replacement. They said 'Yes'.. Really it's "NO"
I was then given some software (DOS) that would 'calibrate'---- a nearly 70 step program that would Create the BBRAM settings for this scope-If you had to replace the BBRAM chip. With a new 1225 chip, I practiced--repeatedly with the program, and I got pretty good at setting up Waveform and Vector display, but I had no component signals to use, and it didn't cover the SCH Option board I was trying to install during the calibrate program, nor after I was done with calibrate and stored the settings to the BBRAM... SCH Option (When I hit the button) did not work.... By this time I was communicating with a thread at TEK,? and I kept revisiting the subject of: How can I ENABLE the SCH Option?? The 3 or so people that did reply from that thread, were continually referencing to "1760SC"?? Which I could not find in the software I received from them....? During this time I was also attempting to upgrade from V2.3 to V2.4 firmware in the scope, as one of them said that would do it... but? the Upgrade software sent would not erase the NVRAM.... Finally, a fellow who had never replied to me from work, did so from home, and pointed me to a TEK U.K. site where the same software that I had been sent before was posted in a public download.... I downloaded it and found out that it was a LATER version of the same software sent to me before, but this one had 80 steps in the CAL Program, and? included references to a "SCOPE BUILD" Document, and covered all the 1700 scope family, and a reference to a TEK Document? "See PRC_LANG.DOC for instructions on building Model, Option, and Title codes from the key...." and the following information: 01 NTSC 02 PAL 04 DUAL 10 MISC STEPS (Main Board) 20 VECTOR BOARD 40 SCH BOARD 80 COMPONENT BOARD and a rem statement of the? models: TEK1740A ? 31 TEK 1741A? 32 TEK1745A?? 34 TEK 1750A? 71 TEK 1751A? 72 TEK 1755A? 74 TEK 1760???? B1 TEK1761????? B2 TEK 1765??? B4 TEK 1760SC F1 TEK1761SC? F2 TEK1765SC? F4 Armed with this information, I contacted everyone at TEK that I had contacted before, asking for the PRC_LANG.DOC..... At this point, ALL Communications with the folks on that support Link STOPPED!!!! After waiting a week... I did an online chat... and was abruptly told I would NOT Get that document.... and demanded to know how I found out about it? "Well, it's a remark statement within the software you released to the public" The chat link abruptly ended....? !!! I then called Support, and was told that that document: PRC_LANG.DOC was a COMPANY DOCUMENT for Company Personnel only.... and I will never get it... That was the end of my support ---- Support!!!?? I was officially on my own... By this time, I had a couple spare 1225 chips, and was awaiting the Ramtoron 16W08 on a DIP Header...? and started "doing experiments" to see if I could find a way to Enable the SCH board from 1750 scope in my 1760 scope...? and I in-advertantly powered up scope WITHOUT the 1225 in the socket, and though the gain calibrations were incorrect---SCH OPTION BOARD WORKED!!!! I then purchased a Programmer, read the contents of "stock" 1225 chip, and then by using 'Powers of 2'? I put "F" in the first half of BBRAM contents, and put that in scope... SCH WORKED! So I then walked the "F" back until I found the ONE LOCATION (Address 0F) where an "F" would allow me to use the SCH Option.... With this (by then Ramtron chip on Dip Header arrived), I then touched up some gains, and scope was good except for SCH Gain on CRT..... (For some reason--1750 SCH Option board in 1760 scope with Firmware Hack to enable SCH Option board would not adjust SCH Gain in the updated CAL Program the guy pointed me to in the U.K.)?? I fixed this by putting a 200K pot in place of R145 (33.2K).... and adjusting SCH Gain.... So, even though TEK Said it would be Plug and Play.... It sure as hell wasn't.... and I had to HACK the BBRAM in order to enable the SCH Option board from a 1750 scope to work in a 1760 scope.... Sorry for being so wordy.... but what? a? "Long, Strange Trip this has been!!!" Jan On 12/6/2017 7:51 AM, Chuck Harris wrote: Tektronix was always good about sharing parts and ideas among their > product line. Looks to me like it could share a lot in common with >the 2465 family. Not bandwidth, so much, but certainly CRT drive and > control, CPU, front panel control mechanisms... > > I'm pretty sure that most anything that uses a 1225 would rely upon > its battery, as that is the only way it has of preserving data. The > earlier 2465B, and the 2465A used a separate battery, control, and > CMOS RAM to provide non volatile storage. > > Thanks for enlightening me. > > -Chuck Harris > > Jan J wrote: >> Does Composite, Component Waveform and Vector/Diamond display.. >> Plus, Time Code, Audio, and SCH.. SCH is an option board... is >> what a good SCH Display looks like..... >> >> On 12/5/2017 9:21 PM, Chuck Harris wrote: >>> Obviously. >>> >>> I was simply expanding on what you stated to include the early >>> 2465B and 2465A. I can't say that I have ever run into the 1700 >>> series.. Isn't it a vector scope, or something like that? >>> >>> -Chuck Harris >>> >>> Jan J wrote: >>>> Also, the Tek 1700 series relies on battery within 1225 only... >>>> no other battery... > > >I don't know if this forum supports pictures... if it does, here |
Re: Ramtron chip in place of Dallas
One of my customer's 2465B's was really a 2445B, ala the
toggle quoted message
Show quoted text
ebay guy's handiwork. It was an early model that had the CMOS RAM and a battery. The handyman removed the CMOS RAM, left the battery and all of its support circuitry in place, and replaced the RAM with a 1225. I guess his idea was that the battery would keep the alert off until the warranty expired. I suppose it is possible he was just asleep when he did the work too... Well, the DS chip really hated having its power stuck at 3.6V, which was a brown out state, and it drew about 0.1ma from VCC in that condition. The DS died early, the battery died, and a new calibration was in order. I think that was the first scope I converted to an FRAM. -Chuck Harris David Hess wrote: When I changed the Dallas NVSRAMs in my 2440, I did some experiments |
Re: Ramtron chip in place of Dallas
When I changed the Dallas NVSRAMs in my 2440, I did some experiments
toggle quoted message
Show quoted text
to see if they could be used as low power battery backed up SRAMs by maintaining the supply voltage above the point where the internal battery would discharge but below the point where the NVSRAMs would be active. There was no such point and the supply current required would have been way too high to use an external battery for more than weeks to months. My backup plan was to restore the original design with the separate battery, control, and low power SRAM. At least on the 2440, the printed circuit boards were the same but the parts were changed or removed. As it ended up, the Cypress STK16C88 EEPROM based NVSRAMs that I used were drop in replacements and worked perfectly so I did not get to trying FRAM. On Wed, 6 Dec 2017 07:51:56 -0500, you wrote:
Tektronix was always good about sharing parts |
Re: Ramtron chip in place of Dallas
Tektronix was always good about sharing parts
toggle quoted message
Show quoted text
and ideas among their product line. Looks to me like it could share a lot in common with the 2465 family. Not bandwidth, so much, but certainly CRT drive and control, CPU, front panel control mechanisms... I'm pretty sure that most anything that uses a 1225 would rely upon its battery, as that is the only way it has of preserving data. The earlier 2465B, and the 2465A used a separate battery, control, and CMOS RAM to provide non volatile storage. Thanks for enlightening me. -Chuck Harris Jan J wrote: Does Composite, Component Waveform and Vector/Diamond display.. Plus, Time Code, |
Re: Ramtron chip in place of Dallas
Jan J
Does Composite, Component Waveform and Vector/Diamond display.. Plus,? Time Code, Audio, and SCH.. SCH is an option board...
toggle quoted message
Show quoted text
I don't know if this forum supports pictures... if it does, here is what a good SCH Display looks like..... On 12/5/2017 9:21 PM, Chuck Harris wrote:
Obviously. |
Re: Ramtron chip in place of Dallas
-CS does more than latch the address. Like DRAM, FRAM uses sense
toggle quoted message
Show quoted text
amplifiers which must be precharged before each read operation. When -CS is high, the sense amplifiers are precharged in preparation for the next read operation. On Tue, 5 Dec 2017 21:17:25 -0500, you wrote:
You got lucky, as the 1225 chip behaves exactly like |
Re: Ramtron chip in place of Dallas
Obviously.
toggle quoted message
Show quoted text
I was simply expanding on what you stated to include the early 2465B and 2465A. I can't say that I have ever run into the 1700 series.. Isn't it a vector scope, or something like that? -Chuck Harris Jan J wrote: Also, the Tek 1700 series relies on battery within 1225 only... no other battery... |
to navigate to use esc to dismiss